On Thu, 16 May 2013 18:25:35 +0100, Frank von Delft 
<frank.vonde...@sgc.ox.ac.uk> wrote:

>Dear Gerard - thanks, very informative!  Two questions:
>
>1.
>Do I understand correctly, that you say XDS will throw together for
>integration counts from many images even if they're spaced widely
>throughout the dataset, i.e. through the various passes?
>

No

>i.e. if I set up my data collection as 5 complete revolutions with low
>beam transmission - will XDS know to combine image 15 and 15+(360deg)
>and 15+(720deg) etc?  By default?  Or do I have tell it to do this
>explicitly, in which case, how?
>

You cannot

>
>2.
>If I have my 5x360 degrees of images, what metric / criterion do I use
>to decide whether to use only data up to 512deg or 839deg or 1469deg?
>

Look at cc1/2 as a function of frame number, for the high-resolution shell. Cut 
before the curve
starts to fall.
You should change some parameter after every 360 degrees, e.g. distance, 
detector shift up/down and left/right, or wavelength to avoid having the same 
systematic errors.
Expose such that you have not less than 1 count/pixel, i.e. do not go overboard 
with low exposure.

HTH,

Kay

>
>Cheers
>Frank
>
>
>
>
>
>On 16/05/2013 18:03, Gerard Bricogne wrote:
>> Dear James,
>>
>>       A week ago I wrote what I thought was a perhaps excessively long and
>> overly dense message in reply to Theresa's initial query, then I thought I
>> should sleep on it before sending it, and got distracted by other things.
>>
>>       I guess you may well have used that whole week composing yours ;-) and
>> reading it just now makes the temptation of sending mine irresistible. I am
>> largely in agreement with you about the need to change mental habits in this
>> field, and hope that the emphasis on various matters in my message below is
>> sufficiently different from yours to make a distinct contribution to this
>> very important discussion. Your analysis of pile-up effects goes well beyond
>> anything I have ever looked at. However, in line with Theresa's initial
>> question, I would say that, while I agree with you that the best strategy
>> for collecting "native data" is no strategy at all, this isn't the case when
>> collecting data for phasing. In that case one needs to go back and consider
>> how to measure accurate differences of intensities, not just accurate
>> intensities on their own. That is another subject, on which I was going to
>> follow up so as to fully answer Theresa's message - but perhaps that should
>> come in another installment!
>>
>>
>>       With best wishes,
>>
>>            Gerard.
>>
>> --
>> On Tue, May 07, 2013 at 12:04:33AM +0100, Theresa Hsu wrote:
>>> Dear crystallographers
>>>
>>> Is there a good source/review/software to obtain tips for good data
>> collection strategy using PILATUS detectors at synchrotron? Do we need to
>> collect sweeps of high and low resolution data separately? For anomalous
>> phasing (MAD), does the order of wavelengths used affect structure solution
>> or limit radiation damage?
>>>
>>> Thank you.
>>>
>>> Theresa
>> --
>>
>> Dear Theresa,
>>
>>       You have had several excellent replies to your question. Perhaps I
>> could venture to add a few more comments, remarks and suggestions, which can
>> be summarised as follows: with a Pilatus, (1) use fine slicing, (2) use
>> strategies combining low exposure with high multiplicity, and (3) use XDS!
>>
>>       As the use of Pilatus detectors has spread widely, it has been rather
>> puzzling to come across so many instances when these detectors are misused,
>> sometimes on the basis of explicit expert advice that is simply misguided. A
>> typical example will be to see images collected on a Pilatus 6M with an
>> image width of 1 degree and an exposure time of 1 second. When you see this,
>> you know that there is some erroneous thinking (or habit) behind it.
>>
>>       When talking to various users who have ended up with such datasets, and
>> with people who advocate this kind of strategy, it seems clear that a number
>> of irrational concerns about fine-slicing and low-exposure+high-multiplicity
>> strategies have tended to override published rational arguments in favour of
>> those strategies: there is a fear that if the images being collected do not
>> show spots discernible by the naked eye to the resolution limit that is
>> being aimed for, the integration software will then somehow not be able to
>> find those spots in order to integrate them, and the final data resolution
>> will be lower than expected. Perhaps this may be of concern in relation with
>> the use of some integration programs, but if you use XDS, which implements a
>> full 3D approach to image integration, this is simply not the case: XDS will
>> collect all the counts belonging to a given reflection, whether those counts
>> are all from a spot on a single 1-degree image exposed for 1 second, or from
>> 10 consecutive images of 0.1 degree width exposed for 0.1 second each, or
>> from 100 images obtained by grouping together the same 10 images as
>> previously collected in 10 successive passes with a 10-fold attenuated beam.
>> The hallmark of the Pilatus detector is to lead to equivalent signal/noise
>> ratios for the last two ways of measuring that reflection, because it is a
>> photon counter and has zero readout noise: therefore the combination
>> Pilatus+XDS is a powerful one.
