Two cents - 

 

A good deal of caution must be exercised when working with composite
particles such as a protein-DNA complex in SAXS because of the contrast
problem.  Simply, protein and DNA scatter differently in x-rays, with a bias
towards the DNA component.  As a result, experimental Rgs could be slightly
deflated versus what their true values would be at infinite contrast.  Mass
estimation by I(0) analysis with a protein standard of known mass and
concentration is not really valid because the contrast terms are different.
Because the particle is heterogeneous in composition and distribution, shape
reconstruction from SAXS alone, which assumes homogeneity, can also be
misleading (although in practice it is still reasonably instructive).  It is
for these reasons that SANS and the contrast variation approach can be
extremely useful.  

 

With those caveats, the strategy you describe - comparison of experimental
and theoretical profiles from an experimental structure using CRYSOL or FoxS
is definitely the best way to go in the case of a protein-DNA complex with
SAXS alone.  Showing comparisons of the experimental with the calculated
should make the point.  Test other possible models inferred from lattice
packing to further your point (if applicable).

 

Regarding populations of monomer and dimer - 

 

.         it is generally good to constrain your interpretation of
scattering data with other orthogonal solution measures which demonstrates
the homogeneity of your complex in comparable experimental conditions, such
as sedimentation velocity or gel filtration.  

 

.         Have some determination of affinity of the complex in the same
solution conditions (including temperature!).  This will allow you to argue
that your sample concentrations are well in excess of any monomer-dimer
association behavior (eg, mixtures!).  Scattering of mixtures can undermine
your ability to accurately assess the structural properties of your complex.

 

.         Collect a concentration series and extrapolate to infinite
dilution, if possible, to ensure elimination of the S(q) term from your
data.  Interparticle interactions can be an issue with complexes containing
DNA if the buffers aren't quite right. (I've seen this a lot)

 

Lastly, remember that the scattering profile represents the solution average
of the particle, not just a single snapshot.  Some discrepancies like those
you note should be expected.  

 

Hope that helps,

 

Kushol

 

Kushol Gupta, Ph.D.

Research Associate - Van Duyne Laboratory 

HHMI / Perelman School of Medicine

University of Pennsylvania 

kgu...@mail.med.upenn.edu

215-573-7260 / 267-259-0082

 

 

-----Original Message-----
From: CCP4 bulletin board [mailto:CCP4BB@JISCMAIL.AC.UK] On Behalf Of Xun Lu
Sent: Saturday, June 16, 2012 2:29 PM
To: CCP4BB@JISCMAIL.AC.UK
Subject: [ccp4bb] Do my SAXS data agree with the crystal structure?

 

Dear all, 

 

   

       I have solved a protein-DNA structure, and I also did SAXS to get
some ideas of the solution structure.  The SAXS data were good, no
aggregation at all three tested concentrations.  I tried to use Crysol to
see if my crystal structure fits the SAXS. The fitting to the scattering
profile seems good to me and the Chi2 is 1~1.4.   Then I wanted to see how
the P(r) looked like (wanted to make a figure for my paper:).  I calculated
the theoretical scattering profile of the crystal structure from an online
server (FOXS).  I then run GNOM to make P(r).  To my surprise, this
theoretical P(r) looks a little different from the P(r) of SAXS data.
There's a very small bump that was peaked at 70A (Dmax is 108A, which seems
reasonable from the crystal structure).   The major peak was at 25A.  As
some people said, P(r) is indeed quite sensitive to subtle differences.  

 

        The protein is a dimer in the crystal, although it can also bind DNA
as a monomer (much  more loosely).  The estimated MW from SAXS indicates
it's a dimer in solution as well.   It seems that I got the information I
wanted from the SAXS experiment, but maybe not.  Due to the low resolution
of SAXS, maybe I can only say that the majority is a dimer??  Would it be
possible to see the monomer if there's only 10% of them in the solution?
How to interpret the discrepancy between the P(r) from crystal and the P(r)
from SAXS?

 

 

Any comments are welcome!

 

 

 

 