Dear Reinhard,

I wouldn't say "solely from the detector side". The biggest improvement is 
synchrotron radiation (+2D detectors🤓).

Best greetings from Lac Leman

Radovan Cerny

Envoyé à partir de Outlook pour Android<https://aka.ms/AAb9ysg>
________________________________
From: rietveld_l-requ...@ill.fr <rietveld_l-requ...@ill.fr> on behalf of 
Reinhard Kleeberg <kleeb...@mineral.tu-freiberg.de>
Sent: Tuesday, September 5, 2023 1:49:57 PM
To: rietveld_l@ill.fr <rietveld_l@ill.fr>
Subject: [EXTERNAL] Re: [EXT] [External] Re: Step-like basline

Dear Luca,
I completely agree and will be very happy to have a 1D detector with
sufficient energy resolution to resolve a "pure" Cu Kalpha1/2 doublet
from W Lalpha1. This would make routine work much easier. IMHO, the
biggest improvements in XRPD instrumentation within the last decades
came solely from the detector side, should be continued.

Regarding search-match by Rietveld: Your FSPM method is a very
inspiring idea. Nicola Doebelin has incorporated a simplified approach
in PROFEX, too:
https://www.youtube.com/watch?v=Oqj71TiifeI

I'm routinely applying a run of ~500 structures/minerals what we are
commonly using (+- the former QPA database of Seifert AUTOQUAN or from
the BGMN webpage) for a phase screening of unknown samples, needs ~ 5
min without interaction. Than running QPA refinement with the best
matching structures, clicking on remaining peaks in the difference
plot for getting a proposal of best matching main line positions in
the database, and adding these structure(s) to the refinement, running
again... This is a typical workflow in our lab, no need for peak
search, background treatment, thinking about artifact peaks...

Best regards

Reinhard

Zitat von Luca Lutterotti <luca.luttero...@unitn.it>:

