Unrestricted Document

Shay,

As someone working for one of the vendors, I agree with most of the advice 
already given.  Most importantly, send samples or at least request data from a 
NIST certified standard that is roughly similar to what you will be studying.  
It is the only way to truly evaluate instrument performance.  As much as we try 
to thoroughly understand diffraction physics, sometimes we are surprised by the 
interaction of a specific sample with our instrumentation.  This also helps you 
avoid getting tricked by the numbers game- it doesn't really matter if a 
detector has 197 vs 255 channels, a 145 mm vs 150 mm radius,  2.17 deg/sec or 
2.5 deg/sec slew rate, or any of the other silly numbers that get promoted as 
the "differentiating" factor.  Any individual number is meaningless if not 
taken into consideration with the entirety of the instrument design (for 
example, a bigger detector isn't necessarily faster if the goniometer radius is 
also bigger).

I recommend that you specify your requested scan parameters as range, total 
time, and minimum number of data points in the top half of the peak (for 
example, 5 to 90 deg 2theta in 10 minutes with at least 7 data points above the 
half maximum) and look at the data in total counts rather than count rate.  
Different vendors use different definitions for counting time which can confuse 
the calculation of count rate.  Also, optimal step size varies with pixel size, 
goniometer radius, and subsampling rate so you will not get identical step 
sizes from the different instruments.

If anybody is interested: for linear position sensitive detectors, I have seen 
counting time defined as the time one data point is observed by one channel or 
as the time one data point is observed by all channels of the detector.  Both 
definitions are correct as long as they are clearly defined, but I do have a 
strong preference for the latter since it inherently normalizes for detectors 
that have a different number of channels.

Also, be clear if you need an instrument that is optimized for a particular 
material class or if you need a versatile instrument for a wide range of 
materials.  Vendors can configure an instrument that performs very well for 
something like low angle clay analysis, but it might compromise other 
measurements.  And, of course, a generalized instrument may not give the best 
possible low angle data.

I do have a couple personal opinions to share, but please keep in mind that I 
am not totally impartial.

I think the decision between theta-theta or omega-2theta goniometer can be 
important.  I have a strong preference for theta-theta goniometers after having 
run a university lab where poorly prepared samples spilled out of the sample 
holders on omega-2theta goniometers.

Finally, do not believe any vendor's marketing material (ours included) that 
claims that specimen fluorescence can be totally solved by energy 
discriminating detectors.  The detector can filter out fluoresced noise, but it 
cannot regain the signal that was lost by absorption of the X-rays and cannot 
correct for the loss of penetration depth that may compromise your particle 
statistics.  Energy discrimination is useful to filter out noise from the 
occasional fluorescent material, but if you are mostly working with Fe-bearing 
materials then you should get a Co tube or other non-Cu anode.  Cu anodes are 
the most common, but they are not the "industry standard" as some people think.

Again, I will be upfront that I work for one of the vendors.  But this advice 
is mostly based as someone who bought instruments from three different vendors 
when I managed a university lab.

Regards,
Scott Speakman
Principal Scientist- XRD
Malvern Panalytical





Unrestricted Document
From: rietveld_l-requ...@ill.fr <rietveld_l-requ...@ill.fr> On Behalf Of Bish, 
David L
Sent: Wednesday, June 14, 2023 6:50 AM
To: Shay Tirosh <stiro...@gmail.com>; iangie <ian...@126.com>
Cc: rietveld_l@ill.fr
Subject: Re: [External] Re:recommended questions to XRD supplier.

Some people who received this message don't often get email from 
b...@indiana.edu<mailto:b...@indiana.edu>. Learn why this is 
important<https://aka.ms/LearnAboutSenderIdentification>
CAUTION: This email originated from outside of the organization. Do not click 
links or open attachments unless you recognize the sender and know the content 
is safe.
Hi Shay,

I echo some of the other comments. A few other things: Learn how alignment will 
be performed (e.g., is it a black-box arrangement?) and how well you can 
produce close to 0.000 zero error. Is an independent water chiller required? 
How expensive and available are replacement X-ray tubes?

