Hello Dan,

on 
https://www.bfarm.de/DE/Medizinprodukte/Aufgaben/Risikobewertung-und-Forschung/Wissenschaftliche-Aufarbeitung/EMV-Studie_Herzschrittmacher.html
 you may find a press release from the German Bundesinstitut für Arzneimittel 
und Medizinprodukte (BfArM; Federal Institute for Drugs and Medical Devices) on 
a "Study on the immunity of implantable pacemakers of the Institut für 
Rundfunktechnik GmbH, Munich". In this study, the EMC immunity of 274 pacemaker 
devices was tested according to European standard EN 45502-2-1 (CENELEC - 
CEN/CLC/JTC 16 
(cencenelec.eu)<https://standards.cencenelec.eu/dyn/www/f?p=CENELEC:110:::::FSP_PROJECT,FSP_ORG_ID:20265,2401823&cs=1B2A778F33C4A9FE02D07532CE3EAC191>)
 and American standard ANSI/AAMI PC69 (ANSI/AAMI PC69:2007 (ANSI/AAMI PC 
69:2007) - Active implantable medical devices - Electromagnetic compatibility - 
EMC test protocols for implantable cardiac pacemakers and implantable 
cardioverter defibrillators (Includes 
Erratum)<https://webstore.ansi.org/Standards/AAMI/ANSIAAMIPC692007PC69>). The 
press release is in German only, but here is a translation of the summary:

"Through the use of a standard-compliant measurement setup and the automated 
measurements of PM, which in their model diversity represent about 95% of the 
market for active pacemakers as of 2004, the study offers the most 
comprehensive overview to date of device behavior under the influence of 
electromagnetic interference at different frequencies. In the area of mobile 
radio frequencies, no failures were observed in the study, but often in the 
frequency range below 450 MHz. Should electromagnetic interference occur in 
practice, the device would automatically be switched to safe mode (backup 
mode), where further operation is possible.

In theBfArM's reporting system, there is no accumulation of pacemaker events 
due to EMC influences, and the BfArM has not reported any individual cases of 
high error severity that show a fatal pacemaker failure caused by interference. 
An event that comes into question for this, such as the coincidence of an 
electromagnetic disturbance and the stimulation in the vulnerable phase, is 
theoretically possible, but unlikely and this justifies the very low 
statistical relevance of such exceptional cases from the point of view of the 
BfArM."

With respect to warnings in the instructions, the press release contains the 
following statement:

"Many instructions for use contain not only an indication of the risks of 
electromagnetic interference, but also a list of sources of interference with 
appropriate warnings for the patient. If there is not even a general indication 
of the possibility of interference by external electromagnetic fields, it is 
necessary from the point of view of the BfArM to include a corresponding 
warning in the device description in accordance with Annex 2.4 of AIMD 
90/385/EEC."

Kind regards,

Bernd


Von: Chas Grasso <[email protected]>
Gesendet: Montag, 7. November 2022 16:16
An: [email protected]
Betreff: Re: [PSES] Magnetic fields, human exposure standards, and pacemakers

Hello Dan,

Dont you think that (irrespective of the warning ) the magnet on the DC cord is 
*much* weaker than a magnet used to penetrate through the thickness of the 
human body and then trigger the reed switches? My first thought is that the 
warning is there as a CYA? Perhaps?

On Sun, Nov 6, 2022 at 2:06 PM Dan Roman 
<[email protected]<mailto:[email protected]>>
 wrote:

 This message originated outside of DISH and was sent by: 
[email protected]<mailto:[email protected]>

________________________________
Thank you everyone for the responses.  Contacting the customer support number 
gets you to someone who basically reads back what is on the website and the 
user manual, so it was not particularly helpful when trying to get to the 
specifics.

In a roundabout way through a friend of a friend I managed to arrange a phone 
call with someone from Boston Scientific that had technical engineering 
knowledge and could answer my questions.  The device in question is a 
pacemaker/defibrillator.  It happens to be MRI compatible but I didn't get into 
what that means with respect to magnetic fields, I think it probably has a lot 
to do with not getting ripped out of your chest if you get an MRI.

