<lurk mode>

 

The answer I think is yes, but partially.  Colin Nave published a very
nice paper on this, "A description of Imperfections in Protein Crystals"
in Acta Cryst D54, 848-853 (1998).  The 'real' mosaicity is a
convolution of the variation in cell dimensions within crystallites or
domains making up a crystal (observed with X-ray topography), the
misalignment of those domains (measured with very parallel beams and
fine phi slicing) and finally the volume variation.  The measured
mosaicity is typically something else entirely as often, even for many
cryocooled samples, the mosaicity of the crystal is masked by the
contribution of the geometrical and spectral effects of the
instrumentation and beam. This is especially the case on a lab source. 

 

</lurk mode>

 

Cheers,

 

Eddie  

Edward Snell Ph.D.
Assistant Prof. Department of Structural Biology, SUNY Buffalo,
Hauptman-Woodward Medical Research Institute
700 Ellicott Street, Buffalo, NY 14203-1102
Phone:     (716) 898 8631         Fax: (716) 898 8660
Email: [EMAIL PROTECTED]  Telepathy: 42.2 GHz

Heisenberg was probably here!

________________________________

From: CCP4 bulletin board [mailto:[EMAIL PROTECTED] On Behalf Of
Jacob Keller
Sent: Tuesday, September 09, 2008 5:24 PM
To: CCP4BB@JISCMAIL.AC.UK
Subject: Re: [ccp4bb] Are cell parameter shifts real?

 

In thinking about one or two responses, I am wondering whether mosaicity
could be the result of having a distribution of each of the cell
parameters within one crystal? Is that what mosaicity really is? That
is, if axis a has a broad distribution of lengths, due to being caught
in the act of changing to the lower-temp state, the spots will be
streaky along a*, and so for the others, resulting in an overall
"mosaicity" value which encompasses all of these distributions, being
used mainly as a data-processing expedient? The actual physical
phenomenon of mosaicity has always puzzled me, but perhaps this is an
answer?

 

Jacob

 

*******************************************
Jacob Pearson Keller
Northwestern University
Medical Scientist Training Program
Dallos Laboratory
F. Searle 1-240
2240 Campus Drive
Evanston IL 60208
lab: 847.491.2438
cel: 773.608.9185
email: [EMAIL PROTECTED]
*******************************************

        ----- Original Message ----- 

        From: Jim Pflugrath <mailto:[EMAIL PROTECTED]>  

        To: CCP4BB@JISCMAIL.AC.UK 

        Sent: Tuesday, September 09, 2008 3:54 PM

        Subject: Re: [ccp4bb] Are cell parameter shifts real?

         

        How would you tell the difference between a unit cell shift and
a wavelength shift when collecting diffraction data at a synchrotron
beamline?  Well, all the cell length would scale by the wavelength, so
that would be one hint that the wavelength changed.  If a got longer and
c got shorter, then it would be less likely to be a wavelength shift.

         

        The crystal-to-detector distance can easily change if the
crystal rotation device (i.e. goniometer) rotates the crystal without
keeping it at the exact same crystal-to-detector distance.  This could
easily happen if the crystal is not centered at the rotation point or if
more crystal volume rotated into the beam during your experiment.
However in a typical experiment the crystal is not going to move by 5 mm
and remain in the X-ray beam, so you would not expect your distance to
change by 5 mm.

         

        Jim

        
________________________________


        From: CCP4 bulletin board [mailto:[EMAIL PROTECTED] On
Behalf Of Jacob Keller
        Sent: Tuesday, September 09, 2008 3:08 PM
        To: CCP4BB@JISCMAIL.AC.UK
        Subject: [ccp4bb] Are cell parameter shifts real?

         

        ...

         

        The bottom line questions are: 1. given that there really are
cases of cell shifts, and that there are also probably experimental
artifactual changes, how is one to decide what to do? 2. Can there be
(or what is?) a plausible mechanism for these shifts?

         

        Jacob Keller

         

Reply via email to