>       Hello,
> 
>       thanks for the data you sent I have found that the main change is that
> shading calibration now happens with motor on, leading to a non pure white
> area. This taints the shading data toward blue. I am currently working on a
> fix which should be available soon.
> 
> Regards,
>       Stef

        Hello,

        I have just commited a fix in SANE's git source tree. You may compile 
and 
test it.

Regards,
        Stef