Anyway, it has two sets of reed switches embedded in it.  One reed switch will 
react to a lower level magnet placed over it and that puts it into a fixed 
pacing mode.  This is used for diagnosing in the doctor's office.  The magnet 
in a fitness tracker is not likely to be strong enough or to get close enough 
to turn on the fixed pacing mode during normal wear, but they still will tell 
you not to wear it to bed in case you contort in such a way when sleeping to 
cause an issue.

The second reed switch reacts only to a much stronger magnetic field and this 
is used to disable the defibrillator if you go in for an operation.  The 
anesthesiologist will apply a strong magnet over the device to disable it for 
the duration of the operation.  Makes sense that as that is easier than 
connecting up a programming device to disable it and then re-enable it and it 
can be easily turned off in an emergency situation with no need for a 
make/model specific programming set.

So the relative is going to keep the fitness tracker and not wear it to bed 
(most of the time).  They lose the sleep tracking but that is not of as much 
interest to them and they don't have it when sleeping every few days to charge 
it overnight anyway.

Not sure if all devices work the same way but sounds like for simplification 
reasons this might be a standardized practice to make surgeon's and other 
health care work easier.

I should have asked about how "MRI compatible" works!  Now I'm curious.

Dan


From: MIKE SHERMAN 
[mailto:[email protected]<mailto:[email protected]>]
Sent: Thursday, November 03, 2022 8:52 PM
To: [email protected]<mailto:[email protected]>
Subject: Re: [PSES] Magnetic fields, human exposure standards, and pacemakers

Dan --

I'd suggest that you have your relative call Support at the pacemaker/defib 
manufacturer. On a slightly different project, I found a lot of on-line 
information about device immunity, and I think we even called Support and got 
additional clarifying information. Go to the source!

Mike Sherman
Sherman PSC LLC
Product Safety and Compliance Consulting
On 11/03/2022 11:32 AM Dan Roman 
<[email protected]<mailto:[email protected]>>
 wrote:


Hello list!

I have a relative with an implanted pacemaker/defibrillator who recently got a 
fitness tracker that has a magnet for attachment/alignment of the charging 
cord.  There are warnings in the user manual advising against people with 
pacemakers using this product because of the magnets, which I assume are fairly 
strong rare earth types.  I looked up the DoC for the tracker and the EM field 
human exposure standards EN 50663 and EN 62479 are listed on the DoC.  My 
question is do these standards and the similar IEC 62233 give a reasonable 
assurance that a pacemaker would not be interfered with or are these standards 
purely concerned with biological effects on the human body only?  I would 
expect the pacemaker to be immune to a certain extent if it is FDA listed and 
complies with IEC 60601-1-2.  I don't know the model number of their device but 
I would expect them all to be similar.  Do they need to send back their fitness 
tracker?  Or is the warning just thrown in because the lawyers wanted it?

Although products I work on need to meet these types of standards, I don't have 
much familiarity with these exposure standards because my products have never 
failed

Regards,

Dan

______________________________________________________
Dan Roman, N.C.E.
Senior Member
IEEE Product Safety Engineering Society
mailto:[email protected]





-
----------------------------------------------------------------

This message is from the IEEE Product Safety Engineering Society emc-pstc 
discussion list. To post a message to the list, send your e-mail to 
<[email protected]<mailto:[email protected]>>

All emc-pstc postings are archived and searchable on the web at: 
http://www.ieee-pses.org/emc-pstc.html<https://eur04.safelinks.protection.outlook.com/?url=http%3A%2F%2Fwww.ieee-pses.org%2Femc-pstc.html&data=05%7C01%7CBernd.Duerrer%40wilo.com%7C4c6ec79bddb14bcad05c08dac0d30c8e%7C39288a38ff19432c80111cd9d0dff445%7C0%7C0%7C638034309922128592%7CUnknown%7CTWFpbGZsb3d8eyJWIjoiMC4wLjAwMDAiLCJQIjoiV2luMzIiLCJBTiI6Ik1haWwiLCJXVCI6Mn0%3D%7C3000%7C%7C%7C&sdata=hVU3gTTi2nyWzbdMKy2kjQjoWoMBSMxmfE2N9SH3dRw%3D&reserved=0>

