Re: [ccp4bb] Nanodrop versus Nanophotomter Pearl versus good old Bradford.

2011-06-17 Thread Arnon Lavie
To add to this discussion: Many of the comments to my question that started this thread do not sufficiently differentiate between accuracy and precision. While we all want an assay that is internally consistent (i.e., high precision), we do care a lot about accuracy ("the degree of closeness of

Re: [ccp4bb] Nanodrop versus Nanophotomter Pearl versus good old Bradford.

2011-06-16 Thread Jan Dohnalek
For accurate absorbance measurements we have been using a 10 ul "tunnel" cuvette for some time. It is still 1 cm pathlength, fits into standard specs, one can manage even with 8 ul of a sample fed into the channel, which can be rescued ~ 80%, no evaporation issues, the cuvette has to be kept clean!

Re: [ccp4bb] Nanodrop versus Nanophotomter Pearl versus good old Bradford.

2011-06-16 Thread Richard Edward Gillilan
Hi Chelsy, yes we had a lot of trouble with the nanoview during that run. Even after going through the calibration procedure with the special fluid provided, we still had inconsistent results even on standards. Finally, I carefully cleaned the return light path of the instrument (a separate s

Re: [ccp4bb] Nanodrop versus Nanophotomter Pearl versus good old Bradford.

2011-06-16 Thread Chelsy Prince
time that I need to use it? Thanks, Chelsy From: CCP4 bulletin board [mailto:CCP4BB@JISCMAIL.AC.UK] On Behalf Of Pascal Egea Sent: June-16-11 8:18 PM To: CCP4BB@JISCMAIL.AC.UK Subject: Re: [ccp4bb] Nanodrop versus Nanophotomter Pearl versus good old Bradford. I would like to add some

Re: [ccp4bb] Nanodrop versus Nanophotomter Pearl versus good old Bradford.

2011-06-16 Thread Machius, Mischa Christian
On Jun 16, 2011, at 8:30 PM, Dima Klenchin wrote: You recommended "determining extinction coefficients experimentally". How is plugging number of specific residues into a formula constitute experimental determination? That is a deeply philosophical question! Eventually, you'll be plugging in

Re: [ccp4bb] Nanodrop versus Nanophotomter Pearl versus good old Bradford.

2011-06-16 Thread Dima Klenchin
The method is that by Edelhoch, mentioned a couple of times already in this discussion. You recommended "determining extinction coefficients experimentally". How is plugging number of specific residues into a formula constitute experimental determination? It's also described in the paper by

Re: [ccp4bb] Nanodrop versus Nanophotomter Pearl versus good old Bradford.

2011-06-16 Thread aaleshin
I see, by the experimental determination of the extinction coefficient you mean correction for the difference between unfolded (which can be computed accurately) and folded proteins. Am I right? Sorry for making this topic viral... Alex On Jun 16, 2011, at 5:06 PM, Machius, Mischa Christian wro

Re: [ccp4bb] Nanodrop versus Nanophotomter Pearl versus good old Bradford.

2011-06-16 Thread Machius, Mischa Christian
Again, the method described by Gill & von Hippel is based on statistical averages. Mach et al. (Anal. Biochem. 1992, 200, 74) later revised these values. Pace et al. (Protein Science, 1995, 4, 2411) again re-determined these averages, so if anything, the values from Pace should be used. Pace als

Re: [ccp4bb] Nanodrop versus Nanophotomter Pearl versus good old Bradford.

2011-06-16 Thread Pascal Egea
I would like to add something about the NanoDrop versus NanoPearl, I don't think that the path length is fixed on this instrument (the NanoDrop) since if I recall well, the instruments sets the path length as it scans through the droplet, hence the characteristic clicky noise that you hear as the

Re: [ccp4bb] Nanodrop versus Nanophotomter Pearl versus good old Bradford.

2011-06-16 Thread Machius, Mischa Christian
The method is that by Edelhoch, mentioned a couple of times already in this discussion. It's also described in the paper by Pace et al., the same paper that the formula in ProtParam is from (ProtParam does not use the values determined by Gill & von Hippel). Last time I looked into this, the con

Re: [ccp4bb] Nanodrop versus Nanophotomter Pearl versus good old Bradford.

