Dear Dr. A. Coelho, dear Rietvelders there are four points to be
considered. 1. Precision of the fit (using the
convolution approach). 2. Theoretical basis of the
convolution approach. 3. Calculation time with the method
in my paper. 4. “… some inaccuracies
of the above mentioned paper” 1. There is no doubt that
convolution approach provides good agreement in profile fitting with ray
traycing. 2. It is not obviously with the
rigorous theoretical validation of the convolution method. I mean the following aspects: a) Strictly
speaking, even the flat specimen aberration coupled with the receiving slit
width can not be mathematically described
as convolution. It is not critique,
but just a statement of fact. Well, the convolution can be used, but this is only
approximation (even if it is good for the certain conditions). b) The same is valid
for the axial aberration. Also the treatment of the axial aberration is an
approximation (“More
recently Cheary and Coelho [3,4] have developed a semi-analytical approach to the calculation
and their results have been incorporated into a profile refinement
procedure.” From R. W. Cheary, A. A. Coelho, J. P. Cline. Fundamental Parameters Line Profile Fitting in Laboratory
Diffractometers, J. Res. Natl. Inst. Stand. Technol. 109, 1-25 (2004)) 3. Calculation time. There is only the approach in my
article was represented. It makes possible to develop a number of the (exact)
modifications and approximations (without lost of accuracy). I had not
yet made optimization. I mean not the code optimization “at an assembler code
level” (From R. W. Cheary, A. A. Coelho, J. P. Cline. Fundamental Parameters Line Profile
Fitting in Laboratory Diffractometers, J. Res. Natl. Inst. Stand. Technol. 109, 1-25 (2004))), but only the mathematical algorithms. Partly I have made it. 4. About “some inaccuracies of the above mentioned
paper” 1) The reference was
made on page 310-311 and to quote the text of "Reefman has
shown, by analyzing the effect of transparency, that the convolution model is
not valid in the general case." This of course is
correct as the convolution approach is not valid for axial divergence greater
than probably 10 degrees in both the
primary and secondary beams and linear absorption coefficients less than 10
(1/cm). Self-explanatory. 2) Back to the paper
by about the influence
of coupling specific instrumental functions in FPA, it is also necessary to
tune the fundamental
parameters to allow a best fit for the experimental data (Cheary et al.,
2004)." Coupling effects
were always investigated during the development of FPA by Cheary and Coelho. This led to the
need to partially number cruch the Full Axial Model which considers primary and secondary axial
divergence together. I would like to cite the other paper:
“Fine tuning is
sometimes necessary to accommodate a monochromator or to compensate for the
fact that certain aberrations are
not completely independent [8].” (From R. W. Cheary, A. A. Coelho, J.
P. Cline. Fundamental
Parameters Line Profile Fitting in Laboratory Diffractometers, J. Res. Natl.
Inst. Stand. Technol. 109, 1-25
(2004)) OK, it may be significant whether the fundamental parameter
are fine tuned intentionally to compensate coupling effect, or to compensate for lack of
knowledge. I would like to emphasize that I had not criticized the
convolution approach, but I was only guided by the opinion of the authors of the convolution approach. With best regards, |
- Re: how to find out POLARISATION Factor martijn . fransen
- Re: how to find out POLARISATION Factor Larry Finger
- Re: how to find out POLARISATION Factor Von Dreele, Robert B.
- Re: how to find out POLARISATION Factor AlanCoelho
- RE: how to find out POLARISATION Factor Zuev
- RE: RE: how to find out POLARISATION Factor Whitfield, Pamela
- RE: how to find out POLARISATION Factor AlanCoelho
- RE: how to find out POLARISATION Factor AlanCoelho