I get it, thanks for the block diagrams.  I will look into doing a lab test 
with a calibrated source.
________________________________
From: Marcus D. Leech <patchvonbr...@gmail.com>
Sent: Tuesday, September 29, 2020 2:23 PM
To: Mark Koenig <mark.koe...@iubelttechnologies.com>
Cc: usrp-users@lists.ettus.com <usrp-users@lists.ettus.com>
Subject: Re: [USRP-users] TWINRX Gain

On 09/29/2020 02:02 PM, Mark Koenig wrote:
Yes, I understand gain varies with temperature and frequency, I just wasnt sure 
if there was any receive chain analysis performed with the daughtecard to give 
the developer an idea of what type of gain is provided over the attenuation 
range at various frequencies.  I am not too concerned about tenths of dBs....I 
was just interested in what the actual gain range provided by the card is.

Thanks

Mark
________________________________

If you look at the first page of this:

https://files.ettus.com/schematics/twinrx/TwinRX%20RF%20Board%20Rev%20D.pdf

You can see the overall block diagram.  You can also see several PE43503 
attenuators, sprinkled among several different MMIC amplifiers, and
 various different RF pathways through switches and filters depending on band.  
 It would be hard for me to unwind all of that and give you
 a definitive answer.

Even for the IF processing, there are two different IFs, depending on the 
frequency band--again with various distributions of gain and
  attenuation (either explicit attenuation, or attenuation via filtering)--all 
of which have considerable uncertainty--due to batch-to-batch
  variability and temperature effects.   I'm fairly sure that even the designer 
of the board couldn't tell you, for any given board configuration
  what the actual gain measured between the antenna input an the ADC input 
actually was, with better than 5dB confidence.  Which is where
  calibration comes in.

https://files.ettus.com/schematics/twinrx/TwinRX%20IF%20Board%20Rev%20C.pdf

In a laboratory instrument, like a spectrum analyser, all of this is 
painstakingly calibrated at the factory, usually using lookup tables (or the
  analog-era equivalent), based on well-characterized calibration sources.  So 
when you set the gain level on the front-panel of the device
  to some dB value, you'll actually get that value at the measurement point and 
when you look at the measurement on the display and it
  says -70dBm, it's actually -70dBm at the input terminal.  SDRs aren't that, 
typically.  Although one could build a fairly nice lab instrument
  *around* an SDR, using all the aforementioned calibration exercises, etc.

Now, this all, I admit, sounds a tad "lecturey".  I know you probably know all 
of this, but many on the list don't, or perhaps haven't thought about
  it much.  So, I'm prompted to deliver this, or a very similar "lecture" a few 
times a year due to similar queries to yours.
_______________________________________________
USRP-users mailing list
USRP-users@lists.ettus.com
http://lists.ettus.com/mailman/listinfo/usrp-users_lists.ettus.com

Reply via email to