Hi Benjamin - The ball-and-stick model doesn't require b-values in the tens of thousands, that would be for DSI type acquisitions.

a.y

On Sat, 11 Jan 2014, Roschinski, Benjamin wrote:

Dear Anastasia,

in the FS Tutorial/Running Freesurfer you say TRACULA uses bedpostX to fit
the ball-and-stick model to the DWI data.
So with TRACULA we use a high angular resolution diffusion imaging model and
you give the recommendation in "Introduction to diffusion MRI" to set the b
value for HARDI ~ 10,000 sec/mm² with a few of 100´s of gradient directions.
The b values in your published articles are much less than what you adviced
and you also used much less
gradient directions.
What does it mean? Why do you suggest a higher b value although you use a
much less one?
Is it because of the upper likelihood of motion effects and the longer scan
time or did I misunderstand the relationship between TRACULA and HARDI?

kind regards
Benjamin

AG Ehrlich
Medizinische Fakultät Carl Gustav Carus
TU Dresden