> This looks like distance between injection point and EUT.
> 
> This diagram is for direct injection through a resistor with no CDN. The 
> resistor forms the ‘C’ or coupling part of the CDN, and the “decoupling 
> device” forms the ‘D’ of decoupling part.

Yes, and it is the horizontal distance, not the cable length.

Clause 7.8 of the standard clearly says:
The injection point shall be located between 0,1 m and 0,3 m from
the geometric projection of the EUT on to the reference ground plane.

Tom


> From: Youngsik Kim <[email protected]> 
> Sent: Tuesday, February 27, 2024 7:57 AM
> To: [email protected]
> Subject: [PSES] Conducted Immunity test for EMC
>  
> Hi All
>  
> I'd like to inquire about the conduction resistance test.
> What is L in the picture below?
> 1. Distance between CDN and EUT
> 2. Length of cable between CDN and EUT

-
----------------------------------------------------------------
This message is from the IEEE Product Safety Engineering Society emc-pstc 
discussion list. To post a message to the list, send your e-mail to 
[email protected]

All emc-pstc postings are archived and searchable on the web at:
https://www.mail-archive.com/[email protected]/

Website:  https://ewh.ieee.org/soc/pses/
Instructions:  https://ewh.ieee.org/soc/pses/list.html (including how to 
unsubscribe)
List rules: https://ewh.ieee.org/soc/pses/listrules.html

For help, send mail to the list administrators:
Mike Sherman at: [email protected]
Rick Linford at: [email protected]

For policy questions, send mail to:
Jim Bacher:  <[email protected]>
_________________________________________________
To unsubscribe from the EMC-PSTC list, click the following link: 
https://listserv.ieee.org/cgi-bin/wa?SUBED1=EMC-PSTC&A=1

Reply via email to