We have used Nanodrop for several years and found the readings are always
accurate. The highest concentration we have measured is around 30 mg/ml. The
differences between diluted and concentrated samples are within dilution
error. Nanodrop spectra at low concentration are noisier. We actually prefer
to have protein concentrations over 0.5 mg/ml.

 

The key to get good Nanodrop reading is to have a nice liquid column. Take a
look whiling waiting for the results. High concentration samples with
detergent or glycerol may produce bubbles (giving an error message
sometimes) or improperly formed liquid column (giving incorrect numbers).
Use more samples (up to 5 ul) and gently push down the head will help the
liquid column formation. 1 ul is not enough for most protein samples. If the
reading is not good, wipe clean the surface and load a new sample. Repeated
reading of same sample gave huge variation. The liquid column probably has
difficulty to form properly after the first reading.

 

We provide protein purification contract services and send proteins to many
users around the globe. The proteins have been used for crystallography,
ITC, DSC, Biacore, and activity assays. So far we haven’t heard any complain
about our concentration measurement using Nanodrop.

 

We also use Bradford to measure concentration if required by our customers
or when concentration can’t be measured by A280. The error range for
Bradford is much wider and easily off by a factor of 2 (by repeating or
comparing to Nanodrop). There are different brands of Bradford reagent. Each
gives a different number. We once had a big discrepancy with a customer. It
turns out the home-made BSA standard lot was off. After adjusting the BSA
concentration with Nanodrop and using the same Bradford reagent and same
curve fitting method, the concentrations agreed. The lesson leaned is to
have good fresh standard solution.

 

The color of Bradford changes over time, it’s better to use a plate reader
than single sample spectrometer. Estimation by eye is just as good as a
spectrometer:-)

 

--Chun

 

  _____  

From: CCP4 bulletin board [mailto:[EMAIL PROTECTED] On Behalf Of
Juergen Bosch
Sent: Saturday, December 06, 2008 12:29 AM
To: CCP4BB@JISCMAIL.AC.UK
Subject: Re: [ccp4bb] AW: [ccp4bb] suggestions for UV spectrometer

 

Hi,

 

really strange, I always dilute my protein when taking an absorption
spectra. I try to adjust my expected concentration to a readout of ~ 0.2-0.3
OD280. And the Bradford is just as well as 'picking house numbers',
depending on your protein, you can underestimate your protein concentration
by a factor of 2.

For my current approach I use a 100 µl volume cuvette and 2 µl of my
concentrated protein (most of the times). The Structural Genomics lab
(MSGPP) has a Nanodrop and I was also always happy with those results (but
again I dilute my protein sample before taking a measurement) 

 

Jürgen

 

On 5 Dec 2008, at 16:00, wangsa tirta ismaya wrote:





Dear all,

Thanks a lot for raising the issue with the not reproducibility of protein
measurement with Nanodrop. We use the instrument as a workhorse in the lab.
Indeed, recently I observed that the protein concentration suggested by
Nanodrop is sometimes differ to the usual colorimetric measurement (Bradford
method, measured with our Pharmacia's Ultrospec 2000 spectrometer). Since
the cell in Nanodrop is very small, could it be due to the homogeneity of
the sample in the cell? Also what I have observed, we have to be sure that
the cell is cleaned properly before use. Another thing is, at high protein
concentration I obtained noisy absorption curve at the peak (like a
seismograf ....) thus the protein concentration measured is doubtful, I have
to dilute the sample to have good curve (thank God it requires only 2
mikroliter for a measurement). Well, I think nanodrop is a good, fast, and
powerful instrument, however it would be better if we established a
reference of our daily practice to a normal spectrometer measurement.

cheers,


Wangsa

2008/12/5 Martin Hallberg <[EMAIL PROTECTED]>

Which brings us back to the Hellma "TrayCell" solution where you can, from
the same spectrometer, have both the cuvette option and the quickness of the
NanoDrop/NanoVue system.

Anyone that can comment on the performance of the TrayCell from Hellma?

