Also depending on the detergent , If you have a detergent like
decyl-maltopyranoside ( DM) or any other glycoside based detergent you
can use the reaction with sulfuric acid  and phenol followed by
Absorption  measurement to quantitate as detailed in

Anal Biochem. 2005 Jan 1;336(1):117-24.
A colorimetric determination for glycosidic and bile salt-based
detergents: applications in membrane protein research.
Urbani A, Warne T.

Using a standard curve against the same detergent  you can quantitate
your detergent concentration in your final protein sample


Hari




On Fri, Oct 23, 2009 at 4:35 PM, Michael Matho <mma...@scripps.edu> wrote:
>
> Weikai,
>
> We did it using NMR but you asked for a simple way so I guess I'm out of 
> topic.
>
> Anyway, since I believe it is the most accurate method, here it is: using a 
> high detergent concentration stock solution you can assign resonance peaks to 
> your detergent molecule bonds.
>
> Then you can set up a standard curve using different known detergent 
> concentrations (for example from 10% down to 0.1%) by calculating the surface 
> of your peak(s) which is directly related to your detergent concentration.
>
> Each time you need to know the concentration of a new sample, you just need 
> to record the peaks, and use the three-click rule to deduct the unknown value.
>
> As a colleague answered you earlier, we noticed that a 50kDa cutoff withheld 
> a lot of detergent during concentration process and consequently your final 
> concentration might increase significantly. For example we started with 0.25% 
> DES and noticed increases of above 1%. Of course this will depend on the 
> concentration factor.
>
> This did not happen when using a 100kDa cutoff, and DES concentration remain 
> pretty much constant.
>
> Now, it will depend on your system: what detergent you are using, since micel 
> size and CMC are obviously the critical parameters here -- but also what 
> maximal cutoff you can use w/o loosing your membrane protein in the flow 
> through...
>
> Good luck,
> Michael
>
> ----- Original Message -----
> From: Patrick Loll
> To: CCP4BB@jiscmail.ac.uk
> Sent: Friday, October 23, 2009 1:12 PM
> Subject: Re: [ccp4bb] measure detergent concentration
> I'll second this.  We've done this as an exercise in NSLS Membrane Protein 
> Crystallization workshop for a few years, and it works like a charm. You can 
> stain in a warm iodine chamber and visualize by scanning the TLC plate on a 
> garden variety scanner (we use an inexpensive Canon LIDE that probably cost 
> less than USD 60 five years ago). We quantify the spot intensity with NIH 
> Image or equivalent, and get lovely linearity down to the CMC, spotting only 
> 1 uL of sample--so we haven't seen any need to concentrate.
> On 23 Oct 2009, at 3:41 PM, Edward A. Berry wrote:
>
> Only easy if you happen to have silica gel TLC plates and
> a chromatography jar lying around, perhaps from some
> phospholipid analysis:
>
> A strategy for identification and quantification of
> detergents frequently used in the purification of membrane proteins
> Laura R. Eriks, June A. Mayor, and Ronald S. Kaplan
> Analytical Biochemistry 323 (2003) 234–241
>
> This paper recommends spotting on a TLC plate and running
> beside standard amounts of the same detergent. From intensity/size
> of the detergent spot after developing you can bracket the detergent
> concentration. (And by the way they found that detergents are concentrated by 
> ultrafiltration). To increase sensitivity, speedvac a volume too large to
> spot on the plate, dissolve the residue in Me0H.
>
> Ed
> wei...@crystal.harvard.edu wrote:
>
> Hi Folks:
>
> After concentrating a membrane protein, is there a (easy) way of measuring
>
> the detergent concentration in the sample?
>
> Regards,
>
> Weikai
>
> ---------------------------------------------------------------------------------------
>
> Patrick J. Loll, Ph. D.
>
> Professor of Biochemistry & Molecular Biology
>
> Director, Biochemistry Graduate Program
>
> Drexel University College of Medicine
>
> Room 10-102 New College Building
>
> 245 N. 15th St., Mailstop 497
>
> Philadelphia, PA  19102-1192  USA
>
> (215) 762-7706
>
> pat.l...@drexelmed.edu

Reply via email to