Re: [ccp4bb] High Rmerge with thin frames

2010-11-06 Thread Kay Diederichs

Am 20:59, schrieb Harry Powell:
...

I think there may be issues with collecting data too finely with a
Pilatus, even in shutterless mode. I don't know where the problems arise
(can't be shutter/rotation axis synchronisation), but it seems to be the
normal thing in XDS (which should have no problems with fine
phi-slicing) to use the "PATCH_SHUTTER_PROBLEM=TRUE" that Martin
Hallberg suggested, which looks a bit like a fudge to me (but I expect
Kay to correct me on that!).

...

Hi Harry,

no, it is a misunderstanding that the normal thing is to use 
PATCH_SHUTTER_PROBLEM=TRUE in XDS; rather, it may be the last resort to 
try before you abandon a dataset, and if the spindle/shutter 
de-synchronization is so poor that this fudge needs to be used, then the 
BL hardware needs to be fixed before other datasets are collected.


Two other points:
a) one may distinguish "weak data resulting in high R-factors" from 
"problems with hardware or radiation damage resulting in high R-factors" 
by looking at (I/sigma)^asymptotic as defined in Acta D66, 733 ( 
http://dx.doi.org/10.1107/S0907444910014836 ).
b) Marcus Müller (SLS) has shown (to me, unambiguously) that the best 
data are obtained from the Pilatus if delta-phi is 1/2 to 1/4 of the 
mosaicity (as reported by XDS). However, for the best data, other 
parameters of the experiment (basically the transmission and the spindle 
speed) have to be optimized as well. The finding of Tassos that 0.5-1 
degree data gave the best result in a particular case might indicate 
that these other factors of his experiment were not optimal (it is 
indeed not trivial to get them right).


best,
Kay



smime.p7s
Description: S/MIME Cryptographic Signature


Re: [ccp4bb] High Rmerge with thin frames

2010-11-06 Thread harry powell

Hi All

Both Martin and Kay ( in a later message) have misinterpreted what I  
wrote - what I meant was that it seems normal in using XDS with  
Pilatus data, not the normal thing to do with data from other detectors.


I had found a number of scripts on the web that deal specifically  
with processing Pilatus data with XDS that seemed to use  
"PATCH_SHUTTER_PROBLEM=TRUE" - what I didn't notice was that the  
commands had all been commented out, and that "FALSE", the default,  
was in operation.


So, in either case I was wrong. Apologies to Kay and Wolfgang, and  
all who write and use scripts for processing Pilatus data with XDS.


On 5 Nov 2010, at 18:33, Martin Hallberg wrote:


Hi Harry,

On Nov 5, 2010, at 5:45 PM, Harry Powell wrote:


I think there may be issues with collecting data too finely with a  
Pilatus, even in shutterless mode. I don't know where the problems  
arise (can't be shutter/rotation axis synchronisation), but it  
seems to be the normal thing in XDS (which should have no problems  
with fine phi-slicing) to use the "PATCH_SHUTTER_PROBLEM=TRUE"  
that Martin Hallberg suggested, which looks a bit like a fudge to  
me (but I expect Kay to correct me on that!).


It is not the normal thing in XDS but it is perhaps a relatively  
common solution to shutter/spindle synchronisation problems  
discovered afterwards (always process directly at the beam line!).  
The default in XDS is indeed PATCH_SHUTTER_PROBLEM=FALSE


Compensating like this is of course not the best (go and  
recollect!) but still way better than unusable data in the  
meantime. In the case Sergei originally described it would at least  
indicate what the problem may be. Sergei did not say which detector  
was used for the data collection so we don't know if it was a  
Pilatus or a CCD. Maybe Sergei can fill us in on the details?


BTW, any views in the community on crystal lifetime with continuous  
data collection like on the Pilatus (or AXIOM) versus letting the  
crystal rest/cool/die(?) a second between frames on a CCD?


Cheers,

Martin







Harry
--
Dr Harry Powell,
MRC Laboratory of Molecular Biology,
Hills Road,
Cambridge,
CB2 0QH


Re: [ccp4bb] High Rmerge with thin frames

2010-11-06 Thread Guenter Fritz

Dear Ronnie,
we are working with weak diffracting crystals. Many tests (and taking 
into account  Marcus Muellers results) showed that 0.3-0.5 of XDS 
mosaicity, (very) low dose, and high redundancy give the best results. 
Our crystals diffract between 7 and 4 A. At low dose I do not check 
diffraction really by eye. I collect 200 images (fine slicing), process 
and check XDS statistics. (200 images is not much when the dataset 
comprises 9000-12000 images). Especially for anomalous data this 
strategy was pivotal.
We did also comparisons between CCD and PILATUS. We used one long 
crystal and collected on end end a MAD dataset using a  CCD and 
afterwards  at the other end a MAD dataset on PILATUS, same dose per 
degree.  Results were absolutely clear that the fine slicing give much 
better results.

