Dear Bryan,

The quick answer is no. As David Waterman mentioned, it has a default value for the gain for each type of detector that it can deal with.

A more detailed answer. An incorrect value for the gain can be indicated by values of the BGRATIO which differ significantly from unity (1.0). BGRATIO is the ratio of the rms variation in the background to the variation expected on the basis of Poisson statistics, using the gain to convert from digitised values in the image to X-ray photons. This is calculated for all measured spots, and binned as a function of intensity for each image measured (and printed in the full logfile). Mosflm prints a warning message if this differs from 1.0 by more than 10% and will suggest an "improved" value for the gain that should be used.

There are a host of caveats in this procedure. For example, if the images contains significant diffuse scatter around the Bragg spots, the BGRATIO may be above 1.0 ... this is probably the commonest effect, but does not mean the gain is wrong. If for any reason the mask definition (defining the boundary between background and spot) has not worked correctly so the spot extends into the background, this will also give a BGRATIO of one (in this case, the BGRATIO will tend to be close to 1.0 for weak spots but greater than 1.0 for strong ones). The boundary is controlled by the "Profile tolerance" parameters, which are sometimes set artificially high to help process images where the spots are not fully resolved.

This is why Mosflm does not automatically update its default value for the gain based on the BGRATIO.

As David has mentioned, this procedure also assumes that adjacent pixels are independent, which they most certainly are not (except possibly for some pixel detectors), due to the point spread function of the detectors and corrections that are applied to the raw images.

Does it matter ? The gain is used to identify outliers in the background plane determination (eg due to zingers, shadows, ice spots, hot pixels etc) which are rejected from the calculation, so if it is significantly in error this will introduce systematic errors in the integrated intensities. This can show up in the cumulative intensity distribution in Truncate if the gain is a very long way off. I have not done a proper study of this, but I think it would need to be out by more than 20% to have a significant effect. The gain is also used to calculate sig(I), however, the sig(I) values from Mosflm are adjusted in SCALA to reflect the true variation between symmetry related reflections so that providing the multiplicity is high enough for this to work correctly this will not have any real effect on the final merged data.

The bottom line is that the estimates for sig(I) that emerge for this procedure seem to be quite good, in that the correction factors that are subsequently applied in SCALA for cases where other systematic errors are small (ie no radiation damage, absorption etc etc) are very close to 1.0.

Best wishes,

Andrew




On 3 Mar 2011, at 20:34, Bryan Lepore wrote:

wondering if mosflm can automatically estimate the gain.

i.e. i gather it is still estimated the usual way.

-Bryan