> Dear Reinhard and Rietvelders,
>
> It is always a compromise. Ideally we would like the fastest
> instrument with a lot of intensity (so, 1D or 2D detectors) but no
> lines outside the Kalpha or even just the Kapha1 like the one of
> James Cline. But this means a monochromator on the incident beam and
> you get fluorescence background. Now I like your solution of the Si
> Drift detector, so you just cut the fluorescence and you don’t need
> the monochromator in the first place.
> So we are working and we are testing some lab prototypes of a 1D Si
> Drift like detector to get the best of both world. It is only a lot
> of electronic, but one day we will have wonderful instrument with
> only the lines we need and no fluorescence for the background.
> Actually we use fluorescence for the chemical analysis (quantitative).
>
> About the search-match problem. Have a look on our FPSM method (
> http://fpsm.radiographema.com/) where we don’t care about extra
> lines as it is a Rietveld search-match. We don’t need to identify or
> search peaks. We don’t use peak positions, we just fit with the
> Rietveld. It is slower indeed, but every day is becoming faster and
> in a few years it will run quickly on our cellphones (I have a
> prototype running there).
>
> Best regards,
>
> Luca
>
>  <http://www.unitn.it/>
>
> Luca Lutterotti
> Dipartimento di Ingegneria Industriale
> Università di Trento
> via Sommarive, 9 - 38123 Trento (Italy)
> tel. +39 0461 2824-14 (Office), -34 (X-Ray lab)
>
>
> 
>
> Maud: http://maud.radiographema.com <http://maud.radiographema/>
>
>
>> Begin forwarded message:
>>
>> From: Reinhard Kleeberg <kleeb...@mineral.tu-freiberg.de>
>> Subject: Re: [EXT] [External] Re: Step-like basline
>> Date: 5 September 2023 at 08:56:15 CEST
>> To: rietveld_l@ill.fr
>> Reply-To: Reinhard Kleeberg <kleeb...@mineral.tu-freiberg.de>
>>
>> Dear Luca,
>> I completely agree with your opinion, and want to add that even for
>> "trivial" tasks in XRPD like phase analysis and standard Rietveld
>> refinements the satellites/spectral impurities do cause significant
>> trouble:
>>
>> - The K beta and W L satellites of strong peaks of major phases
>> like quartz, carbonates or cubic structures in geomaterials are
>> typically not automatically recognized in the standard peak search
>> procedures and therefore misinterpreted to be K alpha peaks and
>> added to the peak list. I can't tell you how often I was asked from
>> colleagues for explanation of such "unidentified peaks", and how
>> much time people have spent for searching for explanation of such
>> artefact lines.
>> - In Rietveld analysis our software must generate the peaks at
>> least for the measured angular range. Depending on the method how
>> the software is doing this, we run into problems with the
>> satellites. If the software generates the reflections from the
>> positions of K alpha peaks from the start lattice parameters and
>> the (extended) upper measured angle, K beta satellites of K alpha
>> peaks outside this angular range will maybe not generated and
>> remain unfitted in the pattern. If the software generates the peaks
>> depending on the shortest wavelength in the wavelength profile
>> (more intelligent approach), the software must generate much more
