Steve
If IEC61010 is insufficient in some way, You might also consider IEC62368-1(AV, 
ICT) or IEC60601-1 (Medical)

Both have stability requirements associated with equipment that is primarily 
moved on castors for basic stability and installation and additional safeguards 
for equipment and carts/stands that are moved during more routine or normal 
operation. I think they both also have specific requirements for castors that 
might be more applicable to address routine movement and thresholds. At least 
in IEC62368-1, the requirements vary by type of person doing the moving 
(ordinary, skilled or instructed persons).



> On Jul 24, 2023, at 12:15 PM, Douglas Nix 
> <[email protected]> wrote:
> 
> Hi Steve,
> 
> IMO, EN 60204-1 is the wrong standard. You said that the cart is used in a 
> lab setting. You need to use a lab equipment standard, and for the EU EN 
> 61010-1 is the correct choice.
> 
> If not that, then you really need to add a bunch of additional machinery 
> related standards, which I also think is the wrong choice since the cart 
> doesn’t sound like it meets the definition of a machine.
> 
> --
> Doug Nix
> [email protected]
> (519) 729-5704
> 
> "Our greatest weakness lies in giving up. The most certain way to succeed is 
> always to try just one more time." - Thomas Edison
> 
> 
> 
>> On Jul 22, 2023, at 15:51, Steve Brody <[email protected]> wrote:
>> 
>> A client has a product which is a mobile cart intended for a laboratory 
>> setting in which it can be moved to whatever lab station it is needed at.  
>> It is on casters and when wheeled to the working location, leveling feet 
>> which are located in the center of the castor mount, are lowered and locked 
>> into position to keep the cart from moving and to ensure it is level.
>> 
>> The product will be evaluated to EN 60204-1, but there is no reference to 
>> stability in this standard.
>> 
>> First question is should EN 61010-1, 7.4, be used to evaluate stability, and 
>> if not then what?
>>  
>> Assuming yes, the first sentence of 7.4 says; 'Equipment and assemblies of 
>> equipment not secured to the building structure before operation should be 
>> physically stable.'
>>  
>> Second question is does this section apply to the cart only when in position 
>> for use, or also when being moved from location to location? 
>>  
>> Because when in use the cart is supported by [properly rated] feet, then the 
>> tests for castors should no longer apply?.
>> 
>> That is the third question.
>>  
>> Your thoughts and comments are requested and appreciated, as always. 
>>  
>> Thanks,
>>  
>> Steve Brody
>> [email protected] <mailto:[email protected]>
>> C - 603 617 9116
>> This message is from the IEEE Product Safety Engineering Society emc-pstc 
>> discussion list. To post a message to the list, send your e-mail to All 
>> emc-pstc postings are archived and searchable on the web at: 
>> https://www.mail-archive.com/[email protected]/
>> 
>> Website: https://ewh.ieee.org/soc/pses/  <https://ewh.ieee.org/soc/pses/>
>> Instructions: https://ewh.ieee.org/soc/pses/list.html (including how to 
>> unsubscribe) <https://ewh.ieee.org/soc/pses/list.html>
>> List rules: https://ewh.ieee.org/soc/pses/listrules.html
>> 
>> For help, send mail to the list administrators:
>> Mike Sherman at: [email protected] <mailto:[email protected]>
>> Rick Linford at: [email protected] <mailto:[email protected]>
>> For policy questions, send mail to:
>> Jim Bacher at: [email protected] <mailto:[email protected]>
>> To unsubscribe from the EMC-PSTC list, click the following link: 
>> https://listserv.ieee.org/cgi-bin/wa?SUBED1=EMC-PSTC&A=1
>> 
> 
> This message is from the IEEE Product Safety Engineering Society emc-pstc 
> discussion list. To post a message to the list, send your e-mail to All 
> emc-pstc postings are archived and searchable on the web at: 
> https://www.mail-archive.com/[email protected]/
> 
> Website: https://ewh.ieee.org/soc/pses/  <https://ewh.ieee.org/soc/pses/>
> Instructions: https://ewh.ieee.org/soc/pses/list.html (including how to 
> unsubscribe) <https://ewh.ieee.org/soc/pses/list.html>
> List rules: https://ewh.ieee.org/soc/pses/listrules.html
> 
> For help, send mail to the list administrators:
> Mike Sherman at: [email protected] <mailto:[email protected]>
> Rick Linford at: [email protected] <mailto:[email protected]>
> For policy questions, send mail to:
> Jim Bacher at: [email protected] <mailto:[email protected]>
> To unsubscribe from the EMC-PSTC list, click the following link: 
> https://listserv.ieee.org/cgi-bin/wa?SUBED1=EMC-PSTC&A=1
> 


-
----------------------------------------------------------------
This message is from the IEEE Product Safety Engineering Society emc-pstc 
discussion list. To post a message to the list, send your e-mail to 
<[email protected]>

All emc-pstc postings are archived and searchable on the web at:
https://www.mail-archive.com/[email protected]/

Website:  https://ewh.ieee.org/soc/pses/
Instructions:  https://ewh.ieee.org/soc/pses/list.html (including how to 
unsubscribe)
List rules: https://ewh.ieee.org/soc/pses/listrules.html

For help, send mail to the list administrators:
Mike Sherman at: [email protected]
Rick Linford at: [email protected]

For policy questions, send mail to:
Jim Bacher:  <[email protected]>
_________________________________________________
To unsubscribe from the EMC-PSTC list, click the following link: 
https://listserv.ieee.org/cgi-bin/wa?SUBED1=EMC-PSTC&A=1

Reply via email to