For accurate absorbance measurements we have been using a 10 ul
"tunnel" cuvette
for some time. It is still 1 cm pathlength, fits into standard specs,
one can manage even with 8 ul of a sample fed into the channel,
which can be rescued ~ 80%, no evaporation issues, the cuvette has to
be kept clean!

Jan D.


On Fri, Jun 17, 2011 at 3:36 AM, Richard Edward Gillilan
<r...@cornell.edu> wrote:
>
> Hi Chelsy, yes we had a lot of trouble with the nanoview during that run.
>  Even after going through the calibration procedure with the special fluid
> provided, we still had inconsistent results even on standards.  Finally, I
> carefully cleaned the return light path of the instrument (a separate set of
> holes towards the back of the glass plates). This seemed to fix the problem.
> The manual warns that users should always wipe from back to front when
> cleaning out sample. The instrument also seems sensitive to how well the
> drop is formed on the glass plate. Once the hydrophobic coat is worn or
> dirty, the drops spread and give poor results, so maybe detergent was also
> an issue.  So these machines can be fussy we are learning.
> Richard Gillilan
> MacCHESS
>
> On Jun 16, 2011, at 9:03 PM, Chelsy Prince wrote:
>
> Hi everyone,
>
> I am working with a membrane protein and normally measure my protein
> concentration by diluting and then reading OD280 (1cm pathlength). I have
> found this to be very consistent, if a bit time consuming.
>
> At the syncatron, I had to use a Nanodrop for the first time to check some
> SAXS dilutions I made on site and it was all over the map. We tested each
> sample 8-10 times and saw large variations (i.e. 2.5-4.1mg/mL) on a single
> sample. I used the Nanodrop to measure OD280 and calibrated it with the same
> extinction coefficient I normally use.
>
> Since I have dodecyl-maltoside in all of my solutions (which is the cause of
> most of my problems), I was wondering if it was also the culprit here.
> The detergent is probably lowering the surface tension of the drops.
>
> It is possible that I was just in-experienced, but I had multiple people
> from the facility helping me and we all had the same issue with my samples.
> In addition, the Nanodrop would regularly complain about inconsistency
> between the two readings and wouldn’t accept readings/blanks at all.
>
> Is it me or the membrane protein? And is there anything that I could do to
> improve the readings the next time that I need to use it?
>