Package: dicomnifti
Version: 2.28.2-1~feist
I have some patient data downloaded from Siemens Vision scanner and dicomnifti
do not work with them (nearly at all). When I type "dicomhead filename", I get
empty output. "dicomnifti -v input output" counts correctly the number of
input files ("Reading 1 DICOM files."), but again, generates no output. Because
recently I do not have the standard anonymizing tool with me, I can only send
the data on request with blanked patient name by hand. Parts of DICOM header
not containing IDs (extracted using ImageJ) follows at the end of the report.
My configuration (but I bet it is not important in this case as data from
Siemens Trio and Siemens Symphony get loaded without problem):
Linux 2.6.20-16-generic #2 SMP i686 GNU/Linux
Glibc version: 2.5-0ubuntu14
I have the same problem with GE Elscint Gyrex 2T ... I am not aware weather
should I create a separate bugreport for that as far as the behavior is the
same.
thank you very much for bug fixing
George Keller, MD
Charles University Prague, Czech Republic
So the header follows:
0002,0002 Media Storage SOP Class UID: 1.2.840.10008.5.1.4.1.1.4
0002,0003 Media Storage SOP Inst UID:
1.3.12.2.1107.5.8.1.12345.200511021445220304413
0002,0010 Transfer Syntax UID: 1.2.840.10008.1.2.1
0002,0012 Implementation Class UID: 2.16.840.1.113669.2.931128
0002,0013 Implementation Version Name: VB33D
0008,0005 Specific Character Set: ISO_IR 100
0008,0008 Image Type: ORIGINAL\PRIMARY\OTHER
0008,0016 SOP Class UID: 1.2.840.10008.5.1.4.1.1.4
0008,0018 SOP Instance UID: 1.3.12.2.1107.5.8.1.12345.200511021445220304413
0008,0032 Acquisition Time: 134225.310000
0008,0033 Image Time: 135103.094000
0008,0050 Accession Number:
0008,0060 Modality: MR
0008,0070 Manufacturer: SIEMENS
0008,1010 Station Name: mrc
0008,103E Series Description: MRAB/MPR_4_140_256
0008,1090 Manufacturer's Model Name: MAGNETOM VISION
0009,0010 ---: SPI RELEASE 1
0009,0012 ---: SIEMENS CM VA0 CMS
0009,0013 ---: SIEMENS CM VA0 LAB
0009,1010 ---: SPI VERSION 01.00
0009,1015 ---: 000S00MR002005110214452123
0009,1040 ---: 0
0009,1041 ---: MRUPNONE
0009,1210 ---: STANDARD
0009,1226 ---: 20051102
0009,1227 ---: 133917.000000
0009,1316 ---: 723a0f08
0009,1320 ---: VC
0011,0010 ---: SPI RELEASE 1
0011,0011 ---: SIEMENS CM VA0 CMS
0011,1110 ---: 20051102
0011,1111 ---: 133530.077000
0011,1123 ---: 66
0018,0015 Body Part Examined:
0018,0020 Scanning Sequence: IR
0018,0021 Sequence Variant: OSP
0018,0022 Scan Options:
0018,0023 MR Acquisition Type: 3D
0018,0024 Sequence Name: mpr
0018,0025 Angio Flag: N
0018,0050 Slice Thickness: 001.000000E+00
0018,0080 Repetition Time: 001.140000E+01
0018,0081 Echo Time: 004.400000E+00
0018,0082 Inversion Time: 002.000000E+01
0018,0083 Number of Averages: 1
0018,0084 Imaging Frequency: 006.361803E+01
0018,0085 Imaged Nucleus: 1H
