Package: dicomnifti
Version:  2.28.2-1~feist

I have some patient data downloaded from Siemens Vision scanner and dicomnifti 
do not work with them (nearly at all). When I type "dicomhead filename", I get 
empty output.  "dicomnifti -v input output" counts correctly the number of 
input files ("Reading 1 DICOM files."), but again, generates no output. Because 
recently I do not have  the standard anonymizing tool with me, I can only send 
the data on request with blanked patient name by hand. Parts of DICOM header 
not containing IDs (extracted using ImageJ) follows at the end of the report. 

My configuration (but I bet it is not important in this case as data from 
Siemens Trio and Siemens Symphony get loaded without problem): 

Linux 2.6.20-16-generic #2 SMP  i686 GNU/Linux
Glibc version: 2.5-0ubuntu14

I have the same problem with GE Elscint Gyrex 2T ... I am not aware weather 
should I create a separate bugreport for that as far as the behavior is the 
same.
  
  thank you very much for bug fixing
                George Keller, MD
                Charles University Prague, Czech Republic

So the header follows:

0002,0002  Media Storage SOP Class UID: 1.2.840.10008.5.1.4.1.1.4 
0002,0003  Media Storage SOP Inst UID: 
1.3.12.2.1107.5.8.1.12345.200511021445220304413 
0002,0010  Transfer Syntax UID: 1.2.840.10008.1.2.1 
0002,0012  Implementation Class UID: 2.16.840.1.113669.2.931128
0002,0013  Implementation Version Name: VB33D 
0008,0005  Specific Character Set: ISO_IR 100
0008,0008  Image Type: ORIGINAL\PRIMARY\OTHER
0008,0016  SOP Class UID: 1.2.840.10008.5.1.4.1.1.4 
0008,0018  SOP Instance UID: 1.3.12.2.1107.5.8.1.12345.200511021445220304413 
0008,0032  Acquisition Time: 134225.310000 
0008,0033  Image Time: 135103.094000 
0008,0050  Accession Number: 
0008,0060  Modality: MR
0008,0070  Manufacturer: SIEMENS 
0008,1010  Station Name: mrc 
0008,103E  Series Description: MRAB/MPR_4_140_256
0008,1090  Manufacturer's Model Name: MAGNETOM VISION 
0009,0010  ---: SPI RELEASE 1 
0009,0012  ---: SIEMENS CM VA0  CMS 
0009,0013  ---: SIEMENS CM VA0  LAB 
0009,1010  ---: SPI VERSION 01.00 
0009,1015  ---: 000S00MR002005110214452123
0009,1040  ---: 0
0009,1041  ---: MRUPNONE
0009,1210  ---: STANDARD
0009,1226  ---: 20051102
0009,1227  ---: 133917.000000 
0009,1316  ---: 723a0f08
0009,1320  ---: VC
0011,0010  ---: SPI RELEASE 1 
0011,0011  ---: SIEMENS CM VA0  CMS 
0011,1110  ---: 20051102
0011,1111  ---: 133530.077000 
0011,1123  ---:     66
0018,0015  Body Part Examined: 
0018,0020  Scanning Sequence: IR
0018,0021  Sequence Variant: OSP 
0018,0022  Scan Options: 
0018,0023  MR Acquisition Type: 3D
0018,0024  Sequence Name: mpr 
0018,0025  Angio Flag: N 
0018,0050  Slice Thickness: 001.000000E+00
0018,0080  Repetition Time: 001.140000E+01
0018,0081  Echo Time: 004.400000E+00
0018,0082  Inversion Time: 002.000000E+01
0018,0083  Number of Averages:      1
0018,0084  Imaging Frequency: 006.361803E+01
0018,0085  Imaged Nucleus: 1H
0018,0086  Echo Numbers(s):      1
0018,0087  Magnetic Field Strength: 1.493806
