>> I have found that their (U,V,W) refined parameters seem to be very correlated.
The Cagliotti formula FWHM**2=W+V.tan(theta)+U.tan**2(theta) with parameters U,V,W represents instrument line width FWHM as a function of scattering angle 2.theta for a conventional "neutron powder diffractometer". This function may approximate line broadening in other situations, but really you should use more appropriate parameters for broadening due to particle size and strain, or other types of diffractometer. See the FullProf manual for alternatives. Particle size and strain may be different for different phases, and even for different crystallographic axes. Such effects may not be seen for low resolution neutron diffractometers. Since Cagliotti's formula is a quadratic equation, U,V,W will naturally be strongly correlated if there is no clear minimum in the line width as a function of angle. You can remove this correlation by simply fixing the angle for this minimum i.e. by differentiation and requiring that V=-2U.tan(theta) where 2.theta is approximately the monochromator angle for a conventional neutron diffractometer. ______________________________________________ Dr Alan Hewat, NeutronOptics, Grenoble, FRANCE <alan.he...@neutronoptics.com> +33.476.98.41.68 http://www.NeutronOptics.com/hewat ______________________________________________