Hi to all,

@Leopoldo, thank you for providing the nice listing of the most relevant quantities determining the reflections' positions. However, I think there is a minor mistake in the description for the sample displacement: The given relation (-s·cos(theta) or in full -2s·cos(theta)/R with R = goniometer radius, e.g. Klug, Alexander 1974) implies that it is 0 at 180°2Theta, i.e., it continuously decreases with growing 2Theta. This also seems to be intuitive when you prepare a small sketch on how a misplaced sample height affects a reflection's position at low, moderate and high diffraction angles.

Best regards from Bremen,
Johannes


Am 29.03.2017 um 15:54 schrieb Cline, James Dr. (Fed):

Hi,

Seems everyone has beaten on this matter quite well. I humbly offer this paper: http://nvlpubs.nist.gov/nistpubs/jres/120/jres.120.013.pdf

which discusses how you can experimentally determine (eliminate) the zero error with considerable certainly.

You should consider what you know the most, and least, about: you can never know the sample displacement or attenuation errors, you might know the lattice parameters, and you can eliminate the zero error if you want to. You can refine all parameters with high angle data; but you must be careful to inspect results for loss of physical plausibility. I do this only as an adventurous exercise.

Regards,

Jim

James P. Cline
Materials Measurement Science Division
National Institute of Standards and Technology
100 Bureau Dr. stop 8520 [ B113 / Bldg 217 ]
Gaithersburg, MD 20899-8523    USA
[email protected] <mailto:[email protected]>
(301) 975 5793

*From:*[email protected] [mailto:[email protected]] *On Behalf Of *David L. Bish
*Sent:* Wednesday, March 29, 2017 9:41 AM
*To:* [email protected]
*Subject:* Re: using Zero

Hi Ana Isabel,

Following up on Stan's comments, we always perform this type of calibration with a certified standard, e.g., an NIST standard (Si or LaB6). We then measure data from below the lowest-angle peak up the high-angle limit of the instrument (about 155 deg. in our case). If you use a certified standard, you can FIX the unit-cell parameter(s) and refine zero error and specimen displacement corrections. I typically obtain good refinements, with small esd's, when I follow this procedure. After all, the purpose is not to obtain the unit-cell parameter of your standard. If you have the ability to adjust the zero error of your instrument, you can then do that after doing your refinements. Ultimately you should be able to achieve a very small zero error.

Make sure you then use that refined zero error in subsequent refinements, until you realign your instrument. The zero error should not be a function of your sample. Specimen displacement is a function of each sample (mount) and should always be refined to obtain more accurate (not precise) unit-cell parameters.

Cheers,

Dave

On 3/29/2017 9:23 AM, Leopoldo Suescun wrote:

    Hola Ana Isabel,

    As Stan said, if you go high enough in 2theta you have enough data
    to de-correlate zero and sample displacement corrections and you
    can refine them together with lattice parameters, and have correct
    values for the three.

    The reason for thi si that delta(2theta) for each peak is a fixed
    amount with zero (for all 2theta), proportional to -Scos(theta)
    with S sample displacement and proportional to arcsin(lambda/2d)
    with change of lattice parameters.

    Since these three terms affecting peak positions have different
    behaviors with 2theta, zero is constant, displacement is maximum
    at low and high 2theta but null at 90 and lattice parameter
    increases the shift with 2theta (as does the separation of Kalfa1
    and Kalfa2 peaks) then with a wide-enough patterns you will see
    the effects of the three affecting differently the low, mid and
    high 2theta peaks, making a refinement possible.

    But again, you need to go very high in 2theta to be able to
    de-correlate the three effects, collecting data from 10 to 80
    degrees won´t allow you to refine more than one of the three with
    confidence (assuming the two you fix are correct).

    Of course, aligning your diffractometer correctly and placing the
    sample at the place where it should be is always the best choice...

    Good luck with your work

    Leo

    2017-03-29 8:38 GMT-03:00 Julian Richard Tolchard
    <[email protected]
    <mailto:[email protected]>>:

        I would also suggest that it is easier to just check that your
        instrument zero is aligned than to mess around with long scans
        and fitting routines to separate correlated variables.