Best,
Guenter

Dear Tassos,

I'm interested in your third point. Do you have any explanation for 
why 0.5-1 degrees oscillation gave better data? Purely due to the fact 
that the crystals survived longer and thus yielded higher redundancy 
data, or also other parameters?
Also do anyone know where the threshold lies for when /not/ to use 
fine phi slicing on the PILATUS? ie, at what level of diffraction 
would one need to increase the exposure (and oscillation in order to 
still get redundant data)?


We'll be in a similar position in the coming weeks with data 
collection using PILATUS detectors, and would like to maximize the 
potential data quality from our weak diffracting crystals. Any input 
on this would be greatly appreciated!


Cheers,
Ronnie Berntsson



On Nov 5, 2010, at 16:16, Anastassis Perrakis wrote:


three additional points:

1.


OTOH, if "The diffraction is quite weak", one may be limited by counting
statistics.  This also cannot be overcome by processing.


As JIm suggests above then, maybe you should look if the 15% Rmerge 
is almost reasonable given the specific I/sigI at low resolution?



2. If there is one thing I do not like in XDS, is that there is no 
(or I have failed to find) statistics of I/sigI and Rmerge as 
function of image.

Have a look at the SCALA output. Maybe some images are bad?

3. making too fine slices of too weak diffraction images ends up with 
either too weak counting statistics or inability to 'lock' the 
refinement.
we did that for one crystal form, collecting 0.1, 0.2, 0.35, 0.5, 
0.7, 1.0 from various crystals (with the same dose per degree, at SLS 
using a PILATUS, mosaicity 0.4-0.6) in an attempt to get better Se 
signal. We miserably failed to get any useful signal at the end, but 
learned that for these very weak diffracting plates (submicron) 
collecting 0.5-1.0 degrees was actually giving at the end better data.


A.





--
***

Priv.Doz.Dr. Guenter Fritz
Fachbereich Biologie
Universitaet Konstanz
http://www.biologie.uni-konstanz.de/fritz

e-mail: guenter.fr...@uni-konstanz.de

e-mail: guenter.fr...@uniklinik-freiburg.de
http://www.uniklinik-freiburg.de/neuropathologie/live/forschung/ag-g-fritz.html
Tel.: +49 761 270 5078
Fax.: +49 761 270 5050



Re: [ccp4bb] High Rmerge with thin frames

2010-11-06 Thread Sergei Strelkov

Dear All,

first of all, I would like to thank the many
good people who have responded to my query.
Yet another truly interesting discussion on this BB!

As a partial summary, two points:

1. Few people suggested that our high Rmerge problem
could be caused by experimental troubles
like phi angle imprecision rather than the crystal itself.
Thus far we could not find indications for that but maybe
we did not look carefully enough. However, the
frames ARE weak, and I tend to think
that this is the main cause of the problem.

2.  Few people argued that data processing programs
(XDS in particular) 'should' handle thin frames
as efficiently as thicker frames...
After digesting these (very useful!) arguments,
I still tend to think that the best proof would be in the pudding,
i.e. trying to pool thin frames into thicker frames
(but I do not have an immediate means of doing this... )

Sergei




Dear All,

I am processing a dataset collected (not by me) with 0.1 degree
oscillations.
The diffraction is quite weak even though there is a clean diffraction
pattern to about 3A.

Either Mosflm or XDS processes the data readily with +/- default settings
but both yield a high overall Rmerge of about 0.23 in the expected symmetry.
Processing in P1 yields an overall Rmerge of ~0.18, but what is
especially disappointing
is that Rmerge is as high as 0.15 at ~5A resolution already.

The question is, how can we process the data so that the merging statistics
becomes more reasonable?

Apparent mosaicity turns out to be ~0.5A. My naive way of thinking is
to try treating each five consecutive frames as a single 0.5 degree frame.
Does anyone have experience with this?

Many thanks in advance,
Sergei





--
Prof. Sergei V. Strelkov
Laboratory for Biocrystallography
Department of Pharmaceutical Sciences, Katholieke Universiteit Leuven
O&N2, Campus Gasthuisberg, Herestraat 49 bus 822, 3000 Leuven, Belgium
Work phone: +32 16 330845  Fax: +32 16 323469 OR +32 16 323460
Mobile: +32 486 294132
Lab pages: http://pharm.kuleuven.be/anafar


Re: [ccp4bb] High Rmerge with thin frames

2010-11-06 Thread Kay Diederichs

Am 20:59, schrieb Sergei Strelkov:
...

I still tend to think that the best proof would be in the pudding,
i.e. trying to pool thin frames into thicker frames
(but I do not have an immediate means of doing this... )


Sergei,

try to locate 2pck and 2pck.man which were part of former XDS 
distributions (there's a link to a Linux binary in article 
http://strucbio.biologie.uni-konstanz.de/xdswiki/index.php/2pck ).


2pck can add several images, and writes them out as one new image. After 
the conversion, you may only have to change the filetype, the 
oscillation range and the number of images in your XDS.INP . Pls tell us 
what you find!


HTH,

Kay




smime.p7s
Description: S/MIME Cryptographic Signature