>> reflections (with maxima outside the measured angular range) and
>> all these peaks must be calculated over an extremely broad angular
>> range. In the case of low symmetry structures with big cells or
>> disordered structures described by partial structure factors and
>> the resulting extreme high numbers of peaks, the "extension effect"
>> will cause significantly prolonged time for calculations, without
>> any positive effect.
>>
>> That's why I prefer to use instrumentation with better
>> monochromatic radiation (monochromators, high energy resolution
>> detectors) even in the daily business of phase analysis. We do use
>> our 1D detector Mythen2 with Fe filter (Co radiation) only for
>> "quick and dirty" measurements.
>>
>> Best regards
>>
>> Reinhard
>>
>>
>> Zitat von Luca Lutterotti <luca.luttero...@unitn.it>:
>>
>>> Dear Habib,
>>>
>>> Reinhard is right, and what he explained is exactly what you
>>> observed. Now I would add that I may not define your Bruker clean
>>> and optimised, because for this kind of samples, wafers and
>>> extremely textured thin films, it would be better to have a
>>> monochromator in the incident beam and not a Ni filtered (I would
>>> more say it is a requirement). It was already showed many times in
>>> the past as a monochromator in the incident beam is a necessity
>>> for this materials to avoid all the "physical artifacts" created
>>> by the strong intensity and not clean Kalpha radiation.
>>>
>>> Best regards,
>>>
>>> Luca
>>>
>>> <http://www.unitn.it/>
>>>
>>> Luca Lutterotti
>>> Dipartimento di Ingegneria Industriale
>>> Università di Trento
>>> via Sommarive, 9 - 38123 Trento (Italy)
>>> tel. +39 0461 2824-14 (Office), -34 (X-Ray lab)
>>>
>>>
>>> 
>>>
>>> Maud: http://maud.radiographema.com <http://maud.radiographema/>
>>>
>>>
>>>> On 4 Sep 2023, at 12:20, Habib Boughzala
>>>> <habib.boughz...@ipein.rnu.tn> wrote:
>>>>
>>>> Many thanks Reinhard,
>>>>
>>>> That's exactly what I wanted to say by  "I can assure that our
>>>> Bruker D8 is clean and optimized!"
>>>> Otherwise, in some other cases of well conditioned thin film no
>>>> similar phenomenon is observed!
>>>> So, yes, it's obviously possible that your point of view is right.
>>>>
>>>> Regards
>>>> Habib
>>>>
>>>>
>>>> ------ Message d'origine ------
>>>> De "Reinhard Kleeberg" <kleeb...@mineral.tu-freiberg.de
>>>> <mailto:kleeb...@mineral.tu-freiberg.de>>
>>>> À "Habib Boughzala" <boughz...@yahoo.com <mailto:boughz...@yahoo.com>>
>>>> Cc rietveld_l@ill.fr <mailto:rietveld_l@ill.fr>
>>>> Date 04/09/2023 11:07:50
>>>> Objet Re: Re[2]: [EXT] Re: [External] Re: Step-like basline
>>>>
>>>>> Dear Habib,
>>>>> the phenomenom "satellites or edges" originates from the
>>>>> diffraction process. The critical parameters are:
>>>>> - spectral pureness of the primary beam (primary beam
>>>>> monochromator, tube spectral contamination like W...)
>>>>> - the use of K beta absorbtion filter (and its thickness)
>>>>> - the energy resolution ("window") of the detector system.
>>>>> Even a D8 system may be equipped with different types of
>>>>> detectors, slits and energy limits can be set differently for an
>>>>> identical configuration, and quite often satellite peaks may
>>>>> appear later in the time of use (aging of the tube produces more