Website: 
http://www.ieee-pses.org/<https://eur04.safelinks.protection.outlook.com/?url=http%3A%2F%2Fwww.ieee-pses.org%2F&data=05%7C01%7CBernd.Duerrer%40wilo.com%7C4c6ec79bddb14bcad05c08dac0d30c8e%7C39288a38ff19432c80111cd9d0dff445%7C0%7C0%7C638034309922128592%7CUnknown%7CTWFpbGZsb3d8eyJWIjoiMC4wLjAwMDAiLCJQIjoiV2luMzIiLCJBTiI6Ik1haWwiLCJXVCI6Mn0%3D%7C3000%7C%7C%7C&sdata=O46lBBxpklS9B3ZS%2Fz4iGZJwwp9J74gZMJ1Oi%2FtCx6M%3D&reserved=0>
Instructions: http://www.ieee-pses.org/list.html (including how to 
unsubscribe)<https://eur04.safelinks.protection.outlook.com/?url=http%3A%2F%2Fwww.ieee-pses.org%2Flist.html&data=05%7C01%7CBernd.Duerrer%40wilo.com%7C4c6ec79bddb14bcad05c08dac0d30c8e%7C39288a38ff19432c80111cd9d0dff445%7C0%7C0%7C638034309922128592%7CUnknown%7CTWFpbGZsb3d8eyJWIjoiMC4wLjAwMDAiLCJQIjoiV2luMzIiLCJBTiI6Ik1haWwiLCJXVCI6Mn0%3D%7C3000%7C%7C%7C&sdata=nbW99kZUXt7Rlxam4s1MFfdS4ZVnkzO952vMn5m5dKY%3D&reserved=0>
List rules: 
http://www.ieee-pses.org/listrules.html<https://eur04.safelinks.protection.outlook.com/?url=http%3A%2F%2Fwww.ieee-pses.org%2Flistrules.html&data=05%7C01%7CBernd.Duerrer%40wilo.com%7C4c6ec79bddb14bcad05c08dac0d30c8e%7C39288a38ff19432c80111cd9d0dff445%7C0%7C0%7C638034309922128592%7CUnknown%7CTWFpbGZsb3d8eyJWIjoiMC4wLjAwMDAiLCJQIjoiV2luMzIiLCJBTiI6Ik1haWwiLCJXVCI6Mn0%3D%7C3000%7C%7C%7C&sdata=rAdvrA7qw6fGJqSMZN%2BvXBF908OS2RVGzhs0tBZcUxk%3D&reserved=0>

For help, send mail to the list administrators:
Mike Cantwell <[email protected]<mailto:[email protected]>>

For policy questions, send mail to:
Jim Bacher <[email protected]<mailto:[email protected]>>
David Heald <[email protected]<mailto:[email protected]>>

________________________________

To unsubscribe from the EMC-PSTC list, click the following link: 
https://listserv.ieee.org/cgi-bin/wa?SUBED1=EMC-PSTC&A=1<https://eur04.safelinks.protection.outlook.com/?url=https%3A%2F%2Flistserv.ieee.org%2Fcgi-bin%2Fwa%3FSUBED1%3DEMC-PSTC%26A%3D1&data=05%7C01%7CBernd.Duerrer%40wilo.com%7C4c6ec79bddb14bcad05c08dac0d30c8e%7C39288a38ff19432c80111cd9d0dff445%7C0%7C0%7C638034309922128592%7CUnknown%7CTWFpbGZsb3d8eyJWIjoiMC4wLjAwMDAiLCJQIjoiV2luMzIiLCJBTiI6Ik1haWwiLCJXVCI6Mn0%3D%7C3000%7C%7C%7C&sdata=3kBFKZ0nitbDx%2BqEOL473w44TNkJBPW3gnu3n4%2B%2BKzs%3D&reserved=0>
-
----------------------------------------------------------------

This message is from the IEEE Product Safety Engineering Society emc-pstc 
discussion list. To post a message to the list, send your e-mail to 
<[email protected]<mailto:[email protected]>>