0018,0091  Echo Train Length: 1 
0018,0093  Percent Sampling: 100 
0018,0094  Percent Phase Field of View: 100 
0018,1000  Device Serial Number:                       7054
0018,1020  Software Versions(s): VB33G 
0018,1200  Date of Last Calibration: 20050803
0018,1201  Time of Last Calibration: 180204.000000 
0018,1250  Receiving Coil: CP Head 
0018,1310  Acquisition Matrix: 0 512 256 0 
0018,1312  Phase Encoding Direction: ROW 
0018,1314  Flip Angle: 10
0018,1316  SAR: 0 
0018,5100  Patient Position: HFS 
0019,0010  ---: SIEMENS CM VA0  CMS 
0019,0012  ---: SIEMENS MR VA0  GEN 
0019,0014  ---: SIEMENS MR VA0  COAD
0019,0015  ---: SIEMENS CM VA0  ACQU
0019,1010  ---:     50
0019,1020  ---: EXAMNONE
0019,1030  ---: A   IRS 
0019,1050  ---:      0
0019,1060  ---: 131072
0019,1210  ---: 005.183947E+02
0019,1211  ---: 005.183947E+02
0019,1212  ---: 003.979688E-43
0019,1213  ---: 001.500000E+01
0019,1214  ---:      1
0019,1220  ---:    256
0019,1221  ---:    256
0019,1226  ---:    127
0019,1228  ---:   -128
0019,1230  ---:    512
0019,1231  ---:    512
0019,1240  ---:      0
0019,1245  ---:      4
0019,1250  ---:      1
0019,1260  ---: 001.000000E+01
0019,1270  ---:      0
0019,1281  ---: NONE    
0019,1283  ---: NONE    
0019,1285  ---: NONE    
0019,1287  ---: NONE  
0019,1290  ---:      0
0019,1294  ---: 000.000000E+00
0019,1298  ---: 000.000000E+00\00.000000E+00\00.000000E+00
0019,1412  ---: 001.493806E+00
0019,1414  ---: 001.000000E+01
0019,1416  ---: -02.780078E+01\-2.842969E+01
0019,1420  ---: 002.200000E+02
0019,1421  ---:      2
0019,1422  ---: 002.500000E+01
0019,1424  ---: 005.545960E+01
0019,1426  ---: 007.656111E+01
0019,1450  ---: 008.252000E+01
0019,1451  ---: 006.479000E+01
0019,1452  ---: 002.803000E+01
0019,1454  ---: -01.300000E+00
0019,1455  ---: 009.926768E+01
0019,1456  ---:  33500
0019,1460  ---: 001.130841E-02
0019,1462  ---: 006.356347E-01
0019,1470  ---: 003.058594E+00
0019,1471  ---: 003.058594E+00
0019,1472  ---: 007.142857E-02
0019,1480  ---: 004.100000E+01\04.400000E+01\04.200000E+01
0019,1482  ---: 002.000000E+02
0019,14A0  ---:      1
0019,14A2  ---: 001.066131E+00
0019,14A5  ---: 001.500000E+00\03.063017E-03\00.000000E+00
0019,14A6  ---: 001.000000E+02\05.372885E-01\04.623279E-01
0019,14D1  ---: 000.000000E+00
0019,14D2  ---: NONE
0019,14D3  ---: 001.000000E+00
0019,14D4  ---:    256
0019,14D5  ---:      0
0019,14D6  ---:      0
0019,14D7  ---:    129
0019,14D8  ---:    180
0019,14D9  ---:      1
0019,14DA  ---: MINUSX
0019,1510  ---: /usr/appl/proto/016/c/head/MRAB/MPR_4_140_256.prg 
0019,1511  ---: /usr/appl/sequence/mpr_ns_t1_4b130.wkc
0019,1512  ---: SIEMENS 
0019,1513  ---: mpr 
0020,000D  Study Instance UID: 1.3.12.2.1107.5.2.4.7054.20051102133530000
0020,000E  Series Instance UID: 1.3.12.2.1107.5.2.4.7054.20051102134223000002 
0020,0010  Study ID:      1
0020,0011  Series Number:      2
0020,0012  Acquisition Number:      1
0020,0013  Image Number:      5
0020,0030  ---: -08.104321E+01\-1.275000E+02\07.555162E+01