0018,0086 Echo Numbers(s): 1
0018,0087 Magnetic Field Strength: 1.493806
0018,0091 Echo Train Length: 1
0018,0093 Percent Sampling: 100
0018,0094 Percent Phase Field of View: 100
0018,1000 Device Serial Number: 7054
0018,1020 Software Versions(s): VB33G
0018,1200 Date of Last Calibration: 20050803
0018,1201 Time of Last Calibration: 180204.000000
0018,1250 Receiving Coil: CP Head
0018,1310 Acquisition Matrix: 0 512 256 0
0018,1312 Phase Encoding Direction: ROW
0018,1314 Flip Angle: 10
0018,1316 SAR: 0
0018,5100 Patient Position: HFS
0019,0010 ---: SIEMENS CM VA0 CMS
0019,0012 ---: SIEMENS MR VA0 GEN
0019,0014 ---: SIEMENS MR VA0 COAD
0019,0015 ---: SIEMENS CM VA0 ACQU
0019,1010 ---: 50
0019,1020 ---: EXAMNONE
0019,1030 ---: A IRS
0019,1050 ---: 0
0019,1060 ---: 131072
0019,1210 ---: 005.183947E+02
0019,1211 ---: 005.183947E+02
0019,1212 ---: 003.979688E-43
0019,1213 ---: 001.500000E+01
0019,1214 ---: 1
0019,1220 ---: 256
0019,1221 ---: 256
0019,1226 ---: 127
0019,1228 ---: -128
0019,1230 ---: 512
0019,1231 ---: 512
0019,1240 ---: 0
0019,1245 ---: 4
0019,1250 ---: 1
0019,1260 ---: 001.000000E+01
0019,1270 ---: 0
0019,1281 ---: NONE
0019,1283 ---: NONE
0019,1285 ---: NONE
0019,1287 ---: NONE
0019,1290 ---: 0
0019,1294 ---: 000.000000E+00
0019,1298 ---: 000.000000E+00\00.000000E+00\00.000000E+00
0019,1412 ---: 001.493806E+00
0019,1414 ---: 001.000000E+01
0019,1416 ---: -02.780078E+01\-2.842969E+01
0019,1420 ---: 002.200000E+02
0019,1421 ---: 2
0019,1422 ---: 002.500000E+01
0019,1424 ---: 005.545960E+01
0019,1426 ---: 007.656111E+01
0019,1450 ---: 008.252000E+01
0019,1451 ---: 006.479000E+01
0019,1452 ---: 002.803000E+01
0019,1454 ---: -01.300000E+00
0019,1455 ---: 009.926768E+01
0019,1456 ---: 33500
0019,1460 ---: 001.130841E-02
0019,1462 ---: 006.356347E-01
0019,1470 ---: 003.058594E+00
0019,1471 ---: 003.058594E+00
0019,1472 ---: 007.142857E-02
0019,1480 ---: 004.100000E+01\04.400000E+01\04.200000E+01
0019,1482 ---: 002.000000E+02
0019,14A0 ---: 1
0019,14A2 ---: 001.066131E+00
0019,14A5 ---: 001.500000E+00\03.063017E-03\00.000000E+00
0019,14A6 ---: 001.000000E+02\05.372885E-01\04.623279E-01
0019,14D1 ---: 000.000000E+00
0019,14D2 ---: NONE
0019,14D3 ---: 001.000000E+00
0019,14D4 ---: 256
0019,14D5 ---: 0
0019,14D6 ---: 0
0019,14D7 ---: 129
0019,14D8 ---: 180
0019,14D9 ---: 1
0019,14DA ---: MINUSX
0019,1510 ---: /usr/appl/proto/016/c/head/MRAB/MPR_4_140_256.prg
0019,1511 ---: /usr/appl/sequence/mpr_ns_t1_4b130.wkc
0019,1512 ---: SIEMENS
0019,1513 ---: mpr
0020,000D Study Instance UID: 1.3.12.2.1107.5.2.4.7054.20051102133530000
0020,000E Series Instance UID: 1.3.12.2.1107.5.2.4.7054.20051102134223000002
0020,0010 Study ID: 1
0020,0011 Series Number: 2
0020,0012 Acquisition Number: 1
0020,0013 Image Number: 5
0020,0030 ---: -08.104321E+01\-1.275000E+02\07.555162E+01
0020,0032 Image Position (Patient): -81.095478\-128\76.048873