All emc-pstc postings are archived and searchable on the web at: 
http://www.ieee-pses.org/emc-pstc.html<https://eur04.safelinks.protection.outlook.com/?url=http%3A%2F%2Fwww.ieee-pses.org%2Femc-pstc.html&data=05%7C01%7CBernd.Duerrer%40wilo.com%7C4c6ec79bddb14bcad05c08dac0d30c8e%7C39288a38ff19432c80111cd9d0dff445%7C0%7C0%7C638034309922128592%7CUnknown%7CTWFpbGZsb3d8eyJWIjoiMC4wLjAwMDAiLCJQIjoiV2luMzIiLCJBTiI6Ik1haWwiLCJXVCI6Mn0%3D%7C3000%7C%7C%7C&sdata=hVU3gTTi2nyWzbdMKy2kjQjoWoMBSMxmfE2N9SH3dRw%3D&reserved=0>

Website: 
http://www.ieee-pses.org/<https://eur04.safelinks.protection.outlook.com/?url=http%3A%2F%2Fwww.ieee-pses.org%2F&data=05%7C01%7CBernd.Duerrer%40wilo.com%7C4c6ec79bddb14bcad05c08dac0d30c8e%7C39288a38ff19432c80111cd9d0dff445%7C0%7C0%7C638034309922128592%7CUnknown%7CTWFpbGZsb3d8eyJWIjoiMC4wLjAwMDAiLCJQIjoiV2luMzIiLCJBTiI6Ik1haWwiLCJXVCI6Mn0%3D%7C3000%7C%7C%7C&sdata=O46lBBxpklS9B3ZS%2Fz4iGZJwwp9J74gZMJ1Oi%2FtCx6M%3D&reserved=0>
Instructions: http://www.ieee-pses.org/list.html (including how to 
unsubscribe)<https://eur04.safelinks.protection.outlook.com/?url=http%3A%2F%2Fwww.ieee-pses.org%2Flist.html&data=05%7C01%7CBernd.Duerrer%40wilo.com%7C4c6ec79bddb14bcad05c08dac0d30c8e%7C39288a38ff19432c80111cd9d0dff445%7C0%7C0%7C638034309922128592%7CUnknown%7CTWFpbGZsb3d8eyJWIjoiMC4wLjAwMDAiLCJQIjoiV2luMzIiLCJBTiI6Ik1haWwiLCJXVCI6Mn0%3D%7C3000%7C%7C%7C&sdata=nbW99kZUXt7Rlxam4s1MFfdS4ZVnkzO952vMn5m5dKY%3D&reserved=0>
List rules: 
http://www.ieee-pses.org/listrules.html<https://eur04.safelinks.protection.outlook.com/?url=http%3A%2F%2Fwww.ieee-pses.org%2Flistrules.html&data=05%7C01%7CBernd.Duerrer%40wilo.com%7C4c6ec79bddb14bcad05c08dac0d30c8e%7C39288a38ff19432c80111cd9d0dff445%7C0%7C0%7C638034309922128592%7CUnknown%7CTWFpbGZsb3d8eyJWIjoiMC4wLjAwMDAiLCJQIjoiV2luMzIiLCJBTiI6Ik1haWwiLCJXVCI6Mn0%3D%7C3000%7C%7C%7C&sdata=rAdvrA7qw6fGJqSMZN%2BvXBF908OS2RVGzhs0tBZcUxk%3D&reserved=0>

For help, send mail to the list administrators:
Mike Cantwell <[email protected]<mailto:[email protected]>>

For policy questions, send mail to:
Jim Bacher <[email protected]<mailto:[email protected]>>
David Heald <[email protected]<mailto:[email protected]>>

________________________________

To unsubscribe from the EMC-PSTC list, click the following link: 
https://listserv.ieee.org/cgi-bin/wa?SUBED1=EMC-PSTC&A=1<https://eur04.safelinks.protection.outlook.com/?url=https%3A%2F%2Flistserv.ieee.org%2Fcgi-bin%2Fwa%3FSUBED1%3DEMC-PSTC%26A%3D1&data=05%7C01%7CBernd.Duerrer%40wilo.com%7C4c6ec79bddb14bcad05c08dac0d30c8e%7C39288a38ff19432c80111cd9d0dff445%7C0%7C0%7C638034309922128592%7CUnknown%7CTWFpbGZsb3d8eyJWIjoiMC4wLjAwMDAiLCJQIjoiV2luMzIiLCJBTiI6Ik1haWwiLCJXVCI6Mn0%3D%7C3000%7C%7C%7C&sdata=3kBFKZ0nitbDx%2BqEOL473w44TNkJBPW3gnu3n4%2B%2BKzs%3D&reserved=0>
-
----------------------------------------------------------------