>>>>> W L, Fe filters may corrode and get perforated...). So it is
>>>>> strictly recommended to check the instrument peridically, by
>>>>> measuring a full pattern of a profile standard (LaB6 or Si or
>>>>> similar).
>>>>> Greetings
>>>>> Reinhard
>>>>>
>>>>> Zitat von Habib Boughzala <boughz...@yahoo.com
>>>>> <mailto:boughz...@yahoo.com>>:
>>>>>
>>>>>> Dear all,
>>>>>> I would like to send you my witness related to this kind of observation.
>>>>>> I can assure that our Bruker D8 is clean and optimized!
>>>>>>
>>>>>> In many cases of well conditioned thin film (spin coating or
>>>>>> controlled diffusion) material this kind of phenomenon is
>>>>>> visible around the highest reflection, especially when the
>>>>>> preferred orientation is drastically present.
>>>>>>
>>>>>> So, in my opinion, Reinhard and Alan are right, and what is
>>>>>> observed is just like reflections broadening, asymmetry,
>>>>>> shifting ...etc ... and can be related to the material behavior.
>>>>>> Now, what is the physical (crystallographic!) property
>>>>>> responsible of this phenomenon? let's open the floor for a
>>>>>> large discussion.
>>>>>>
>>>>>>
>>>>>> Habib
>>>>>>
>>>>>>
>>>>>> ------ Message d'origine ------
>>>>>> De "Alan W Hewat" <alan.he...@neutronoptics.com
>>>>>> <mailto:alan.he...@neutronoptics.com>>
>>>>>> À "Reinhard Kleeberg" <kleeb...@mineral.tu-freiberg.de
>>>>>> <mailto:kleeb...@mineral.tu-freiberg.de>>
>>>>>> Cc rietveld_l@ill.fr <mailto:rietveld_l@ill.fr>
>>>>>> Date 04/09/2023 09:29:08
>>>>>> Objet Re: [EXT] Re: [External] Re: Step-like basline
>>>>>>
>>>>>>> Reinhard is right that it is best to improve the instrument to
>>>>>>> produce cleaner data. I'm concerned about the advice to model
>>>>>>> all kinds of features whose origin is not fully understood,
>>>>>>> simply to obtain a better fit. Shay has told us nothing about
>>>>>>> his instrument or his conditions of data collection. He asks
>>>>>>> "Is it a sample preparation problem", to which the obvious
>>>>>>> reply is "Do you see this with other samples or different
>>>>>>> materials" ? Only he can answer that. If the answer is yes, he
>>>>>>> might try modifying his instrument (remove filters etc) to see
>>>>>>> what effect that has on the pattern from a simple well
>>>>>>> characterised material. Again only he can do that. Data
>>>>>>> collection is an experimental science, and data refinement
>>>>>>> should not be reduced to a "black box" computer program where
>>>>>>> extra parameters can be added to reduce the R-factor.
>>>>>>> Alan.
>>>>>>>
>>>>>>> On Mon, 4 Sept 2023 at 08:18, Reinhard Kleeberg
>>>>>>> <kleeb...@mineral.tu-freiberg.de
>>>>>>> <mailto:kleeb...@mineral.tu-freiberg.de>> wrote:
>>>>>>>> Can be modeled in the BGMN peak profile model as well, by modifying
>>>>>>>> the *.lam file by a series of additional Lorentzians on the 1/lambda
>>>>>>>> scale, see figure.
>>>>>>>> The same can be done for other spectral impurities, e.g. W L
>>>>>>>> satellites. Also "electronic effects" on the wavelength distribution
>>>>>>>> profile like the "edges" from the ROI settings of Si drift detectors
>>>>>>>> can be modelled in such a convolution based approach.
>>>>>>>> However, better to have a pure/simple wavelength distribution (clear
>>>>>>>> alpha1/2 doublet) by a monochromator or high energy resolution
