Thank you for the tips Brent.

I will try what you suggest. There are actually multiple lattices. And I
actually only processed the dominant one in XDS.
On May 8, 2014 2:26 PM, "Segelke, Brent W." <segel...@llnl.gov> wrote:

>  Completeness looks pretty good and you have a good enough resolution
> that you should be able to get this eventually.
>
>
>
> A couple of things to look at to start. First, suspect your data
> processing. As others have already said, double check your space group. You
> could even do an initial sanity check by expanding your MR solution into P1
> and refining with strong NCS in P1 reduced data. You will automatically get
> a lower R because you are introducing more parameters, but if you use
> strong NCS and both R and Rfree are better behaved, then something is going
> on with the merging (e.g., wrong space group).
>
>
>
> Look in your log and see if there are an unusual number of reflection or
> frames thrown out, it should be only a few percentage. If you have more
> than 10-15% rejected, something probably went awry. Tweak your error model
> and/or summing box, etc.
>
>
>
> You could also try refining to a less ambitious resolution to start,
> assuming anisotropy will affect the higher resolution to a greater degree.
>
>
>
> Hope this helps.
>
>
>
> Brent
>
>
>
> *From:* yarrowmadr...@gmail.com [mailto:yarrowmadr...@gmail.com] *On
> Behalf Of *Yarrow Madrona
> *Sent:* Thursday, May 08, 2014 1:50 PM
> *To:* Segelke, Brent W.; CCP4BB@JISCMAIL.AC.UK
> *Subject:* Re: [ccp4bb] stalled refinement after MR solution
>
>
>
> Hi Brent,
>
>
>
> I forgot to mention the resolution and other statistics. Here they are
> (XDS -unmerged data P2 below). Overall completeness is 93.4(85.2)% for 2.2A.
>
> I do have some anisotropy in the b-direction. I have tried running phaser
> with data scaled by the UCLA anisotropy server and obtained the same
> results.
>
>
>
> -Yarrow
>
>
>
> SUBSET OF INTENSITY DATA WITH SIGNAL/NOISE >= -3.0 AS FUNCTION OF
> RESOLUTION
>
>  RESOLUTION     NUMBER OF REFLECTIONS    COMPLETENESS R-FACTOR  R-FACTOR
> COMPARED I/SIGMA   R-meas  CC(1/2)  Anomal  SigAno   Nano LIMIT
> OBSERVED  UNIQUE  POSSIBLE     OF DATA   observed  expected Corr
>
>
>
>      6.40        4788    1487      1718       86.6%       2.7%      2.9%
>   4675   33.78      3.2%    99.8*    20*   0.895     873
>
>      4.60        8644    2667      2852       93.5%       3.0%      3.1%
>   8511   30.77      3.6%    99.8*     9    0.856    1461
>
>      3.78       10659    3301      3598       91.7%       3.0%      3.1%
>  10559   31.05      3.6%    99.8*     0    0.789    1747
>
>      3.28       12398    3874      4261       90.9%       3.7%      3.7%
>  12319   25.46      4.4%    99.7*    -4    0.779    1981
>
>      2.94       14970    4587      4766       96.2%       4.8%      4.9%
>  14886   20.15      5.8%    99.7*    -2    0.779    2476
>
>      2.69       16718    5129      5292       96.9%       6.7%      6.9%
>  16638   15.44      8.1%    99.4*    -2    0.767    2681
>
>      2.49       17842    5536      5722       96.7%       9.3%      9.4%
>  17734   11.89     11.1%    98.9*    -2    0.755    2779
>
>      2.33       19743    6017      6133       98.1%      12.0%     12.6%
>  19623    9.27     14.3%    98.4*    -4    0.737    3132
>
>      2.20       16777    5566      6533       85.2%      17.4%     16.9%
>  16420    6.77     21.2%    96.8*    -8    0.749    2334
>
>     total      122539   38164     40875       93.4%       5.3%      5.4%
>   121365   17.41      6.4%    99.7*    -2    0.776   19464
>
>
>
>
>
>  NUMBER OF REFLECTIONS IN SELECTED SUBSET OF IMAGES  134095
>
>  NUMBER OF REJECTED MISFITS                           11536
>
>  NUMBER OF SYSTEMATIC ABSENT REFLECTIONS                  0
>
>  NUMBER OF ACCEPTED OBSERVATIONS                     122559
>
>  NUMBER OF UNIQUE ACCEPTED REFLECTIONS                38170
>
>
>
>
>
> On Thu, May 8, 2014 at 10:48 AM, Segelke, Brent W. <segel...@llnl.gov>
> wrote:
>
> You didn’t report your resolution and completeness. Also, is there
> anisotropy in your data?
>
>
>
> Brent
>
>
>
> *From:* CCP4 bulletin board [mailto:CCP4BB@JISCMAIL.AC.UK] *On Behalf Of 
> *Yarrow
> Madrona
> *Sent:* Thursday, May 08, 2014 10:12 AM
> *To:* CCP4BB@JISCMAIL.AC.UK
> *Subject:* [ccp4bb] stalled refinement after MR solution
>
>
>
> Hello CCP4 community,
>
>
>
> I am stumped and would love some help. I have a molecular replacement
> solution that has Rfree stuck around 40% while Rwork is aorund 30%. The
> model is actually the same enzyme with a similar inhibitor bound. Relevant
> information is below.
>
>
>
> -Yarrow
>
>
>
> I have solved a structure in a P21 spacegroup:
>
>
>
> 51.53 88.91 89.65, beta = 97.1.
>
>
>
> Processing stats (XDS) are very good with low Rmerge (~5% overall) and
> good completeness.
>
>
>
> I don't think twinning is an option with these unit cell dimensions. My
> data was highly aniosotropic. I ran the data through the UCLA anisotropic
> server to scale in the B- direction (
> http://services.mbi.ucla.edu/anisoscale/)
>
>
>
> I get a small (a little over 5) patterson peak suggesting there is not
> much t-NCS to worry about. However, the output structure does have 2 fold
> symmetry (see below) and as Dale Tronrud pointed out, there is always tNCS
> in a P21 space group with two monomers related by a 2-fold axis.
>
> I calculated the translation to be unit cell fractions of 0.36 0.35, 0.32.
>
>
>
> rota_matrix   -0.9860   -0.1636   -0.0309
>
> rota_matrix   -0.1659    0.9511    0.2605
>
> rota_matrix   -0.0132    0.2620   -0.9650
>
> tran_orth      34.3310  -24.0033  107.0457
>
>
>
> center_orth   15.7607    7.2426   77.7512
>
>
>
> *Phaser stats:*
>
> *  SOLU SET  RFZ=20.3 TFZ=19.5 PAK=0 LLG=1314 RF++ TFZ=63.8 PAK=0 LLG=4745
> LLG=4947*
>
>
>
>
>
>
>

Reply via email to