Cheers,

Martin



On Dec 5, 2008, at 9:06 AM, Gregor Witte wrote:

Agree!
I think for crystallographic use the nanodrop is perfectly okay to see if
the protein is 5mg/ml or 30mg/ml. But in fact I also do not trust our
instrument if it comes to more important issues like preparing solutions for
titrations or assays. And due to the small pathlength I do not trust
absorptions of small concentrated samples at all. I always prefer a "real"
2-beam spectrophotometer (monochromators) with a quarz-cuvette and a nice
pathlength. Of course, you cannot reliably measure solutions exceeding Abs 1
or maybe 1.5 OD in a spectrophotometer with 1cm pathlength.

There's also one quite strange thing about the nanodrop – they sell the
"calibration check solution" (which is some kind of yellow chromate-solution
with known absorption), then you check your nanodrop with it and maybe find
out that it's off to some certain extend: But then you're stuck(!), because
you cannot calibrate it on your own. I guess it would be quite easy to
integrate a calibration-option into the software, but at the moment the
instrument tells you "calibration failure" and you have to call the service
guys who then carry it home and calibrate it by turning of the two small
screws at the top of it and then glue them with locktite.
Anyway, at least for our mid to high concentrated samples the nanodrop is
not showing large fluctuations so we are happy with it. But everyone using a
nanodrop should check it from time to time – as I found out that ours was
off more than 20% at one day - which raised some trouble of course…

Cheers,

Gregor


Von: CCP4 bulletin board [mailto:[EMAIL PROTECTED] Im Auftrag von Filip
Van Petegem
Gesendet: Donnerstag, 4. Dezember 2008 22:20
An: CCP4BB@JISCMAIL.AC.UK
Betreff: Re: [ccp4bb] suggestions for UV spectrometer

I want to add I absotely hate the nanodrop.  We've had a demo for it, and
found the readouts to be very unreliable.  Fluctuations of 20% and more.
Just leaving the same drop in and measuring the sample multiple times gives
different values (going in both directions, so not only due to
evaportations). Sure, it's easy and fast, and maybe good to have a rough
idea about your protein concentration, but I would never want to use it for
exact measurements such as needed for e.g. a CD or an ITC instrument. I've
heard other labs in our department have similar issues.  We've also had a
demo for the Nanovue from GE Healthcare:  same issues - very large
fluctuations from one sample to another.  I suppose this is simply an
inherent problem with small volumes...

Cheers

Filip Van Petegm


On Thu, Dec 4, 2008 at 12:48 PM, Patrick Loll <[EMAIL PROTECTED]> wrote:
At the risk of dragging this discussion even further afield from
crystallography:

How can you get realistic numbers for concentrated solutions using the
Nanodrop?  I understand that the instrument reduces absorbance by using a
very short path length. However, I thought that in order for the
Beer-Lambert formalism to be applicable, the solution needs to be
sufficiently dilute so that the chance of molecules "shadowing" one another
is negligible. Isn't this condition violated for concentrated solutions
(even with short path lengths)?

Pat

On 4 Dec 2008, at 1:27 PM, Michael Giffin wrote:


We also like the Nanodrop...
----------------------------------------------------------------------------
-----------
Patrick J. Loll, Ph. D.
Professor of Biochemistry & Molecular Biology
Director, Biochemistry Graduate Program
Drexel University College of Medicine
Room 10-102 New College Building
245 N. 15th St., Mailstop 497
Philadelphia, PA  19102-1192  USA

(215) 762-7706
[EMAIL PROTECTED]




-- 
Filip Van Petegem, PhD
Assistant Professor
The University of British Columbia
Dept. of Biochemistry and Molecular Biology
2350 Health Sciences Mall - Rm 2.356
Vancouver, V6T 1Z3

phone: +1 604 827 4267
email: [EMAIL PROTECTED]
http://crg.ubc.ca/VanPetegem/

 

.
B. Martin Hallberg, PhD
Assistant professor
Department of Cell and Molecular Biology
Medical Nobel Institute
Karolinska Institutet
Von Eulersv. 1
SE-171 77 Stockholm
Sweden
Fax: +46-8-339380




-- 
Wangsa Tirta Ismaya
=====================================
Josef-Israelstraat 66, 9718 GN Groningen
The Netherlands

 

-

Jürgen Bosch
University of Washington
Dept. of Biochemistry, K-426
1705 NE Pacific Street
Seattle, WA 98195
Box 357742
Phone:     +1-206-616-4510
FAX:        +1-206-685-7002
Web:     http://faculty.washington.edu/jbosch

 

Reply via email to