>>>>>>>> detector, as any satellites make trouble in trace phase analysis and
>>>>>>>> do cause prolonged calculation time in complicated Rietveld
>>>>>>>> refinements.
>>>>>>>>
>>>>>>>> Reinhard
>>>>>>>>
>>>>>>>>
>>>>>>>>
>>>>>>>> Zitat von Matthew Rowles <rowle...@gmail.com
>>>>>>>> <mailto:rowle...@gmail.com>>:
>>>>>>>>
>>>>>>>>> Topas can model them quite well. The functionality was introduced in
>>>>>>>>> version 5.
>>>>>>>>>
>>>>>>>>> On Mon, 4 Sep 2023, 00:54 Kurt Leinenweber, <ku...@asu.edu
>>>>>>>>> <mailto:ku...@asu.edu>> wrote:
>>>>>>>>>
>>>>>>>>>> Hi, Are these things modeled in Rietveld programs, by
>>>>>>>>>> chance? It seems
>>>>>>>>>> like a lot of baggage to put in a refinement but if it
>>>>>>>>>> makes the results
>>>>>>>>>> better…
>>>>>>>>>>
>>>>>>>>>>
>>>>>>>>>>
>>>>>>>>>> - Kurt
>>>>>>>>>>
>>>>>>>>>>
>>>>>>>>>>
>>>>>>>>>> *From:* rietveld_l-requ...@ill.fr
>>>>>>>>>> <mailto:rietveld_l-requ...@ill.fr>
>>>>>>>>>> <rietveld_l-requ...@ill.fr
>>>>>>>>>> <mailto:rietveld_l-requ...@ill.fr>> *On Behalf
>>>>>>>>>> Of *Thomas Gegan
>>>>>>>>>> *Sent:* Sunday, September 3, 2023 9:16 AM
>>>>>>>>>> *To:* Bish, David L <b...@indiana.edu
>>>>>>>>>> <mailto:b...@indiana.edu>>; Shay Tirosh <stiro...@gmail.com
>>>>>>>>>> <mailto:stiro...@gmail.com>>;
>>>>>>>>>> Fernando Igoa <fer.igoa.1...@gmail.com
>>>>>>>>>> <mailto:fer.igoa.1...@gmail.com>>
>>>>>>>>>> *Cc:* Rietveld List (rietveld_l@ill.fr
>>>>>>>>>> <mailto:rietveld_l@ill.fr>) <rietveld_l@ill.fr
>>>>>>>>>> <mailto:rietveld_l@ill.fr>>
>>>>>>>>>> *Subject:* RE: [EXT] Re: [External] Re: Step-like basline
>>>>>>>>>>
>>>>>>>>>>
>>>>>>>>>>
>>>>>>>>>> I agree with a Ni absorption edge, possibly with a Kβ peak
>>>>>>>>>> around 38° 2θ.
>>>>>>>>>>
>>>>>>>>>>
>>>>>>>>>>
>>>>>>>>>> *Tom Gegan*
>>>>>>>>>> Chemist III
>>>>>>>>>>
>>>>>>>>>>
>>>>>>>>>>
>>>>>>>>>> Phone: +1 732 205-5111, Email: tom.ge...@basf.com
>>>>>>>>>> <mailto:tom.ge...@basf.com>
>>>>>>>>>> Postal Address: BASF Corporation, , 25 Middlesex Essex
>>>>>>>>>> Turnpike, 08830
>>>>>>>>>> Iselin, United States
>>>>>>>>>>
>>>>>>>>>>
>>>>>>>>>>
>>>>>>>>>> *From:* rietveld_l-requ...@ill.fr
>>>>>>>>>> <mailto:rietveld_l-requ...@ill.fr>
>>>>>>>>>> <rietveld_l-requ...@ill.fr
>>>>>>>>>> <mailto:rietveld_l-requ...@ill.fr>> *On Behalf
>>>>>>>>>> Of *Bish, David L
>>>>>>>>>> *Sent:* Sunday, September 3, 2023 7:08 AM
>>>>>>>>>> *To:* Shay Tirosh <stiro...@gmail.com
>>>>>>>>>> <mailto:stiro...@gmail.com>>; Fernando Igoa <
>>>>>>>>>> fer.igoa.1...@gmail.com <mailto:fer.igoa.1...@gmail.com>>
>>>>>>>>>> *Cc:* Rietveld List (rietveld_l@ill.fr
>>>>>>>>>> <mailto:rietveld_l@ill.fr>) <rietveld_l@ill.fr
>>>>>>>>>> <mailto:rietveld_l@ill.fr>>
>>>>>>>>>> *Subject:* [EXT] Re: [External] Re: Step-like basline
>>>>>>>>>>
>>>>>>>>>>
>>>>>>>>>>
>>>>>>>>>> Some people who received this message don't often get email from
>>>>>>>>>> b...@indiana.edu <mailto:b...@indiana.edu>. Learn why this
>>>>>>>>>> is important
>>>>>>>>>> <https://urldefense.com/v3/__https:/aka.ms/LearnAboutSenderIdentification__;!!IKRxdwAv5BmarQ!fFhSHn4S5iEzkW-O9lvWG-OzoqK_2SKhRniGa71nxuOL3GcxiyD83i2mnNN0Z48HPkn4zjKqH-aqqA$>
>>>>>>>>>>
>>>>>>>>>> Hello Shay,
>>>>>>>>>>
>>>>>>>>>> I think it is probably related to "tube tails". You can
>>>>>>>>>> read about this in
