Hi Daniel

The peak shape variation comes from exponential absorption of the incident
and directed X-rays due to their interaction below the specimen's surface.
In the case of a finite thickness sample, this exponential gets truncated.
I'm pretty sure I talk about this in Rowles & Buckley.

Intensity is the main thing. Peak shape is going to be dominated by
crystallite size/strain, but in its absence, youll see cut off in the peak
trails due to finite thickness.

Matthew



On Thu, 7 Nov 2024, 16:39 Daniel Chateigner, <daniel.chateig...@ensicaen.fr>
wrote:

> Dear All,
>
> Just to be sure on two of the Q&As:
> Le 07/11/2024 à 05:03, Matthew Rowles a écrit :
>
> Great questions!
>
>
> > Is it worthy to use the parameter known as "Sample Thickness",
> associated with the Absorption Correction in TOPAS?
>
> Generally, no. You only need to worry about it if the penetration depth of
> your incident beam is larger than the thickness (t) of the specimen. If
> that is the case, then "Sample Thickness" and "Scale Intensity" is
> necessary to correct for the change in peak shape due to the finite depth,
> and the change in diffracting volume.
>
> - Seems from your answer Matthew, that diffracting volume changes (Vd)
> through the scan in BB geometry gives rise to some peak shape variations (I
> assume here that "shape" includes "width"). I agree that the diffracting
> volume changes with theta, and with the sample thickness, but these two
> effects should not contribute the same on the peak shape:
>
> -- For a given sample thickness, the diffracting volume varies with Theta,
> but to me the main contribution to the peak shape is due to the irradiated
> surface (beam track on the sample surface). Since deeper dVd will be more
> absorbed than the ones closer to the surface. You can picture this
> measuring a 2Theta scan at fixed incident angle, up to 2Theta=90°, the peak
> width increases notably with 2Theta because of this.
>
> -- For the same theta on the diffraction diagram, varying the sample
> thickness will probably not modify this much the shape of the peaks, but
> intensities. For the same reasons on dVd. For instance, if measuring 2
> samples with a 100 nm and 200 nm thickness respectively, the peak shape
> variation will probably not be visible using lab instruments, while the
> intensities will differ significantly.
>
> - I do not like much the term "penetration depth" (p), since the depth
> reached by x-rays is always infinite. For instance, if as usual p is
> defined as 1/mu (mu = absorption coefficient of the sample, the one we
> never fully know exactly), using a Theta-2Theta scan in BB on a t=infinity
> sample, the measured diffracted intensity is larger than the one coming
> from p only. And, on a thin specimen, both absorption and Vd are varying
> during the scans, each of them bringing a contribution to the intensity. On
> the contrary, for t=infinite specimen, they compensate each-other and their
> contribution is the same on the diffracted intensity throughout the diagram
> (but the intensity does not come from the same irradiated volume). I was
> then wondering if Topas was simply corrected from volume variations or also
> including the absorption variations in the case of "thin enough" samples ?
>
>
>
>
> > Is the negative intensity in the difference curve...
>
> It is showing that there is some mismatch between your model and the data.
> The intensities could be off because the peaks positions are wrong, the
> peak-shape is wrong, the atomic displacement parameters are wrong (don't
> refine these from normal QPA data), the LP factor correction is wrong, the
> site occupancies are wrong, the atomic scattering factors are wrong. Could
> also be that you have forgotten to refine a parameter that you should be.
>
> What about preferred orientations ? Anisotropic crystallite shapes and
> microstrains ? Improperly defined film thickness can give rise to
> variations of the ADPs also.
>
>
> My 2€cts !
>
> Daniel
>
>
>
>
>
>
>
>
>
>
>
> On Wed, 6 Nov 2024 at 19:12, Łukasz Kruszewski <lkruszew...@twarda.pan.pl>
> wrote:
>
>> Dear Rietvelders,
>>
>> after some years of being a happy user of TOPAS & Rietveld method (after
>> working-out a good approach, thanks to a great help from users of this
>> mailing list for which I am really thanful, and having it checked by
>> attendance in Reynolds Cup) I was recently confronted with few questions
>> I couldn't answer. I thus have a kind request of a help here:
>>
>> - does the Absorption correction also influences microabsorption? This
>> question arose due to somewhat lowered expected wt.% of pyrite (19
>> instead of 25), FeS2, even though all the used parameters (Absorption,
>> CrySize, Strain - for all the sample components) were checked and proven
>> to give value > error, i.e., being physically meaningful for the
>> refinement model. I was wondering how this supposed non-precision for
>> pyrite could be addressed...
>>