0020,0032  Image Position (Patient): -81.095478\-128\76.048873 
0020,0035  ---: 
000.000000E+00\01.000000E+00\-0.000000E+00\01.045285E-01\-0.000000E+00\-9.945219E-01
0020,0037  Image Orientation (Patient): 0\1\-0\0.1045285\-0\-0.9945219
0020,0050  ---: -07.270195E+01
0020,0052  Frame of Reference UID: 1.3.12.2.1107.5.2.4.7054.20051102133917000
0020,0070  ---: PLANAR
0020,0080  Masking Image UID: 
0020,1001  ---:      1
0020,1040  Position Reference Indicator: 
0020,1041  Slice Location: -07.270195E+01
0020,4000  Image Comments:                                                     
0021,0010  ---: SIEMENS MED 
0021,0011  ---: SIEMENS CM VA0  CMS 
0021,0013  ---: SIEMENS MR VA0  GEN 
0021,0023  ---: SIEMENS MR VA0  RAW 
0021,1020  ---: 0
0021,1120  ---: 002.560000E+02\02.560000E+02
0021,1122  ---: 001.000000E+00
0021,1124  ---: 000.000000E+00\00.000000E+00
0021,1126  ---:      0
0021,1130  ---: LTOR
0021,1132  ---: HEAD
0021,1160  ---: -06.771583E+01\00.000000E+00\05.124993E+01
0021,1161  ---: 009.945219E-01\00.000000E+00\-1.045285E-01
0021,1163  ---: -07.270195E+01
0021,1165  ---: 1
0021,116A  ---: 000.000000E+00\-1.000000E+00\00.000000E+00
0021,116B  ---: 001.045285E-01\00.000000E+00\09.945219E-01
0021,1170  ---: A \HR\RF
0021,1171  ---: P \FL\LH
0021,1182  ---: MEA 
0021,1322  ---:      0
0021,1324  ---:      0
0021,1330  ---:    140
0021,1331  ---:    140
0021,1334  ---:    140
0021,1336  ---:      2
0021,1339  ---: 001.400000E+02
0021,1340  ---:      1
0021,1341  ---:      1
0021,1342  ---:      1
0021,1343  ---:      1
0021,134F  ---: ASCENDING   
0021,1356  ---: 001.140000E+01
0021,1370  ---:      1
0021,2300  ---: IMAG
0021,2301  ---:    512
0021,2302  ---:    512
0021,2303  ---: 003.333333E+04
0021,2304  ---: 000.000000E+00\00.000000E+00\00.000000E+00
0021,2305  ---: 000.000000E+00\00.000000E+00\00.000000E+00
0021,2306  ---:    256\  256\   64
0021,2307  ---:    256\  256\   64
0021,2308  ---: 000.000000E+00\00.000000E+00\00.000000E+00
0021,2309  ---: 000.000000E+00
0021,2310  ---: 004.095000E+03
0021,2331  ---: 
002.803000E+01\02.803000E+01\02.803000E+01\02.803000E+01\02.803000E+01\02.803000E+01\02.803000E+01\02.803000E+01\02.803000E+01\02.803000E+01\02.803000E+01\02.803000E+01\02.803000E+01\02.803000E+01\02.803000E+01\02.803000E+01
0028,0002  Samples per Pixel: 1
0028,0004  Photometric Interpretation: MONOCHROME2 
0028,0010  Rows: 256
0028,0011  Columns: 256
0028,0030  Pixel Spacing: 001.000000E+00\01.000000E+00
0028,0040  ---: RECT
0028,0100  Bits Allocated: 16
0028,0101  Bits Stored: 12
0028,0102  High Bit: 11
0028,0103  Pixel Representation: 0
0028,1050  Window Center:   242
0028,1051  Window Width:   446
0029,0011  ---: SIEMENS CM VA0  CMS 
0029,1110  ---: STD 1       
0029,1120  ---: NONE      \NONE       \NONE         
0029,1152  ---:      1
0051,0010  ---: SIEMENS CM VA0  CMS 
0051,1010  ---: ***PATIENT ID containing data***
7FE0,0010  Pixel Data: 5824
------------------------
Width:  256 pixels
Height:  256 pixels
Coordinate origin:  0,0
Bits per pixel: 16 (unsigned)




-- 