Spots don't fall off with the inverse square law. It is a very easy experiment to do. Just take exposures at several distances and scale the data together, noting the correction for air absorption. A good reference for the underlying theory is Chapter 6 of M. M. Woolfson's book (1997). But briefly: Diffuse scattering falls off with the inverse square law as one would naturally expect from anything that spreads out in all directions. Spots, however, are "reflections" and their source point is the source of the incident beam, which is usually far away or at worst comparable to the sample-to-detector distance.

For those who do not like my "long posts" I now direct you to the "delete" button which should have come with your email client. Everyone else, read on.


The source of the confusion on this topic seems largely due to the fact that people keep dropping steradians from the units of intensity. ;) The units of diffuse scattered intensity are photons/steradian, and so the number of photons falling onto a pixel depends on the solid angle subtended by the pixel at the sample. The keyword to use to learn more about this is "differential cross section". The atomic form factor is an example of a differential cross section (after it is squared and multiplied by the Thomson cross section).

Spot intensity also arises from scattering and therefore fundamentally has units of photons/steradian. Since all the atoms in the crystal scatter coherently together, the intensity in photons/steradian can be very very large, but only in certain directions! The solid angle over which this high photons/steradian applies is proportionately small, and the integrated intensity is preserved. In fact, the fundamental "grain" of a spot is the reciprocal size of a mosaic domain (or Fresnel zone, whichever is smaller) which can be up to 10 microns across (10,000 A) and this corresponds to a solid angle of about 2.5e-9 steradian (using Bragg's law). This is a LOT smaller than a pixel at normal detector distances, and this is why the spots do not appear to spread out with distance. The angle- and area-integrated intensity (in photons/spot) does not depend on the distance (after correcting for air absorption). Think of your monochromator crystals. As long as the beam is monochromatic and the crystal is not bent or otherwise messed up, the reflected x-ray beam will not spread out any more than the incident beam did. In practice, of course, the spots are broadened by other things like mosaic spread, Nave disorder (unit cell variations, see Nave 1998), horizontal or vertical beam divergence, spectral dispersion and the detector point-spread function (PSF). The latter, along with the physical size of the illuminated crystal tends to be the dominant effect for good crystals, but obviously the PSF and crystal size does not depend on detector distance. All the others, taken individually, tend to spread out spots by an amount that is proportional to distance, not the distance squared. Hence there is an advantage in "photons/pixel from Bragg scattering" vs "photons/pixel from background" with increasing detector distance. Different factors become "important" as soon as their contribution to spot shape becomes comparable with the "starting" spot shape. All the effects are literally convoluted, so the result is the "RMS" of all contributing factors. However, if your crystal is messed up to the point of approaching fiber diffraction (say, with significant mosaic spread and significant Nave disorder), then the advantage of having a bigger detector farther away over a small, close detector starts to approach a "wash" in terms of signal/background.

It is an interesting point, however, that simultaneously increasing the detector distance and increasing the photon energy to keep the resolution at the edge of the detector constant has absolutely NO effect on the photons/spot from Bragg vs diffuse scattering. It took me a while to become convinced of this, and I still owe Colin Nave a few beers as a result. Nevertheless, the experimental evidence was given by Gonzalez, Denny & Nave in 1994.

HTH

-James Holton
MAD Scientist

Richard Gillilan wrote:
It seems to be widely known and observed that diffuse background scattering decreases more rapidly with increasing detector-to-sample distance than Bragg reflections. For example, Jim Pflugrath, in his 1999 paper (Acta Cryst 1999 D55 1718-1725) says "Since the X-ray background falls off as the square of the distance, the expectation is that a larger crystal-to-detector distance is better for reduction of the x-ray background. ..."

Does anyone know of a more rigorous discussion of why background scatter fades while Bragg reflections remain collimated with distance?


Richard Gillilan
MacCHESS

Reply via email to