>> - Is it worthy to use the parameter known as "Sample Thickness",
>> associated with the Absorption Correction in TOPAS? I was taught in a
>> 5-day course of TOPAS that it's not a good idea, especially with the
>> Bragg-Brentano geometry, but I'm not sure...
>>
>> - Is there any difference, to be expected, in the calculated unit cell
>> parameters, between approximating a phase using (1) a chemically pure
>> STRUCTURE model (i.e., taken directly from a database / CIF file), and
>> (2) a STRUCTURE with Occupancy information updated with the known
>> content of impurities? I always thought that it shouldn't matter, as the
>> software simply "iteratively" compares the sample' reflections with the
>> structural information contained in STRUCTURE models (?). I have
>> compared these two situations, and option "2" actually gave me worse
>> fit. I thus wondered: does it really matter that much which model
>> (approximation) of the structure - chemically pure or with added
>> impurity info - is used? For most substances, the reflection positions
>> vary even for different standards found in databases like PDF or COD...
>>
>> - Is the negative intensity in the difference curve only informing us of
>> underestimation of the content of a particular phase, suggesting to
>> replace the Structure model used with another one?
>>
>> We have been working using few computers with, supposedly the very same
>> version of the software. However, even though we've been using exactly
>> the same Structure models, results were varying depending on the
>> computer stand used. I was wondering if this could have to do with some
>> errors in the JAVA script, or maybe a non-identical installation
>> conditions, or something about the Windows system?
>>
>> I would be thankful for any tips.
>>
>> Best regards,
>>
>> --
>> Łukasz Kruszewski, Ph.D., Associate Professor
>> Polish Academy of Sciences
>> Institute of Geological Sciences
>> Twarda 51/55
>> <https://www.google.com/maps/search/Twarda+51%2F55?entry=gmail&source=g>
>> str.
>> 00-818 Warsaw
>> Poland
>> ++++++++++++++++++++++++++++++++++++++++++++++++++++++++++++++
>> Please do NOT attach files to the whole list
>> <alan.he...@neutronoptics.com> <alan.he...@neutronoptics.com>
>> Send commands to <lists...@ill.fr> eg: HELP as the subject with no body
>> text
>> The Rietveld_L list archive is on
>> http://www.mail-archive.com/rietveld_l@ill.fr/
>> ++++++++++++++++++++++++++++++++++++++++++++++++++++++++++++++
>>
>>
> ++++++++++++++++++++++++++++++++++++++++++++++++++++++++++++++
> Please do NOT attach files to the whole list <alan.he...@neutronoptics.com> 
> <alan.he...@neutronoptics.com>
> Send commands to <lists...@ill.fr> <lists...@ill.fr> eg: HELP as the subject 
> with no body text
> The Rietveld_L list archive is on 
> http://www.mail-archive.com/rietveld_l@ill.fr/
> ++++++++++++++++++++++++++++++++++++++++++++++++++++++++++++++
>
>
> --
>
> ----------------------------------------------------------------------
> Daniel Chateigner
> Professeur Normandie Universitechateigner.ensicaen.fr/
> ----------------------------------------------------------------------
> Editor: "Combined Analysis", Wiley-ISTE: 
> http://eu.wiley.com/WileyCDA/WileyTitle/productCd-1848211988.html
> Workshops on Combined Analysis: chateigner.ensicaen.fr/formation/
> ----------------------------------------------------------------------
> Address:
> CRISMAT-ENSICAEN and IUT-Caen,
> Universite de Caen Normandie, campus 2
> 6, Bd. M. Juin 14050 Caen, France
> tel: 33 (0)2 31 45 26 11daniel.chateig...@ensicaen.fr
> ----------------------------------------------------------------------
> Open Databases:
> Material Properties Open Database (MPOD): mpod.cimav.edu.mx
> Crystallography Open Database     (COD): www.crystallography.net/cod/
>                   Theoretical     (TCOD): www.crystallography.net/tcod/
>                   Predicted       (PCOD): www.crystallography.net/pcod/
> Raman Open Database (ROD): solsa.crystallography.net/rod/
> Full-Profile <http://solsa.crystallography.net/rod/Full-Profile> Search-Match 
> (FPSM): cod.iutcaen.unicaen.fr/
> ----------------------------------------------------------------------
>
> ++++++++++++++++++++++++++++++++++++++++++++++++++++++++++++++
> Please do NOT attach files to the whole list <alan.he...@neutronoptics.com
> >
> Send commands to <lists...@ill.fr> eg: HELP as the subject with no body
> text
> The Rietveld_L list archive is on
> http://www.mail-archive.com/rietveld_l@ill.fr/
> ++++++++++++++++++++++++++++++++++++++++++++++++++++++++++++++
>
>
++++++++++++++++++++++++++++++++++++++++++++++++++++++++++++++
Please do NOT attach files to the whole list <alan.he...@neutronoptics.com>
Send commands to <lists...@ill.fr> eg: HELP as the subject with no body text
The Rietveld_L list archive is on http://www.mail-archive.com/rietveld_l@ill.fr/
++++++++++++++++++++++++++++++++++++++++++++++++++++++++++++++

Reply via email to