>>
>>       What is different between these three strategies, however, is the
>> quality of the overall dataset they will produce. There is nothing new in
>> what I am describing below: it is all in the references that Bob Sweet gave
>> you in his reply, or is an obvious consequence of what is found in these
>> references.
>>
>>       In case 1 (1-degree, 1 second - "coarse slicing") you would presumably
>> also be (mis-)advised to use a strategy aiming at collecting a complete
>> dataset in the minimum number of images. These strategies used to make sense
>> in the days of films, of image plates, and even of CCDs because of the image
>> readout noise, but they have no place any longer in the context of Pilatus
>> detectors. First of all, using 1-degree image widths can only degrade the
>> precision with which 2D spots on images are lifted to 3D reciprocal space
>> for indexing, and hence worsen the quality of that indexing and therefore
>> the accuracy with which the spot locations will be predicted (unless you
>> carefully "post-refine") - then the integration step perhaps does need to
>> "hunt" for those spots locally, and needs them to be somewhat visible.
>> Secondly, 1 degree is usually greater than the angular width of a typical
>> reflection: the integration process will therefore pick up more background
>> noise (variance) than it would have done with a smaller image width.
>> Thirdly, by collecting only enough images to reach completeness you will
>> have substantial radiation damage in your late images compared to the early
>> ones (if you don't, it means you have under-exposed your crystal) and will
>> therefore end up with internal inconsistencies in your dataset, as well as
>> perhaps some extra, spurious anisotropy of diffraction limits as a result of
>> having to impose increasingly stringent resolution cut-offs in the later
>> images. This will affect the internal scaling of that dataset and the final
>> quality of the merged data.
>>
>>       In case 2 (0.1 degree, 0.1 second - "fine slicing") you will have a
>> more precise sampling of the 3D shape of each spot, hence more accurate
>> indexing and prediction of spot positions if you use a genuinely 3D
>> integration program like XDS. Thanks to that increased precision, spots can
>> be integrated "blind", even if they are not terribly visible in the images,
>> and the same number of photons will be collected with no penalty in terms of
>> noise level, thanks to the photon-counting noiseless-readout nature of the
>> Pilatus detector. An improvement will be that the finely sampled 3D shape of
>> the spots will be used by XDS to minimise the impact of background variance
>> on the integrated intensities. On the other hand, the differential radiation
>> damage between early and late images will still be the same as in case 1 if
>> you have chosen one of those old-style strategies (and associated beam
>> intensity setting) that aim at just about exhausting the useful lifetime of
>> the crystal by the time you reach completeness.
>>
>>       In case 3 (like case 2, but collecting n times more images with an
>> n-fold attenuated beam once you have collected a few "characterisation
>> images" without that attenuation to carry out the initial indexing) you
>> still have the two advantages of case 2 (the same total number of photons
>> will be picked up by XDS, even if the individual images are now so weak that
>> you can't see anything) but you are spreading the radiation damage so thinly
>> over multiple successive complete datasets that you can choose to later
>> apply a cut-off on image number at the processing stage, when the statistics
>> tell you that diffraction quality has become degraded beyond some critical
>> level. This is much preferable to having to apply different resolution
>> cut-offs to different images towards the end of a barely complete dataset,
>> as in cases 1 and 2. The impact of radiation damage will be quite smoothly
>> and uniformly distributed across the final unique reflections, and your
>> scaling problems (as well as any spurious anisotropy in your diffraction
>> limits) will be minimised.
>>
>>
>>       This is becoming quite a long message: you can see why I included a
>> summary of it at the beginning! Returning to it for a conclusion: Pilatus
>> detectors, fine-slicing with low-exposure and high-multiplicity strategies,
>> and XDS are a unique winning combination. If fears that another integration
>> program may not perform as well as XDS on fine-sliced data make you feel
>> tempted to revert to old-fashioned strategies (case 1) because it supposedly
>> makes no difference: resist the temptation! Switch to those Pilatus-adapted
>> strategies and to XDS, and enjoy the very real difference in the results!
>>
>>
>>       With best wishes,
>>
>>            Gerard
>>
>> and colleagues at Global Phasing.
>>
>> --
>> On Tue, May 07, 2013 at 12:04:33AM +0100, Theresa Hsu wrote:
>>> Dear crystallographers
>>>
>>> Is there a good source/review/software to obtain tips for good data
>> collection strategy using PILATUS detectors at synchrotron? Do we need to
>> collect sweeps of high and low resolution data separately? For anomalous
>> phasing (MAD), does the order of wavelengths used affect structure solution
>> or limit radiation damage?
>>> Thank you.
>>>
>>> Theresa

Reply via email to