Be careful you havent picked up the staraniso output which has already
truncated data..
Sometimes this is a good thing but as users we need to know the
alternatives!
Eleanor

On Tue, 26 Nov 2019 at 10:16, Matthias Oebbeke <
[email protected]> wrote:

> Thank you all, for your replies! Some of you suggested that there
> could be reflections at the corners of the detector, which are 'cutted
> by the edges'. This also came to my mind, but the data was collected
> until 1.7A at the edges.
>
> I will take the advice to refine against the low and the high
> resolution cutted datasets and compare them. (For high resolution
> limit refinement I got a R-work of 0.2098 and R-free of 0.2655 after a
> few refinement cycles with phenix using isotropic parameters)
>
> Thanks again for your help.
>
>
> Best regards,
> Matthias Oebbeke
>
>
>
>
> Zitat von "Winter, Graeme (DLSLtd,RAL,LSCI)" <[email protected]
> >:
>
> > The data are > 90% complete at low resolution, where the
> > completeness is more important, so this is probably fine - you would
> >  probably benefit from better detector placement in the future as
> > the data are good to the edge.
> >
> > Was the detector offset or just too far away? 120° for an oP lattice
> > should be relatively complete - as appears to be the case for low
> > angle data here.
> >
> > The reflections you have measured are almost certainly a better
> > estimate of F^2 than 0, which is what you are saying if you do throw
> > them away :-)
> >
> > Best wishes Graeme
> >
> >> On 25 Nov 2019, at 13:11, Matthias Oebbeke
> >> <[email protected]> wrote:
> >>
> >> Dear ccp4 Bulletin Board,
> >>
> >> I collected a dataset at a synchrotron beamline and got the
> >> statistics (CORRECT.LP) after processing (using xds) shown in the
> >> attached pdf-file.
> >>
> >> Do you think this dataset is usable, due to its low completeness?
> >> As you can see in the attached file the completeness is just 50% in
> >> the highest resolution shell, whereas the I over Sigma is above 2
> >> and also the CC 1/2 (80%) and the R factors (36.8%) have reasonable
> >> values. Furthermore the overall statistic are good regarding R
> >> factor, CC 1/2 and I over Sigma. The only problem seems to be the
> >> completeness. If I would set the cut-off at a lower resolution to
> >> get good completeness, I would throw away nearly half of my
> >> reflections.
> >>
> >> I'm happy to here your opinion. In Addition to that: The space
> >> group is orthorhombic and the dataset was collected over 120° using
> >> fine slicing (0.1°).
> >>
> >>
> >> Best regards,
> >>
> >> Matthias Oebbeke
> >>
> >>
> >> --
> >> Matthias Oebbeke, M.Sc.
> >> Research Group of Professor Dr. G. Klebe
> >> Institute of Pharmaceutical Chemistry
> >> Philipps-University Marburg
> >> Marbacher Weg 6, 35032 Marburg, Germany
> >> Phone: +49-6421-28-21392
> >> Mail: [email protected]
> >> www.agklebe.de
> >>
> >> ########################################################################
> >>
> >> To unsubscribe from the CCP4BB list, click the following link:
> >> https://www.jiscmail.ac.uk/cgi-bin/webadmin?SUBED1=CCP4BB&A=1
> >> <CORRECT_LP.pdf>
> >
> >
> > --
> > This e-mail and any attachments may contain confidential, copyright
> > and or privileged material, and are for the use of the intended
> > addressee only. If you are not the intended addressee or an
> > authorised recipient of the addressee please notify us of receipt by
> > returning the e-mail and do not use, copy, retain, distribute or
> > disclose the information in or attached to the e-mail.
> > Any opinions expressed within this e-mail are those of the
> > individual and not necessarily of Diamond Light Source Ltd.
> > Diamond Light Source Ltd. cannot guarantee that this e-mail or any
> > attachments are free from viruses and we cannot accept liability for
> > any damage which you may sustain as a result of software viruses
> > which may be transmitted in or with the message.
> > Diamond Light Source Limited (company no. 4375679). Registered in
> > England and Wales with its registered office at Diamond House,
> > Harwell Science and Innovation Campus, Didcot, Oxfordshire, OX11
> > 0DE, United Kingdom
>
>
>
> --
> Matthias Oebbeke, M.Sc.
> Research Group of Professor Dr. G. Klebe
> Institute of Pharmaceutical Chemistry
> Philipps-University Marburg
> Marbacher Weg 6, 35032 Marburg, Germany
> Phone: +49-6421-28-21392
> Mail: [email protected]
> www.agklebe.de
>
> ########################################################################
>
> To unsubscribe from the CCP4BB list, click the following link:
> https://www.jiscmail.ac.uk/cgi-bin/webadmin?SUBED1=CCP4BB&A=1
>

########################################################################

To unsubscribe from the CCP4BB list, click the following link:
https://www.jiscmail.ac.uk/cgi-bin/webadmin?SUBED1=CCP4BB&A=1