2011-06-16 Thread Scott Pegan
Here is also a very effective method: 1Gill, S. & Hippel, P. v. Calculation of protein extinction coefficients from amino acid sequence data. Analytical Biochemistry 182, 319-326, (1989). On Thu, Jun 16, 2011 at 5:56 PM, Filip Van Petegem < filip.vanpete...@gmail.com> wrote: > A convenient

Re: [ccp4bb] Nanodrop versus Nanophotomter Pearl versus good old Bradford.

2011-06-16 Thread Shaun Lott
Just to add my 2c worth... The department here has a couple of nanodrops as a shared facility, one for DNA/RNA and one for protein. It has been noticeable that over time people has been getting decreased reliability of measurements on the latter machine cf cuvette measurements, presumably due t

Re: [ccp4bb] Nanodrop versus Nanophotomter Pearl versus good old Bradford.

2011-06-16 Thread Filip Van Petegem
A convenient fast way is the earlier mentioned Edelhoch method, as described in this paper which is referenced on the popular Protparam tool: http://onlinelibrary.wiley.com/doi/10.1002/pro.5560041120/pdf Filip On Thu, Jun 16, 2011 at 4:45 PM, aaleshin wrote: > Mischa, > You intrigued me. What

Re: [ccp4bb] Nanodrop versus Nanophotomter Pearl versus good old Bradford.

2011-06-16 Thread aaleshin
Sorry for misprint, I meant evaporating water from a protein solution... On Jun 16, 2011, at 4:45 PM, aaleshin wrote: > Mischa, > You intrigued me. What is the experimental technique for the Extinction > Coefficient measurement (which requires knowledge of protein concentration)? > Let me gue

Re: [ccp4bb] Nanodrop versus Nanophotomter Pearl versus good old Bradford.

2011-06-16 Thread aaleshin
Mischa, You intrigued me. What is the experimental technique for the Extinction Coefficient measurement (which requires knowledge of protein concentration)? Let me guess, Bradford? Protein evaporation and weighing? Alex On Jun 16, 2011, at 4:22 PM, Machius, Mischa Christian wrote: > With re

Re: [ccp4bb] Nanodrop versus Nanophotomter Pearl versus good old Bradford.

2011-06-16 Thread Oganesyan, Vaheh
bulletin board [mailto:CCP4BB@JISCMAIL.AC.UK] On Behalf Of Machius, Mischa Christian Sent: Thursday, June 16, 2011 7:23 PM To: CCP4BB@JISCMAIL.AC.UK Subject: Re: [ccp4bb] Nanodrop versus Nanophotomter Pearl versus good old Bradford. With respect to the Edelhoch method and the ProtParam server, I would stro

Re: [ccp4bb] Nanodrop versus Nanophotomter Pearl versus good old Bradford.

2011-06-16 Thread Machius, Mischa Christian
With respect to the Edelhoch method and the ProtParam server, I would strongly recommend determining extinction coefficients experimentally and not rely on the ProtParam values. The reason is that the underlying extinction coefficients in the formula used by ProtParam and referenced there are st

Re: [ccp4bb] Nanodrop versus Nanophotomter Pearl versus good old Bradford.

2011-06-16 Thread Petr Leiman
Totally support the statements below. We have had several proteins with A280 absorbance of 0.1 or less (at 1 mg/ml). You _have_ to use Bradford in the Nanodrop or whatnot to measure the concentration. Before purchasing the Nanodrop we used a Hellma TrayCell and a "normal" UV/Vis instrument. Sim

Re: [ccp4bb] Nanodrop versus Nanophotomter Pearl versus good old Bradford.

2011-06-16 Thread Jacob Keller
5% accuracy mentioned by Flip is pretty good for biological >>>> samples.  Using 50 ul cuvette in a traditional spectrophotometer will not >>>> give this accuracy because cleanness of the cuvette will be a big issue... >>>> >>>> Alex >>>> &

Re: [ccp4bb] Nanodrop versus Nanophotomter Pearl versus good old Bradford.