        Jools



        -----Original Message-----
        From: [email protected]
        <mailto:[email protected]>
        [mailto:[email protected]
        <mailto:[email protected]>] On Behalf Of "Lukasz
        Kruszewski"
        Sent: 29. mars 2017 13:25
        To: Ana Isabel Becerro Nieto <[email protected]
        <mailto:[email protected]>>
        Cc: [email protected] <mailto:[email protected]>
        Subject: Re: using Zero

        Hi. You can do that, of course, but you have to choose: you
        can use Zero error OR sample displacement, but never use both
        (this induces some physical impossibilities). However, I
        assume your diffractometer is calibrated, and the zero
        position of the detector is fine; the sample displacement may
        (I suppose) be connected with preparation-derived errors, and
        I'd use the latter parameter instead of the zero error.

        Hope this helps anyhow. Good luck!

        Luke Kruszewski


        > Dear all,
        >
        > I am using Si as internal standard to calibrate my pattern.
        Should I
        > refine the "Zero error" of the diffractometer if I am using the
        > calibrated pattern?
        >
        > Thanks a lot,
        >
        > Ana
        >
        >
        >
        > Dra. Ana Isabel Becerro
        >
        > Instituto de Ciencia de Materiales de Sevilla
        >
        > CSIC-US
        >
        >
        >
        >
        >
        >
        >
        > ---
        > El software de antivirus Avast ha analizado este correo
        electrónico en
        > busca de virus.
        > https://www.avast.com/antivirus
        > ++++++++++++++++++++++++++++++++++++++++++++++++++++++++++++++
        > Please do NOT attach files to the whole list
        > <[email protected]>
        <mailto:[email protected]> Send commands to
        <[email protected] <mailto:[email protected]>> eg:
        > HELP as the subject with no body text The Rietveld_L list
        archive is
        > on http://www.mail-archive.com/[email protected]/
        > ++++++++++++++++++++++++++++++++++++++++++++++++++++++++++++++
        >
        >


        --
        Łukasz Kruszewski, Ph.D., adjunct
        Polish Academy of Sciences
        Institute of Geological Sciences
        Twarda 51/55 str.
        00-818 Warsaw
        Poland


        ++++++++++++++++++++++++++++++++++++++++++++++++++++++++++++++
        Please do NOT attach files to the whole list
        <[email protected]>
        <mailto:[email protected]>
        Send commands to <[email protected] <mailto:[email protected]>>
        eg: HELP as the subject with no body text
        The Rietveld_L list archive is on
        http://www.mail-archive.com/[email protected]/
        <http://www.mail-archive.com/[email protected]/>
        ++++++++++++++++++++++++++++++++++++++++++++++++++++++++++++++



--
    Dr. Leopoldo Suescun
    Prof. Agr (Assoc. Prof.) de Física  Tel: (+598) 29290705/29249859
    Cryssmat-Lab./Cátedra de Fisica/DETEMA  Fax: (+598) 29241906*

    e-mail: [email protected]/[email protected]
    <http://[email protected]/[email protected]>
    Facultad de Quimica, Universidad de la Republica. Montevideo, Uruguay


    Ahora la cristalografía importa más (www.iucr.org
    <http://www.iucr.org/>) Crystallography Matters more.




    ++++++++++++++++++++++++++++++++++++++++++++++++++++++++++++++

    Please do NOT attach files to the whole list<[email protected]> 
<mailto:[email protected]>

    Send commands to<[email protected]> <mailto:[email protected]>  eg: HELP as the 
subject with no body text

    The Rietveld_L list archive is 
onhttp://www.mail-archive.com/[email protected]/

    ++++++++++++++++++++++++++++++++++++++++++++++++++++++++++++++



--
David L. Bish
Department of Geological Sciences
Indiana University
1001 E. 10th St.
Bloomington, IN 47405
812-855-2039


++++++++++++++++++++++++++++++++++++++++++++++++++++++++++++++
Please do NOT attach files to the whole list <[email protected]>
Send commands to <[email protected]> eg: HELP as the subject with no body text
The Rietveld_L list archive is on http://www.mail-archive.com/[email protected]/
++++++++++++++++++++++++++++++++++++++++++++++++++++++++++++++


--
Johannes Birkenstock
Universität Bremen - FB 5 (GEO 2300)
Klagenfurter Str. 2-4
28359 Bremen
Tel. +49 421 218 65165

++++++++++++++++++++++++++++++++++++++++++++++++++++++++++++++
Please do NOT attach files to the whole list <[email protected]>
Send commands to <[email protected]> eg: HELP as the subject with no body text
The Rietveld_L list archive is on http://www.mail-archive.com/[email protected]/
++++++++++++++++++++++++++++++++++++++++++++++++++++++++++++++

Reply via email to