It is also important that you obtain data for samples representative of the 
things you anticipate examining. For example, if you will run many Fe-bearing 
samples, make sure to include one in your test suite (to evaluate 
fluorescence). I suggest that you obtain data for materials with low-angle 
peaks, to evaluate the performance. Many of the newer instruments have poor (my 
assessment) performance at low angles, below perhaps 12 degrees 2-theta in many 
cases, and analysis of things like clay minerals is significantly impacted. I 
assume any manufacturer will provide data for the NIST LaB6, to allow you to 
assess instrument resolution and profile shapes. You might consider obtaining 
data on a very small sample, e.g., a few mg or smaller, and data obtained very 
quickly, e.g., < 1 minute for a reasonable scan range. For all of these, make 
sure you receive the full run conditions, e.g., slit sizes, run parameters 
(time, step size), tube power, etc. All of these should give you a good picture 
of instrument performance.

Regarding sample changers, I believe some other manufacturers supply these with 
their benchtop instruments. Personally, I find these useful primarily in a 
production mode rather than for research. I have one and I seldom use it. An 
advantage of some is that they add the ability to spin samples.

Best of luck with your search, it's always exciting to be able to purchase a 
new instrument. Rather than focus your search on a single instrument, it might 
be better in many ways to look at several different instruments/companies. 
Doing so often yields improved responsiveness and even price. If a company 
knows that you have decided on their instrument, most incentive is lost. I have 
not used the Panalytical benchtop instrument, but I have had good experiences 
with several others.

Regards,
David Bish
________________________________
From: rietveld_l-requ...@ill.fr<mailto:rietveld_l-requ...@ill.fr> 
<rietveld_l-requ...@ill.fr<mailto:rietveld_l-requ...@ill.fr>> on behalf of 
iangie <ian...@126.com<mailto:ian...@126.com>>
Sent: Wednesday, June 14, 2023 5:07 AM
To: Shay Tirosh <stiro...@gmail.com<mailto:stiro...@gmail.com>>
Cc: rietveld_l@ill.fr<mailto:rietveld_l@ill.fr> 
<rietveld_l@ill.fr<mailto:rietveld_l@ill.fr>>
Subject: [External] Re:recommended questions to XRD supplier.

This message was sent from a non-IU address. Please exercise caution when 
clicking links or opening attachments from external sources.

Hi Shay,

Every company will answer that their product is the best.
I suggest you send typical samples you are going to run in the future, to each 
potential company and compare the demo data quality.

In your contract state that "demo data need be resonably reproducible before 
acceptance", so the company won't dare to cheet with high configure data.
This is how I buy machines.


--
Dr. Xiaodong (Tony) Wang




At 2023-06-14 14:50:26, "Shay Tirosh" 
<stiro...@gmail.com<mailto:stiro...@gmail.com>> wrote:

Dear Rietvelds,

We are currently exploring the possibility of acquiring a new benchtop XRD 
instrument, specifically the Benchtop XRD from Malvern-Pananalytical. We have 
scheduled a Zoom meeting with the supplier to discuss this further. Before the 
meeting takes place, I would like to prepare by gathering relevant information.

Could you please provide some recommended questions that we can ask the 
supplier? We aim to gain a comprehensive understanding of the instrument and 
compare its performance to other options available.

Feel free to suggest and comment.

Thank you for your assistance.

Best regards,

Shay

--

Dr. Shay Tirosh

Materials Scientist.

With focusing on Photovoltaics, Electrochemistry, Thin film coatings, and 
nanotechnology.





Mobile: +972-(0)54-8834533

Email: stiro...@gmail.com<mailto:stiro...@gmail.com>

This email and any files transmitted with it are confidential and maybe legally 
privileged. Such message is intended solely for the use of the individual or 
entity to whom they are addressed. Please notify the originator of the message 
if you are not the intended recipient and destroy all copies of the message. 
Please note that any use, dissemination, or reproduction is strictly prohibited 
and may be unlawful.
++++++++++++++++++++++++++++++++++++++++++++++++++++++++++++++
Please do NOT attach files to the whole list <alan.he...@neutronoptics.com>
Send commands to <lists...@ill.fr> eg: HELP as the subject with no body text
The Rietveld_L list archive is on http://www.mail-archive.com/rietveld_l@ill.fr/
++++++++++++++++++++++++++++++++++++++++++++++++++++++++++++++