> I'm intrigued: how come this apparently excellent idea has not become > standard best practice in the 14 years since it was published?
It would seem because too few people know about it, and it is not implemented in any software in the usual pipeline. Maybe it could be? Perhaps the way to do it would be always to integrate to ridiculously-high resolution, give that to Refmac, and starting from lower resolution, to iterate to higher resolution according the most recent sigma a calculation, and cutoff according to some reasonable sigma a value? JPK > > phx > > > > On 30/01/2012 09:40, Randy Read wrote: > > Hi, > > Here are a couple of links on the idea of judging resolution by a type of > cross-validation with data not used in refinement: > > Ling et al, 1998: http://pubs.acs.org/doi/full/10.1021/bi971806n > Brunger et al, > 2008: http://journals.iucr.org/d/issues/2009/02/00/ba5131/index.html > (cites earlier relevant papers from Brunger's group) > > Best wishes, > > Randy Read > > On 30 Jan 2012, at 07:09, arka chakraborty wrote: > > Hi all, > > In the context of the above going discussion can anybody post links for a > few relevant articles? > > Thanks in advance, > > ARKO > > On Mon, Jan 30, 2012 at 3:05 AM, Randy Read <rj...@cam.ac.uk> wrote: >> >> Just one thing to add to that very detailed response from Ian. >> >> We've tended to use a slightly different approach to determining a >> sensible resolution cutoff, where we judge whether there's useful >> information in the highest resolution data by whether it agrees with >> calculated structure factors computed from a model that hasn't been refined >> against those data. We first did this with the complex of the Shiga-like >> toxin B-subunit pentamer with the Gb3 trisaccharide (Ling et al, 1998). >> From memory, the point where the average I/sig(I) drops below 2 was around >> 3.3A. However, we had a good molecular replacement model to solve this >> structure and, after just carrying out rigid-body refinement, we computed a >> SigmaA plot using data to the edge of the detector (somewhere around 2.7A, >> again from memory). The SigmaA plot dropped off smoothly to 2.8A >> resolution, with values well above zero (indicating significantly better >> than random agreement), then dropped suddenly. So we chose 2.8A as the >> cutoff. Because there were four pentamers in the asymmetric unit, we could >> then use 20-fold NCS averaging, which gave a fantastic map. In this case, >> the averaging certainly helped to pull out something very useful from a very >> weak signal, because the maps weren't nearly as clear at lower resolution. >> >> Since then, a number of other people have applied similar tests. Notably, >> Axel Brunger has done some careful analysis to show that it can indeed be >> useful to take data beyond the conventional limits. >> >> When you don't have a great MR model, you can do something similar by >> limiting the resolution for the initial refinement and rebuilding, then >> assessing whether there's useful information at higher resolution by using >> the improved model (which hasn't seen the higher resolution data) to compute >> Fcalcs. By the way, it's not necessary to use a SigmaA plot -- the >> correlation between Fo and Fc probably works just as well. Note that, when >> the model has been refined against the lower resolution data, you'll expect >> a drop in correlation at the resolution cutoff you used for refinement, >> unless you only use the cross-validation data for the resolution range used >> in refinement. >> >> ----- >> Randy J. Read >> Department of Haematology, University of Cambridge >> Cambridge Institute for Medical Research Tel: +44 1223 336500 >> Wellcome Trust/MRC Building Fax: +44 1223 336827 >> Hills Road >> E-mail: rj...@cam.ac.uk >> Cambridge CB2 0XY, U.K. >> www-structmed.cimr.cam.ac.uk >> >> On 29 Jan 2012, at 17:25, Ian Tickle wrote: >> >> > Jacob, here's my (personal) take on this: >> > >> > The data quality metrics that everyone uses clearly fall into 2 >> > classes: 'consistency' metrics, i.e. Rmerge/meas/pim and CC(1/2) which >> > measure how well redundant observations agree, and signal/noise ratio >> > metrics, i.e. mean(I/sigma) and completeness, which relate to the >> > information content of the data. >> > >> > IMO the basic problem with all the consistency metrics is that they >> > are not measuring the quantity that is relevant to refinement and >> > electron density maps, namely the information content of the data, at >> > least not in a direct and meaningful way. This is because there are 2 >> > contributors