Wojciech pointed out an important issue regarding ADS measurements:

Wojciech Paszkowicz schrieb:

Dear All,

Additional comment concerning the usefulness of variable slits for structure
refinement: The angular distribution of the radiation from an X-ray tube is
not necessarily uniform, so using variable slits for collection of
Rietveld-refined (Bragg-Brentano geometry) data may imply the need for an
additional intensity correction for this effect.
Because the intensity distribution versus take-off angle of a tube is a priori unknown, one must find an working compromise, see again our comments at
http://www.bgmn.de/vardiv.html
Our comment: "Maximum divergence angle should be much less than take-off angle of X-rays from tube anode. The latter angle usually is 6°. "

An additional restriction of the maximum slit opening may be given by the limited length of a secondary beam monochromator. The crystal may act like a fixed antiscatter slit and may cut the intensity if a certain ADS divergence is reached. But this can be modeled by the Monte-Carlo approach.

Moreover, the opening of
the variable slits requires an extremely high precision mechanics so there
may be a doubt  concerning the smoothness of the flux variation with varying
two theta, at the given instrument.

This is true for conventional stepper-motor driven slits at low angles, say below 15 °2theta. In that range the smallest steps of the slit may be too large to fix the irradiated sample length over two theta. But the intensity data at very low angles may be biased by a number of additonal factors like sample roughness and misalignment of the sample surface. Thus, high accuracy/reproducibility of the intensity at low angles cannot be expected at all, and excluding this range from the refinement is common practice. For higher angles, my personal experiences with ADS are positive, even for older instruments (built from 1986 to 1995, from 3 different manufacturers, 3 different types of stepper-motor driven ADS systems). Other people have been less happy with their ADS system, see
Peplinski, B. & Wenzel, J. (2000) Mat.Sc.Forum, Vols. 321-324, pp. 144-149.
It should be pointed out that it is necessary to check the a alignment and programming of the ADS in the way that Ian Madsen recommended in a previous mail (refining the temperature factors of a known standard). See also the literatur cited above. For a first coarse check of the primary beam ADS, a simple look on a fluorescent screen with a scale during a measurement (2° step, 2 sec per step) helps to see coarse angular fluctuations of the irradiated length, for example by a wrong programming. To check the zero point alignment of the slit, a "divergence slit scan" from negative to positive opening at 0 °2theta without sample (of course at reduced tube power and with absorber) can be done. If the instrument or the control software is not flexible enough to do this, one should ask the manufacturer service for help respective for a demonstration that the ADS is correctly aligned. It is clear that such additional efforts that are neccessary for the correct use of the ADS system (compared to fixed slit systems) must be counterbalanced to the advantages of an ADS. Here, everyone must do his own judging.
Regards
Reinhard Kleeberg