This message is from the IEEE Product Safety Engineering Society emc-pstc 
discussion list. To post a message to the list, send your e-mail to 
&LT;[email protected]<mailto:[email protected]>&GT;

All emc-pstc postings are archived and searchable on the web at: 
http://www.ieee-pses.org/emc-pstc.html<https://eur04.safelinks.protection.outlook.com/?url=http%3A%2F%2Fwww.ieee-pses.org%2Femc-pstc.html&data=05%7C01%7CBernd.Duerrer%40wilo.com%7C4c6ec79bddb14bcad05c08dac0d30c8e%7C39288a38ff19432c80111cd9d0dff445%7C0%7C0%7C638034309922128592%7CUnknown%7CTWFpbGZsb3d8eyJWIjoiMC4wLjAwMDAiLCJQIjoiV2luMzIiLCJBTiI6Ik1haWwiLCJXVCI6Mn0%3D%7C3000%7C%7C%7C&sdata=hVU3gTTi2nyWzbdMKy2kjQjoWoMBSMxmfE2N9SH3dRw%3D&reserved=0>

Website: 
http://www.ieee-pses.org/<https://eur04.safelinks.protection.outlook.com/?url=http%3A%2F%2Fwww.ieee-pses.org%2F&data=05%7C01%7CBernd.Duerrer%40wilo.com%7C4c6ec79bddb14bcad05c08dac0d30c8e%7C39288a38ff19432c80111cd9d0dff445%7C0%7C0%7C638034309922128592%7CUnknown%7CTWFpbGZsb3d8eyJWIjoiMC4wLjAwMDAiLCJQIjoiV2luMzIiLCJBTiI6Ik1haWwiLCJXVCI6Mn0%3D%7C3000%7C%7C%7C&sdata=O46lBBxpklS9B3ZS%2Fz4iGZJwwp9J74gZMJ1Oi%2FtCx6M%3D&reserved=0>
Instructions: http://www.ieee-pses.org/list.html (including how to 
unsubscribe)<https://eur04.safelinks.protection.outlook.com/?url=http%3A%2F%2Fwww.ieee-pses.org%2Flist.html&data=05%7C01%7CBernd.Duerrer%40wilo.com%7C4c6ec79bddb14bcad05c08dac0d30c8e%7C39288a38ff19432c80111cd9d0dff445%7C0%7C0%7C638034309922128592%7CUnknown%7CTWFpbGZsb3d8eyJWIjoiMC4wLjAwMDAiLCJQIjoiV2luMzIiLCJBTiI6Ik1haWwiLCJXVCI6Mn0%3D%7C3000%7C%7C%7C&sdata=nbW99kZUXt7Rlxam4s1MFfdS4ZVnkzO952vMn5m5dKY%3D&reserved=0>
List rules: 
http://www.ieee-pses.org/listrules.html<https://eur04.safelinks.protection.outlook.com/?url=http%3A%2F%2Fwww.ieee-pses.org%2Flistrules.html&data=05%7C01%7CBernd.Duerrer%40wilo.com%7C4c6ec79bddb14bcad05c08dac0d30c8e%7C39288a38ff19432c80111cd9d0dff445%7C0%7C0%7C638034309922285246%7CUnknown%7CTWFpbGZsb3d8eyJWIjoiMC4wLjAwMDAiLCJQIjoiV2luMzIiLCJBTiI6Ik1haWwiLCJXVCI6Mn0%3D%7C3000%7C%7C%7C&sdata=dfFJ%2FFfiSdjmEZPuIXW1a9w%2BYB2%2FTDmMLzBcKzoSJyI%3D&reserved=0>

For help, send mail to the list administrators:
Mike Cantwell &LT;[email protected]<mailto:[email protected]>&GT;