2011-06-16 Thread David Briggs
'Fine' for me means comparable to SEC-MALLS measurements and reproducible. I use the E calculated from the sequence using the protparam server at Expasy. David C. Briggs PhD Father, Structural Biologist and Sceptic University of Manchester

Re: [ccp4bb] Nanodrop versus Nanophotomter Pearl versus good old Bradford.

2011-06-16 Thread Filip Van Petegem
reasonable >>>> speed (if you put a drop there then lower the lever and click measure >>>> before >>>> you do anything else) there will be no issues. At very high concentrations >>>> the accuracy and therefore consistency may become lower. Conce

Re: [ccp4bb] Nanodrop versus Nanophotomter Pearl versus good old Bradford.

2011-06-16 Thread Edward A. Berry
Arnon Lavie wrote: ~~~ We have been considering buying a Nanodrop machine (small volume, no dilution needed, fast, easy). However, while testing our samples using a colleague's machine, we have gotten readings up to 100% different to our Bradford assay (all fully purified proteins). For example,

Re: [ccp4bb] Nanodrop versus Nanophotomter Pearl versus good old Bradford.

2011-06-16 Thread Justin Hall
lower. Concentrations between 5 and 10 mg/ml should be fine. The instrument is pricey though. Vaheh From: CCP4 bulletin board [mailto:CCP4BB@JISCMAIL.AC.UK] On Behalf Of Filip Van Petegem Sent: Thursday, June 16, 2011 3:34 PM To: CCP4BB@JISCMAIL.AC.UK Subject: Re: [ccp4bb] Nanodro

Re: [ccp4bb] Nanodrop versus Nanophotomter Pearl versus good old Bradford.

2011-06-16 Thread Chun Luo
ed in publications will help the research community. Cheers, Chun -Original Message- From: CCP4 bulletin board [mailto:CCP4BB@JISCMAIL.AC.UK] On Behalf Of Bjørn Panyella Pedersen Sent: Thursday, June 16, 2011 1:19 PM To: CCP4BB@JISCMAIL.AC.UK Subject: Re: [ccp4bb] Nanodrop versus Nanophotomter

Re: [ccp4bb] Nanodrop versus Nanophotomter Pearl versus good old Bradford.

2011-06-16 Thread Prince, D Bryan
board [mailto:CCP4BB@JISCMAIL.AC.UK] On Behalf Of Arnon Lavie Sent: Thursday, June 16, 2011 3:16 PM To: CCP4BB@JISCMAIL.AC.UK Subject: [ccp4bb] Nanodrop versus Nanophotomter Pearl versus good old Bradford. Dear fellow crystallographers - a question about spectrophotometers for protein concentration

Re: [ccp4bb] Nanodrop versus Nanophotomter Pearl versus good old Bradford.

2011-06-16 Thread Tommi Kajander
e lower. >>> Concentrations between 5 and 10 mg/ml should be fine. The instrument is >>> pricey though. >>> >>> Vaheh >>> >>> >>> >>> From: CCP4 bulletin board [mailto:CCP4BB@JISCMAIL.AC.UK] On Behalf Of Filip

Re: [ccp4bb] Nanodrop versus Nanophotomter Pearl versus good old Bradford.

2011-06-16 Thread David Briggs
I'll give my backing to the Nanodrop as well. I've used it in two different labs, for general yield checking use as well as prior to ITC experiments, and haven't found there to be any issues. That said, I've also used cuvettes, and I find that one the whole, cuvette-derived and nanodrop-derived m

Re: [ccp4bb] Nanodrop versus Nanophotomter Pearl versus good old Bradford.

2011-06-16 Thread aaleshin
> >> >> >> From: CCP4 bulletin board [mailto:CCP4BB@JISCMAIL.AC.UK] On Behalf Of Filip >> Van Petegem >> Sent: Thursday, June 16, 2011 3:34 PM >> To: CCP4BB@JISCMAIL.AC.UK >> Subject: Re: [ccp4bb] Nanodrop versus Nanophotomter Pearl versus go

Re: [ccp4bb] Nanodrop versus Nanophotomter Pearl versus good old Bradford.