to any consistency metric: the systematic errors (e.g. >> > differences in illuminated volume and absorption) and the random >> > errors (from counting statistics, detector noise etc.). If the data >> > are collected with sufficient redundancy the systematic errors should >> > hopefully largely cancel, and therefore only the random errors will >> > determine the information content. Therefore the systematic error >> > component of the consistency measure (which I suspect is the biggest >> > component, at least for the strong reflections) is not relevant to >> > measuring the information content. If the consistency measure only >> > took into account the random error component (which it can't), then it >> > would be essentially be a measure of information content, if only >> > indirectly (but then why not simply use a direct measure such as the >> > signal/noise ratio?). >> > >> > There are clearly at least 2 distinct problems with Rmerge, first it's >> > including systematic errors in its measure of consistency, second it's >> > not invariant with respect to the redundancy (and third it's useless >> > as a statistic anyway because you can't do any significance tests on >> > it!). The redundancy problem is fixed to some extent with Rpim etc, >> > but that still leaves the other problems. It's not clear to me that >> > CC(1/2) is any better in this respect, since (as far as I understand >> > how it's implemented), one cannot be sure that the systematic errors >> > will cancel for each half-dataset Imean, so it's still likely to >> > contain a large contribution from the irrelevant systematic error >> > component and so mislead in respect of the real data quality exactly >> > in the same way that Rmerge/meas/pim do. One may as well use the >> > Rmerge between the half dataset Imeans, since there would be no >> > redundancy effect (i.e. the redundancy would be 2 for all included >> > reflections). >> > >> > I did some significance tests on CC(1/2) and I got silly results, for >> > example it says that the significance level for the CC is ~ 0.1, but >> > this corresponded to a huge Rmerge (200%) and a tiny mean(I/sigma) >> > (0.4). It seems that (without any basis in statistics whatsoever) the >> > rule-of-thumb CC > 0.5 is what is generally used, but I would be >> > worried that the statistics are so far divorced from the reality - it >> > suggests that something is seriously wrong with the assumptions! >> > >> > Having said all that, the mean(I/sigma) metric, which on the face of >> > it is much more closely related to the information content and >> > therefore should be a more relevant metric than Rmerge/meas/pim & >> > CC(1/2), is not without its own problems (which probably explains the >> > continuing popularity of the other metrics!). First and most obvious, >> > it's a hostage to the estimate of sigma(I) used. I've never been >> > happy with inflating the counting sigmas to include effects of >> > systematic error based on the consistency of redundant measurements, >> > since as I indicated above if the data are collected redundantly in >> > such a way that the systematic errors largely cancel, it implies that >> > the systematic errors should not be included in the estimate of sigma. >> > The fact that then the sigma(I)'s would generally be smaller (at >> > least for the large I's), so the sample variances would be much larger >> > than the counting variances, is irrelevant, because the former >> > includes the systematic errors. Also the I/sigma cut-off used would >> > probably not need to be changed since it affects only the weakest >> > reflections which are largely unaffected by the systematic error >> > correction. >> > >> > The second problem with mean(I/sigma) is also obvious: i.e. it's a >> > mean, and as such it's rather insensitive to the actual distribution >> > of I/sigma(I). For example if a shell contained a few highly >> > significant intensities these could be overwhelmed by a large number >> > of weak data and give an insignificant mean(I/sigma). It seems to me >> > that one should be considering the significance of individual >> > reflections, not the shell averages. Also the average will depend on >> > the width of the resolution bin, so one will get the strange effect >> > that the apparent resolution will depend on how one bins at the data! >> > The assumption being made in taking the bin average is that I/sigma(I) >> > falls off smoothly with d* but that's unlikely to be the reality. >> > >> > It seems to me that a chi-square statistic which takes into account >> > the actual