>>>>>>>>>> the literature (e.g., on the BGMN web site) and you can
>>>>>>>>>> model it in some
>>>>>>>>>> Rietveld software such as Topas. You don't normally notice
>>>>>>>>>> this but it
>>>>>>>>>> becomes apparent with higher-intensity peaks.
>>>>>>>>>>
>>>>>>>>>>
>>>>>>>>>>
>>>>>>>>>> Regards,
>>>>>>>>>>
>>>>>>>>>> Dave
>>>>>>>>>> ------------------------------
>>>>>>>>>>
>>>>>>>>>> *From:* rietveld_l-requ...@ill.fr
>>>>>>>>>> <mailto:rietveld_l-requ...@ill.fr>
>>>>>>>>>> <rietveld_l-requ...@ill.fr
>>>>>>>>>> <mailto:rietveld_l-requ...@ill.fr>> on behalf
>>>>>>>>>> of Fernando Igoa <fer.igoa.1...@gmail.com
>>>>>>>>>> <mailto:fer.igoa.1...@gmail.com>>
>>>>>>>>>> *Sent:* Sunday, September 3, 2023 3:06 AM
>>>>>>>>>> *To:* Shay Tirosh <stiro...@gmail.com <mailto:stiro...@gmail.com>>
>>>>>>>>>> *Cc:* Rietveld List (rietveld_l@ill.fr
>>>>>>>>>> <mailto:rietveld_l@ill.fr>) <rietveld_l@ill.fr
>>>>>>>>>> <mailto:rietveld_l@ill.fr>>
>>>>>>>>>> *Subject:* [External] Re: Step-like basline
>>>>>>>>>>
>>>>>>>>>>
>>>>>>>>>>
>>>>>>>>>> This message was sent from a non-IU address. Please
>>>>>>>>>> exercise caution when
>>>>>>>>>> clicking links or opening attachments from external sources.
>>>>>>>>>>
>>>>>>>>>>
>>>>>>>>>>
>>>>>>>>>> Hey Shay,
>>>>>>>>>>
>>>>>>>>>>
>>>>>>>>>>
>>>>>>>>>> Are you using a motorized slit during the measurement?
>>>>>>>>>> These may open up
>>>>>>>>>> abruptly to compensate for the angular dependence of the
>>>>>>>>>> footprint and thus
>>>>>>>>>> generate an abrupt increase in the intensity.
>>>>>>>>>>
>>>>>>>>>>
>>>>>>>>>>
>>>>>>>>>> Hope it helps :)
>>>>>>>>>>
>>>>>>>>>>
>>>>>>>>>>
>>>>>>>>>> On Sun, Sep 3, 2023, 8:50 AM Shay Tirosh
>>>>>>>>>> <stiro...@gmail.com <mailto:stiro...@gmail.com>> wrote:
>>>>>>>>>>
>>>>>>>>>> Dear Rietvelders
>>>>>>>>>>
>>>>>>>>>> I am attaching a zoom-in on a diffraction profile.
>>>>>>>>>>
>>>>>>>>>> My question is what is the origin of the step-like profile
>>>>>>>>>> next to a very
>>>>>>>>>> large reflection peak?
>>>>>>>>>>
>>>>>>>>>> Is it a sample preparation problem?
>>>>>>>>>>
>>>>>>>>>> Is it part of the baseline?
>>>>>>>>>>
>>>>>>>>>> Thanks
>>>>>>>>>>
>>>>>>>>>> Shay
>>>>>>>>>>
>>>>>>>>>> --
>>>>>>>>>>
>>>>>>>>>>
>>>>>>>>>>
>>>>>>>>>>
>>>>>>>>>>
>>>>>>>>>>
>>>>>>>>>>
>>>>>>>>>>
>>>>>>>>>>
>>>>>>>>>>
>>>>>>>>>>
>>>>>>>>>> ++++++++++++++++++++++++++++++++++++++++++++++++++++++++++++++
>>>>>>>>>> Please do NOT attach files to the whole list
>>>>>>>>>> <alan.he...@neutronoptics.com
>>>>>>>>>> <mailto:alan.he...@neutronoptics.com>
>>>>>>>>>> >
>>>>>>>>>> Send commands to <lists...@ill.fr <mailto:lists...@ill.fr>>
>>>>>>>>>> eg: HELP as the subject with no body
>>>>>>>>>> text
>>>>>>>>>> The Rietveld_L list archive is on
>>>>>>>>>> http://www.mail-archive.com/rietveld_l@ill.fr/
>>>>>>>>>> <https://urldefense.com/v3/__http:/www.mail-archive.com/rietveld_l@ill.fr/__;!!IKRxdwAv5BmarQ!fFhSHn4S5iEzkW-O9lvWG-OzoqK_2SKhRniGa71nxuOL3GcxiyD83i2mnNN0Z48HPkn4zjJTf8rNHg$>
>>>>>>>>>> ++++++++++++++++++++++++++++++++++++++++++++++++++++++++++++++
>>>>>>>>>>
>>>>>>>>>> ++++++++++++++++++++++++++++++++++++++++++++++++++++++++++++++
>>>>>>>>>> Please do NOT attach files to the whole list
>>>>>>>>>> <alan.he...@neutronoptics.com
>>>>>>>>>> <mailto:alan.he...@neutronoptics.com>
>>>>>>>>>> >
