It's an interesting discussion.

1.Usually it's not possible to use mass-spec for non-covalent complexes. 

2. Most methods depend on macromolecule shape and concentration. 

3. SAXS method looks limited to me. It uses diluted monodisperse solutions. 
That excludes complexes that can associate/dissociate. How can you calculate Kd 
from it?

4. All methods for determination of multimeric state using separation technique 
depend on three different cases: time of equilibrium (teq)>> time of separation 
(tsep), or teq << tsep or teq ~ tsep. Even without equilibrium, you cannot have 
only one component. For a successful separation, you would want teq >> tsep, 
which is less likely in AUC method.

5. There are papers on capillary electrophoresis methods where they study 
exactly these effects (time of equilibrium vs time of separation). 

Maia



----- Original Message ----- 
From: "aidong" <a...@xmu.edu.cn>
To: <CCP4BB@JISCMAIL.AC.UK>
Sent: Sunday, July 04, 2010 3:01 AM
Subject: Re: [ccp4bb] monomeric coiled coil--updated


> In light of several wonderful responses,I would like to provide an  
> update for this question:
> 
> 1. I would agree that SEC might not be able to identify monomer vs  
> multimer forms for this likely rod-shaped protein.
> 
> 2. It is extremely low kd for dimer. AUC and SAXS experiments have  
> measured its kd at ~0.1 mM.
> 
> 3. MALS might not be able to pick up dimer form since it might be only  
> a few percent when the concentration is low. We might overcome  
> concentration effect by direct injection to dawn heleos and refraction  
> index.
> 
> 4. Mass spec has found both monomer and dimer forms although the  
> abundance of each one is not known.
> 
> 5. Intramolecular coiled coil is quite possible since intermolecular  
> dimer is unstable. We hope our structure might provide an answer.
> 
> Many thanks for your time and ideas
> 
> Cheers
> 
> Aidong
> 
> 
> On Jul 4, 2010, at 1:09 AM, Anastassis Perrakis wrote:
> 
>> A few thoughts on these, since I do not fully agree.
>>
>> 1. Detection by light scattering is a method that can be used either  
>> without separation, or while separating.
>> If you have a scattering detector, you can stick in a cuvette, or  
>> stick it to the end of a column, your choice.
>>
>> 2. Sec is not a good method to show if especially a coiled coil is  
>> monomer-multimer. A long coil, will
>> have a hydrodynamic radius bigger than its MW, thus any prediction  
>> based on SEC will be misleading,
>> especially for this class of proteins.
>>
>> 3. In AUC (although I am not an expert at it at all) I cant see the  
>> connection between the disassociation time
>> and the run time. In sedimentation or equilibrium runs, depending on  
>> what you want to see, I think you can look
>> at monomer-multimer equilibrium over a wide range of kD and  
>> combinations of k(on) and k(off).
>>
>> 4. The physiological concentration is a bit misleading. First, its  
>> clear now that cells have microenvironments,
>> and 'physiological' concentrations are hard to define. Also, in a  
>> cell, I think (and I think others tend to agree)
>> that kD plays little role at the end. kD is a combination of k(on) -  
>> which is concentration dependent but in a cell
>> very likely diffusion limited - and of k(off) which I think is what  
>> matters most in the cell.
>>
>> Going to Aidong's question, I think that MALLS was a good  
>> experiment. The fact that these constructs do no associate,
>> can mean that
>>
>> a. the prediction is wrong - likely with these scores, but not  
>> necessary
>> b. the kD in solution is indeed higher that the concentration you  
>> used for MALLS
>> c. The constructs are not well chosen for some reason
>>
>> You could use AUC to detect kD as high as ~100uM, depending on the  
>> concentration of the start sample of course.
>> The next question will anyway be if that kD has any sort of  
>> physiological significance - which you cannot tell by magnitude -
>> so you are back at the drawing board for mutants. Three years later  
>> the referees will still not believe it ... sorry, now it gets  
>> personal,
>> so I stop here.
>>
>> My two cents.
>>
>> A.
>>
>>
>> On 3 Jul 2010, at 18:10, chern wrote:
>>
>>> The multimeric state depends on a protein concentration. You can  
>>> get any
>>> multimer to dissociate if you dilute it to low enough  
>>> concentration.  If
>>> your complex is a homodimer, then Kdiss=[complex]/[monomer]^2.  
>>> Let's say
>>> your Kdiss~10^(-3)M,  and your protein concentration is ~10^(-4)M,  
>>> then
>>> [complex]=Kdiss/[monomer]^2=10^(-3)/10^(-4)^2=10^(-5), that means,  
>>> the dimer
>>> concentration is approximately ~10 times less then the monomer  
>>> concentration
>>> at this particular protein concentration. Let's say, the mol weight  
>>> is 50
>>> kDa, then at 5mg/ml you will have only about ~10% of the dimer. Of  
>>> course,
>>> if your Kdiss~10^(-4)M, then you will have approximately similar
>>> concentrations of monomers and dimers at 10^(-4).
>>> Because this is a dynamic equlibrium between multimers and  
>>> monomers, some
>>> methods are not good for the determination of a multimeric state.  
>>> Some
>>> reviewers demand to prove the multimeric state by size-exclusion
>>> chromatography (SEC) or analytical centrifugation. The analytical
>>> ultracentrifugation method will not work, as the characteristic  
>>> time of the
>>> dissociation/association is much lower than the centrifugation time  
>>> (`24
>>> hours). The separated monomer will start association and the  
>>> separated dimer
>>> will start dissociation according to Kdiss and the bands will be  
>>> smeared.
>>> SEC is faster, like half an hour, it gives you a better chance. The  
>>> methods
>>> without separation are the best Like light scattering), just make  
>>> protein
>>> concentration high. Here comes the other question. What is the  
>>> physiological
>>> concentration. You want to be close to it. I read some literature  
>>> on this
>>> and it looks like it is between 10^-(4) to 10^-(6) for majority of  
>>> proteins.
>>>
>>>
>>>
>>>
>>>
>>>
>>>
>>> ----- Original Message -----
>>> From: "aidong" <a...@xmu.edu.cn>
>>> To: <CCP4BB@JISCMAIL.AC.UK>
>>> Sent: Saturday, July 03, 2010 6:26 AM
>>> Subject: [ccp4bb] monomeric coiled coil
>>>
>>>
>>>> Sorry for this ccp4 unrelated question.
>>>>
>>>> We recently have a protein that a multicoil program
>>>> (http://groups.csail.mit.edu/cb/multicoil/cgi-bin/multicoil.cgi/cgi-bin/multicoil
>>>> ) predicts to have very high probability for dimer and trimer.   
>>>> Their
>>>> scores are close to 0.4 and 0.6 for lengths of more than 60 amino   
>>>> acids.
>>>> However, two constructs that cover this region have  demonstrated  
>>>> monomers
>>>> in solutions by Multiangle light scattering?!    For the same  
>>>> question, we
>>>> could not get any response from this program  manager therefore we  
>>>> turn to
>>>> ccp4 for help.  We wonder whether some of  you might have similar
>>>> experience. Thank you in advance.
>>>>
>>>> Sincerely,
>>>>
>>>> Aidong
>>>>
>