0020,0035 ---:
000.000000E+00\01.000000E+00\-0.000000E+00\01.045285E-01\-0.000000E+00\-9.945219E-01
0020,0037 Image Orientation (Patient): 0\1\-0\0.1045285\-0\-0.9945219
0020,0050 ---: -07.270195E+01
0020,0052 Frame of Reference UID: 1.3.12.2.1107.5.2.4.7054.20051102133917000
0020,0070 ---: PLANAR
0020,0080 Masking Image UID:
0020,1001 ---: 1
0020,1040 Position Reference Indicator:
0020,1041 Slice Location: -07.270195E+01
0020,4000 Image Comments:
0021,0010 ---: SIEMENS MED
0021,0011 ---: SIEMENS CM VA0 CMS
0021,0013 ---: SIEMENS MR VA0 GEN
0021,0023 ---: SIEMENS MR VA0 RAW
0021,1020 ---: 0
0021,1120 ---: 002.560000E+02\02.560000E+02
0021,1122 ---: 001.000000E+00
0021,1124 ---: 000.000000E+00\00.000000E+00
0021,1126 ---: 0
0021,1130 ---: LTOR
0021,1132 ---: HEAD
0021,1160 ---: -06.771583E+01\00.000000E+00\05.124993E+01
0021,1161 ---: 009.945219E-01\00.000000E+00\-1.045285E-01
0021,1163 ---: -07.270195E+01
0021,1165 ---: 1
0021,116A ---: 000.000000E+00\-1.000000E+00\00.000000E+00
0021,116B ---: 001.045285E-01\00.000000E+00\09.945219E-01
0021,1170 ---: A \HR\RF
0021,1171 ---: P \FL\LH
0021,1182 ---: MEA
0021,1322 ---: 0
0021,1324 ---: 0
0021,1330 ---: 140
0021,1331 ---: 140
0021,1334 ---: 140
0021,1336 ---: 2
0021,1339 ---: 001.400000E+02
0021,1340 ---: 1
0021,1341 ---: 1
0021,1342 ---: 1
0021,1343 ---: 1
0021,134F ---: ASCENDING
0021,1356 ---: 001.140000E+01
0021,1370 ---: 1
0021,2300 ---: IMAG
0021,2301 ---: 512
0021,2302 ---: 512
0021,2303 ---: 003.333333E+04
0021,2304 ---: 000.000000E+00\00.000000E+00\00.000000E+00
0021,2305 ---: 000.000000E+00\00.000000E+00\00.000000E+00
0021,2306 ---: 256\ 256\ 64
0021,2307 ---: 256\ 256\ 64
0021,2308 ---: 000.000000E+00\00.000000E+00\00.000000E+00
0021,2309 ---: 000.000000E+00
0021,2310 ---: 004.095000E+03
0021,2331 ---:
002.803000E+01\02.803000E+01\02.803000E+01\02.803000E+01\02.803000E+01\02.803000E+01\02.803000E+01\02.803000E+01\02.803000E+01\02.803000E+01\02.803000E+01\02.803000E+01\02.803000E+01\02.803000E+01\02.803000E+01\02.803000E+01
0028,0002 Samples per Pixel: 1
0028,0004 Photometric Interpretation: MONOCHROME2
0028,0010 Rows: 256
0028,0011 Columns: 256
0028,0030 Pixel Spacing: 001.000000E+00\01.000000E+00
0028,0040 ---: RECT
0028,0100 Bits Allocated: 16
0028,0101 Bits Stored: 12
0028,0102 High Bit: 11
0028,0103 Pixel Representation: 0
0028,1050 Window Center: 242
0028,1051 Window Width: 446
0029,0011 ---: SIEMENS CM VA0 CMS
0029,1110 ---: STD 1
0029,1120 ---: NONE \NONE \NONE
0029,1152 ---: 1
0051,0010 ---: SIEMENS CM VA0 CMS
0051,1010 ---: ***PATIENT ID containing data***
7FE0,0010 Pixel Data: 5824
------------------------
Width: 256 pixels
Height: 256 pixels
Coordinate origin: 0,0
Bits per pixel: 16 (unsigned)
--
To UNSUBSCRIBE, email to [EMAIL PROTECTED]
with a subject of "unsubscribe". Trouble? Contact [EMAIL PROTECTED]