distribution of I/sigma(I) would be a better bet than the >> > bin average, though it's not entirely clear how one would formulate >> > such a metric. One would have to consider subsets of the data as a >> > whole sorted by increasing d* (i.e. not in resolution bins to avoid >> > the 'bin averaging effect' described above), and apply the resolution >> > cut-off where the chi-square statistic has maximum probability. This >> > would automatically take care of incompleteness effects since all >> > unmeasured reflections would be included with I/sigma = 0 just for the >> > purposes of working out the cut-off point. I've skipped the details >> > of implementation and I've no idea how it would work in practice! >> > >> > An obvious question is: do we really need to worry about the exact >> > cut-off anyway, won't our sophisticated maximum likelihood refinement >> > programs handle the weak data correctly? Note that in theory weak >> > intensities should be handled correctly, however the problem may >> > instead lie with incorrectly estimated sigmas: these are obviously >> > much more of an issue for any software which depends critically on >> > accurate estimates of uncertainty! I did some tests where I refined >> > data for a known protein-ligand complex using the original apo model, >> > and looked at the difference density for the ligand, using data cut at >> > 2.5, 2 and 1.5 Ang where the standard metrics strongly suggested there >> > was only data to 2.5 Ang. >> > >> > I have to say that the differences were tiny, well below what I would >> > deem significant (i.e. not only the map resolutions but all the map >> > details were essentially the same), and certainly I would question >> > whether it was worth all the soul-searching on this topic over the >> > years! So it seems that the refinement programs do indeed handle weak >> > data correctly, but I guess this should hardly come as a surprise (but >> > well done to the software developers anyway!). This was actually >> > using Buster: Refmac seems to have more of a problem with scaling & >> > TLS if you include a load of high resolution junk data. However, >> > before anyone acts on this information I would _very_ strongly advise >> > them to repeat the experiment and verify the results for themselves! >> > The bottom line may be that the actual cut-off used only matters for >> > the purpose of quoting the true resolution of the map, but it doesn't >> > significantly affect the appearance of the map itself. >> > >> > Finally an effect which confounds all the quality metrics is data >> > anisotropy: ideally the cut-off surface of significance in reciprocal >> > space should perhaps be an ellipsoid, not a sphere. I know there are >> > several programs for anisotropic scaling, but I'm not aware of any >> > that apply anisotropic resolution cutoffs (or even whether this would >> > be advisable). >> > >> > Cheers >> > >> > -- Ian >> > >> > On 27 January 2012 17:47, Jacob Keller <j-kell...@fsm.northwestern.edu> >> > wrote: >> >> Dear Crystallographers, >> >> >> >> I cannot think why any of the various flavors of Rmerge/meas/pim >> >> should be used as a data cutoff and not simply I/sigma--can somebody >> >> make a good argument or point me to a good reference? My thinking is >> >> that signal:noise of >2 is definitely still signal, no matter what the >> >> R values are. Am I wrong? I was thinking also possibly the R value >> >> cutoff was a historical accident/expedient from when one tried to >> >> limit the amount of data in the face of limited computational >> >> power--true? So perhaps now, when the computers are so much more >> >> powerful, we have the luxury of including more weak data? >> >> >> >> JPK >> >> >> >> >> >> -- >> >> ******************************************* >> >> Jacob Pearson Keller >> >> Northwestern University >> >> Medical Scientist Training Program >> >> email: j-kell...@northwestern.edu >> >> ******************************************* > > > > > -- > > ARKA CHAKRABORTY > CAS in Crystallography and Biophysics > University of Madras > Chennai,India > > > ------ > Randy J. Read > Department of Haematology, University of Cambridge > Cambridge Institute for Medical Research Tel: + 44 1223 336500 > Wellcome Trust/MRC Building Fax: + 44 1223 336827 > Hills Road E-mail: rj...@cam.ac.uk > Cambridge CB2 0XY, U.K. www-structmed.cimr.cam.ac.uk > -- ******************************************* Jacob Pearson Keller Northwestern University Medical Scientist Training Program email: j-kell...@northwestern.edu *******************************************