Hey Dave,

The first thing I would suggest is upgrading to NI-VISA 3.1 to make
sure your not running into any old VISA issues.

http://digital.ni.com/softlib.nsf/websearch/8E7D54CA3A1FDECF86256E4D007ED2E0?opendocument&node=132060_US

In addition, here are some helpful links about setting this system up.
http://digital.ni.com/public.nsf/websearch/80C28CC795611E4A86256AB100209521?OpenDocument
http://digital.ni.com/public.nsf/websearch/D7D8E3E1B682022286256D4A00771A52?OpenDocument
http://digital.ni.com/public.nsf/websearch/22F1F2ABC640D5BA86256CD1003BA06D?OpenDocument
http://digital.ni.com/public.nsf/websearch/3DD043326CF5BF6786256C10005151DC?OpenDocument

I'm not sure that GetInstrSession will return for a raw socket.  The
raw socket connections don't contain as much information as the
instrument connections.

The two methods that you have listed act differently.  The first
method you mention will only work on instruments that are VXI-11
compliant and the second method is designed for the instruments that
don't support the TCP/IP instrument protocol.

Sense both of your instruments seem to use VXI-11 I would recommend
using the first method that you've listed.

I hope this helps out.

JoshuaP