For policy questions, send mail to:
Jim Bacher &LT;[email protected]<mailto:[email protected]>&GT;
David Heald &LT;[email protected]<mailto:[email protected]>&GT;

________________________________

To unsubscribe from the EMC-PSTC list, click the following link: 
https://listserv.ieee.org/cgi-bin/wa?SUBED1=EMC-PSTC&A=1<https://eur04.safelinks.protection.outlook.com/?url=https%3A%2F%2Flistserv.ieee.org%2Fcgi-bin%2Fwa%3FSUBED1%3DEMC-PSTC%26A%3D1&data=05%7C01%7CBernd.Duerrer%40wilo.com%7C4c6ec79bddb14bcad05c08dac0d30c8e%7C39288a38ff19432c80111cd9d0dff445%7C0%7C0%7C638034309922285246%7CUnknown%7CTWFpbGZsb3d8eyJWIjoiMC4wLjAwMDAiLCJQIjoiV2luMzIiLCJBTiI6Ik1haWwiLCJXVCI6Mn0%3D%7C3000%7C%7C%7C&sdata=KKcmcT68dzVjZMCUUgOwRDiV1BPdz%2Fu2OOc%2BgDNHamo%3D&reserved=0>


--
Charles Grasso
Dish Technologies
 (c) 303-204-2974
(w) 303-706-5467
(h) 303-317-5530
(e ) [email protected]<mailto:[email protected]>
(e2) [email protected]<mailto:[email protected]>
-
----------------------------------------------------------------

This message is from the IEEE Product Safety Engineering Society emc-pstc 
discussion list. To post a message to the list, send your e-mail to 
<[email protected]<mailto:[email protected]>>

All emc-pstc postings are archived and searchable on the web at: 
http://www.ieee-pses.org/emc-pstc.html<https://eur04.safelinks.protection.outlook.com/?url=http%3A%2F%2Fwww.ieee-pses.org%2Femc-pstc.html&data=05%7C01%7CBernd.Duerrer%40wilo.com%7C4c6ec79bddb14bcad05c08dac0d30c8e%7C39288a38ff19432c80111cd9d0dff445%7C0%7C0%7C638034309922285246%7CUnknown%7CTWFpbGZsb3d8eyJWIjoiMC4wLjAwMDAiLCJQIjoiV2luMzIiLCJBTiI6Ik1haWwiLCJXVCI6Mn0%3D%7C3000%7C%7C%7C&sdata=C%2FklHcViPKOjPm31e19U0oKJtDHCm%2F80H%2FgDS3J1NZQ%3D&reserved=0>

Website: 
http://www.ieee-pses.org/<https://eur04.safelinks.protection.outlook.com/?url=http%3A%2F%2Fwww.ieee-pses.org%2F&data=05%7C01%7CBernd.Duerrer%40wilo.com%7C4c6ec79bddb14bcad05c08dac0d30c8e%7C39288a38ff19432c80111cd9d0dff445%7C0%7C0%7C638034309922285246%7CUnknown%7CTWFpbGZsb3d8eyJWIjoiMC4wLjAwMDAiLCJQIjoiV2luMzIiLCJBTiI6Ik1haWwiLCJXVCI6Mn0%3D%7C3000%7C%7C%7C&sdata=aXvP33hq%2BsJKvXf%2FVvq4Zyfv%2B8cCgDohAscCckg9F88%3D&reserved=0>
Instructions: http://www.ieee-pses.org/list.html (including how to 
unsubscribe)<https://eur04.safelinks.protection.outlook.com/?url=http%3A%2F%2Fwww.ieee-pses.org%2Flist.html&data=05%7C01%7CBernd.Duerrer%40wilo.com%7C4c6ec79bddb14bcad05c08dac0d30c8e%7C39288a38ff19432c80111cd9d0dff445%7C0%7C0%7C638034309922285246%7CUnknown%7CTWFpbGZsb3d8eyJWIjoiMC4wLjAwMDAiLCJQIjoiV2luMzIiLCJBTiI6Ik1haWwiLCJXVCI6Mn0%3D%7C3000%7C%7C%7C&sdata=o46atVyRyS3abxhw4lQ3vdHHEXZZE1b3FHev%2F8nvsGI%3D&reserved=0>
List rules: 
http://www.ieee-pses.org/listrules.html<https://eur04.safelinks.protection.outlook.com/?url=http%3A%2F%2Fwww.ieee-pses.org%2Flistrules.html&data=05%7C01%7CBernd.Duerrer%40wilo.com%7C4c6ec79bddb14bcad05c08dac0d30c8e%7C39288a38ff19432c80111cd9d0dff445%7C0%7C0%7C638034309922285246%7CUnknown%7CTWFpbGZsb3d8eyJWIjoiMC4wLjAwMDAiLCJQIjoiV2luMzIiLCJBTiI6Ik1haWwiLCJXVCI6Mn0%3D%7C3000%7C%7C%7C&sdata=dfFJ%2FFfiSdjmEZPuIXW1a9w%2BYB2%2FTDmMLzBcKzoSJyI%3D&reserved=0>

