Hi,

Two things i would like to add:

1) Due to the dependence of A280 on amino acid composition, a simple two-wavelength 280 and 340 or 320 comparison is not very ideal for determining the scatter component of the UV absorption of protein/protein aggregate particles (the usefulness of this simple method for determining the degree of aggregation is not questioned). Instead, we use data points in the 320-350nm range from a UV absorption/scattering scan to plot a curve for determining the scattering component, which serves two purposes: a) to be used for subtracting the A280 to get real UV absorption of the protein at 280nm, b) to have an idea of how aggregated the sample is.

To do this, we plot a double log curve with absorption values A and wavelengths lambda at 320 330 340 350 nm, and fit it to the equation
log(A)=C - k*log(lambda)
Then the resulting equation can be used for extrapolating the scatter component at 280 nm. Of course the larger this component is, the more aggregated the protein is. The degree of aggregation (particle size and population) is also reflected by the k value. The larger the k is, the less aggregated the solution is.

An explanation of the theory behind this can be found here:
http://www.chem.agilent.com/Library/applications/59633927.pdf
Our favorite method is the one in the "scattering model" part, note that in figure 7 their molecule has UV absorption peak at 300, while in most of our cases, the proteins have peak at 280. There should be an academic article on this method. However I can not locate it right now. Apologies for that.

In fact, a quick look at the UV curve in the 300-350 nm range can directly tell us how badly aggregated a protein solution is. A perfectly not aggregating protein solution should have a UV absorption curve that quickly drops to nearly zero at 300 nm. One can test this by measuring an aggregation sample before and after filtration or centrifugation and see the dramatic change on the UV curve. (0.22um filter's pore sizes are quite comparable to the 200-400nm UV wavelengths, while most globular proteins are roughly 2-10 nm in diameter.)


2) Nanodrop's baseline function in the scan mode ("Protein 280" for example) is actually an all-point zeroing. At the blanking step, the instrument records a blank absorption curve and later uses this curve to subtract the sample curve. So when using a proper solution as blank, the sample curve is quite faithfully reflecting the UV absorption + scattering. I believe most UV spectrometers allow the users to do the same in the scan mode. So they are all good for this task.

Zhijie



-----Original Message----- From: Michael C. Wiener
Sent: Friday, February 21, 2014 12:16 PM
To: CCP4BB@JISCMAIL.AC.UK
Subject: Re: [ccp4bb] Aggregation assay

Analogous to Dr. Mirella, we've used the ratio of A320/A280 (for membrane proteins). In response to a fussy reviewer, I located some relevant references (refs. 10-13, copied below) when we referred to this in "A high-throughput differential filtration assay to screen and select detergents
for membrane proteins," Vergis et al., Anal. Biochem 407:1 (2010).

[10] S.J. Leach, H.A. Scheraga, Effect of light scattering on ultraviolet difference
spectra, J. Am. Chem. Soc. 82 (1960) 4790–4792.
[11] Y. Cordeiro, F. Machado, L. Juliano, M.A. Juliano, R.R. Brentani, D. Foguel, J.L.
Silva, DNA converts cellular prion protein into the b-sheet conformation and
inhibits prion peptide aggregation, J. Biol. Chem. 276 (2001) 49400–49409.
[12] G.J. Lee, A.M. Roseman, H.R. Saibil, E. Vierling, A small heat shock protein stably
binds heat-denatured model substrates and can maintain a substrate in a
folding-competent state, EMBO J. 16 (1997) 659–671.
[13] Y. Panyukov, I. Yudin, V. Drachev, E. Dobrov, B. Kurganov, The study of
amorphous aggregation of tobacco mosaic virus coat protein by dynamic light
scattering, Biophys. Chem. 127 (2007) 9–18.

Regards,

-MW

Michael C. Wiener, Ph.D.
Professor
Department of Molecular Physiology
and Biological Physics
University of Virginia
PO Box 800886
Charlottesville, VA 22908-0886
434-243-2731
434-982-1616 (FAX)

On Fri, 21 Feb 2014 15:05:46 +0000
"Tanner, John J." <tanne...@missouri.edu> wrote:
I am aware of an assay for aggregation (aggregation rate) that is based on fluorescence measurements. It involves excitation at 280 and emission in the range 260-400. Is there a reference for the absorbance method?

See

Nominé Y, Ristriani T, Laurent C, Lefèvre JF, Weiss E, Travé G. A strategy for optimizing the monodispersity of fusion proteins: application to purification of recombinant HPV E6 oncoprotein. Protein Eng. 2001 Apr;14(4):297-305. PubMed PMID: 11391022.

http://www.ncbi.nlm.nih.gov/pubmed/11391022

Jack Tanner

Sent from Jack's iPad

On Feb 21, 2014, at 7:23 AM, "Vivoli, Mirella" <m.viv...@exeter.ac.uk<mailto:m.viv...@exeter.ac.uk>> wrote:

Dear Prerana,

before starting the crystallization you could try to check the state of you protein, simply determining the Aggregation Index (AI) from measuring the absorbance at 280 nm and 340 nm. The AI is then computed using this simple formula: AI= 100 x (Abs340/(Abs 280 - Abs 340)). Soluble and non-aggregated proteins solutions typically have an Aggregation Index of 2 and lower, whereas for some aggregation will be 2-5; heavily aggregated proteins show an aggregation index >5. Of course you cannot use Nanodrop for it (because the wavelength for the baseline normalization is 340 nm). I would try to decrease the concentration of your protein to 4-5 mg/ml.Good luck.

Best,

Mirella


Vivoli Mirella

Postdoctoral Fellow
University of Exeter,
Biosciences
Biocatalysis Centre, Henry Wellcome Building
Stocker Road,
Exeter,
Ex4 4QD
Tel:+ 44 (0)1392 726121


________________________________
From: CCP4 bulletin board [CCP4BB@JISCMAIL.AC.UK<mailto:CCP4BB@JISCMAIL.AC.UK>] on behalf of Prerana G. [tracy...@gmail.com<mailto:tracy...@gmail.com>]
Sent: Friday, February 21, 2014 11:14 AM
To: CCP4BB@JISCMAIL.AC.UK<mailto:CCP4BB@JISCMAIL.AC.UK>
Subject: [ccp4bb]

Hi, Sorry for asking an off-topic question,
I have recently purified a protein having a molecular weight of 40kDa and concentration of the protein was 8mg/ml. When I tried to set the protein for crystallisation using micobatch method, the protein started precipitating in most of the buffer conditions of Crystal screen and Peg ion. The precipitation took place very quickly (within 5-10 mins).
How should I overcome this problem?


Regards
Prerana

Reply via email to