2011-06-16 Thread Bjørn Panyella Pedersen
On 2011-06-16 13:06, Filip Van Petegem wrote: Even if evaporation is not an issue, one has to take pipetting errors into account when dealing with small volumes. The relative error on 1-2ul is a lot bigger than on 50ul. True, but the nanodrop works independent of volumes, since it has a fix

Re: [ccp4bb] Nanodrop versus Nanophotomter Pearl versus good old Bradford.

2011-06-16 Thread Quyen Hoang
We also have not experienced any problems with a Nanodrop 2000C. No one in my touched the two boxes of Bradford and BCA kits that we have, because we have been very happy with the Nanodrop. Quyen ___ Quyen Hoang, Ph.D Assistant Professor Department of Biochemistry a

Re: [ccp4bb] Nanodrop versus Nanophotomter Pearl versus good old Bradford.

2011-06-16 Thread Filip Van Petegem
:CCP4BB@JISCMAIL.AC.UK] *On Behalf Of > *Filip > Van Petegem > *Sent:* Thursday, June 16, 2011 3:34 PM > *To:* CCP4BB@JISCMAIL.AC.UK > *Subject:* Re: [ccp4bb] Nanodrop versus Nanophotomter Pearl versus good > old Bradford. > > > Dear Arnon, > > > the Bradford

Re: [ccp4bb] Nanodrop versus Nanophotomter Pearl versus good old Bradford.

2011-06-16 Thread Bosch, Juergen
AC.UK<mailto:CCP4BB@JISCMAIL.AC.UK> Subject: Re: [ccp4bb] Nanodrop versus Nanophotomter Pearl versus good old Bradford. Dear Arnon, the Bradford method is not recommended for accurate measurements. The readings are strongly dependent on the amino acid composition. A much better method is using the

Re: [ccp4bb] Nanodrop versus Nanophotomter Pearl versus good old Bradford.

2011-06-16 Thread Francis E Reyes
Never had problems with evaporation (and this is in the relatively dry climate of Denver, CO, especially in the winter when the relative humidity is in the low 20%). Using the Thermo Scientific Nanodrop 2000c. We use it also as a prerequisite for ITC, which can be very sensitive to proper

Re: [ccp4bb] Nanodrop versus Nanophotomter Pearl versus good old Bradford.

2011-06-16 Thread aaleshin
gt; > From: CCP4 bulletin board [mailto:CCP4BB@JISCMAIL.AC.UK] On Behalf Of Filip > Van Petegem > Sent: Thursday, June 16, 2011 3:34 PM > To: CCP4BB@JISCMAIL.AC.UK > Subject: Re: [ccp4bb] Nanodrop versus Nanophotomter Pearl versus good old > Bradford. > > Dear Arnon, > > t

Re: [ccp4bb] Nanodrop versus Nanophotomter Pearl versus good old Bradford.

2011-06-16 Thread Oganesyan, Vaheh
tegem Sent: Thursday, June 16, 2011 3:34 PM To: CCP4BB@JISCMAIL.AC.UK Subject: Re: [ccp4bb] Nanodrop versus Nanophotomter Pearl versus good old Bradford. Dear Arnon, the Bradford method is not recommended for accurate measurements. The readings are strongly dependent on the amino acid compositio

Re: [ccp4bb] Nanodrop versus Nanophotomter Pearl versus good old Bradford.

2011-06-16 Thread Filip Van Petegem
Dear Arnon, the Bradford method is not recommended for accurate measurements. The readings are strongly dependent on the amino acid composition. A much better method is using the absorption at 280nm under denaturing conditions (6M Guanidine), and using calculated extinction coefficients based on

[ccp4bb] Nanodrop versus Nanophotomter Pearl versus good old Bradford.

2011-06-16 Thread Arnon Lavie
Dear fellow crystallographers - a question about spectrophotometers for protein concentration determination. We are so last millennium - using Bradford reagent/ 1 ml cuvette for protein conc. determination. We have been considering buying a Nanodrop machine (small volume, no dilution needed,