>>>>>>>>>> Send commands to <lists...@ill.fr <mailto:lists...@ill.fr>>
>>>>>>>>>> eg: HELP as the subject with no body
>>>>>>>>>> text
>>>>>>>>>> The Rietveld_L list archive is on
>>>>>>>>>> http://www.mail-archive.com/rietveld_l@ill.fr/
>>>>>>>>>> ++++++++++++++++++++++++++++++++++++++++++++++++++++++++++++++
>>>>>>>>>>
>>>>>>>>>>
>>>>>>>>
>>>>>>>>
>>>>>>>> --
>>>>>>>> TU Bergakademie Freiberg
>>>>>>>> Dr. R. Kleeberg
>>>>>>>> Mineralogisches Labor
>>>>>>>> Brennhausgasse 14
>>>>>>>> D-09596 Freiberg
>>>>>>>>
>>>>>>>> Tel. ++49 (0) 3731-39-3244
>>>>>>>> Fax. ++49 (0) 3731-39-3129
>>>>>>>> ++++++++++++++++++++++++++++++++++++++++++++++++++++++++++++++
>>>>>>>> Please do NOT attach files to the whole list
>>>>>>>> <alan.he...@neutronoptics.com
>>>>>>>> <mailto:alan.he...@neutronoptics.com>>
>>>>>>>> Send commands to <lists...@ill.fr <mailto:lists...@ill.fr>>
>>>>>>>> eg: HELP as the subject with no body text
>>>>>>>> The Rietveld_L list archive is on
>>>>>>>> http://www.mail-archive.com/rietveld_l@ill.fr/
>>>>>>>> ++++++++++++++++++++++++++++++++++++++++++++++++++++++++++++++
>>>>>>>>
>>>>>>>
>>>>>>>
>>>>>>> --
>>>>>>> ______________________________________________
>>>>>>> Dr Alan Hewat, NeutronOptics, Grenoble, FRANCE
>>>>>>> <alan.he...@neutronoptics.com
>>>>>>> <mailto:alan.he...@neutronoptics.com>> +33.476.98.41.68
>>>>>>> http://www.NeutronOptics.com/hewat
>>>>>>> <http://www.neutronoptics.com/hewat>
>>>>>>> <http://www.neutronoptics.com/hewat>
>>>>>>> ______________________________________________
>>>>>
>>>>>
>>>>> -- TU Bergakademie Freiberg
>>>>> Dr. R. Kleeberg
>>>>> Mineralogisches Labor
>>>>> Brennhausgasse 14
>>>>> D-09596 Freiberg
>>>>>
>>>>> Tel. ++49 (0) 3731-39-3244
>>>>> Fax. ++49 (0) 3731-39-3129
>>>>>
>>>> ++++++++++++++++++++++++++++++++++++++++++++++++++++++++++++++
>>>> Please do NOT attach files to the whole list
>>>> <alan.he...@neutronoptics.com
>>>> <mailto:alan.he...@neutronoptics.com>>
>>>> Send commands to <lists...@ill.fr <mailto:lists...@ill.fr>> eg:
>>>> HELP as the subject with no body text
>>>> The Rietveld_L list archive is on
>>>> http://www.mail-archive.com/rietveld_l@ill.fr/
>>>> ++++++++++++++++++++++++++++++++++++++++++++++++++++++++++++++
>>
>>
>> --
>> TU Bergakademie Freiberg
>> Dr. R. Kleeberg
>> Mineralogisches Labor
>> Brennhausgasse 14
>> D-09596 Freiberg
>>
>> Tel. ++49 (0) 3731-39-3244
>> Fax. ++49 (0) 3731-39-3129
>>
>> ++++++++++++++++++++++++++++++++++++++++++++++++++++++++++++++
>> Please do NOT attach files to the whole list <alan.he...@neutronoptics.com>
>> Send commands to <lists...@ill.fr> eg: HELP as the subject with no body text
>> The Rietveld_L list archive is on
>> http://www.mail-archive.com/rietveld_l@ill.fr/
>> ++++++++++++++++++++++++++++++++++++++++++++++++++++++++++++++
>>


--
TU Bergakademie Freiberg
Dr. R. Kleeberg
Mineralogisches Labor
Brennhausgasse 14
D-09596 Freiberg

Tel. ++49 (0) 3731-39-3244
Fax. ++49 (0) 3731-39-3129

++++++++++++++++++++++++++++++++++++++++++++++++++++++++++++++
Please do NOT attach files to the whole list <alan.he...@neutronoptics.com>
Send commands to <lists...@ill.fr> eg: HELP as the subject with no body text
The Rietveld_L list archive is on http://www.mail-archive.com/rietveld_l@ill.fr/
++++++++++++++++++++++++++++++++++++++++++++++++++++++++++++++

Reply via email to