For help, send mail to the list administrators:
Mike Cantwell <[email protected]<mailto:[email protected]>>

For policy questions, send mail to:
Jim Bacher <[email protected]<mailto:[email protected]>>
David Heald <[email protected]<mailto:[email protected]>>

________________________________

To unsubscribe from the EMC-PSTC list, click the following link: 
https://listserv.ieee.org/cgi-bin/wa?SUBED1=EMC-PSTC&A=1

[https://wilo-download.cdn.mediamid.com/footer/150_years_email_footer_350x260px.jpg]
 <https://wilo1872.com/>
Experience Wilo on www.wilo1872.com<https://wilo1872.com/en/150-years/>

________________________________

WILO SE
Wilopark 1, 44263 Dortmund
Amtsgericht Dortmund, HRB 21356
www.wilo.com

Vorstand/Executive Board:
Oliver Hermes (President & CEO), Dr. Patrick Niehr, Georg Weber, Mathias Weyers
Aufsichtsratsvorsitzender/Chairman of the Supervisory Board:
Lars Roßner

Dieses Dokument ist vertraulich zu behandeln. Die Weitergabe sowie 
Vervielfaeltigung, Verwertung und Mitteilung seines Inhalts ist nur mit unserer 
ausdruecklichen Genehmigung gestattet. Alle Rechte vorbehalten, insbesondere 
fuer den Fall der Schutzrechtsanmeldung.
This document has to be treated confidentially. Its contents are not to be 
passed on, duplicated, exploited or disclosed without our expressed permission. 
All rights reserved, especially the right to apply for protective rights.

Datenschutz ist uns wichtig! Daher gehen wir sorgsam mit Ihren Daten um. 
Näheres finden Sie in unseren 
Datenschutzhinweisen<https://wilo.com/de/de/Wilo-Deutschland/Datenschutzhinweise.html>.
 Sollten Sie keinen Kontakt mehr mit uns wünschen, wenden Sie sich bitte an 
[email protected]<mailto:[email protected]>.
Data Protection is important to us! Therefore we handle your data carefully. 
Please find the details in our Note on Data 
Protection<https://wilo.com/de/de/Note-on-data-protection.html>. If you do not 
wish to stay in touch with us anymore, please write to 
[email protected]<mailto:[email protected]>.

-
----------------------------------------------------------------
This message is from the IEEE Product Safety Engineering Society emc-pstc 
discussion list. To post a message to the list, send your e-mail to 
<[email protected]>

All emc-pstc postings are archived and searchable on the web at:
http://www.ieee-pses.org/emc-pstc.html

Website:  http://www.ieee-pses.org/
Instructions:  http://www.ieee-pses.org/list.html (including how to unsubscribe)
List rules: http://www.ieee-pses.org/listrules.html

For help, send mail to the list administrators:
Mike Cantwell <[email protected]>

For policy questions, send mail to:
Jim Bacher:  <[email protected]>
David Heald: <[email protected]>
_________________________________________________
To unsubscribe from the EMC-PSTC list, click the following link: 
https://listserv.ieee.org/cgi-bin/wa?SUBED1=EMC-PSTC&A=1

Reply via email to