Note the density of air is approximately 1000 times less than a protein
crystal. The total scatter for a beam going through a 50 micron thick
crystal will be similar to that from 50mm air. Most beamlines will have
a path length less than this but nevertheless the air scatter will be
significant with small crystals.
In principle, with smaller beams one can have smaller beamstops nearer
the sample thus reducing the path length through the air.

 Colin
-----Original Message-----
From: CCP4 bulletin board [mailto:[EMAIL PROTECTED] On Behalf Of
Richard Gillilan
Sent: 26 November 2007 16:24
To: CCP4BB@JISCMAIL.AC.UK
Subject: Re: [ccp4bb] To bathe or not to bathe.

>

I just noticed this thread. I should make a few comments.

We regularly provide microbeam with and without Helium here at MacCHESS.
Yes, there are cases in which microbeam can give you good diffraction on
large crystals when a larger beam cannot. Just last week we had a user
group collecting on column-shaped crystals about 40 microns wide by
200-300 microns long. When exposed using a 150 micron diameter beam, the
diffraction patterns were a mess, producing multiple lattices. When they
exposed the crystals using a 20 micron beam, they were able to find
enough "sweet spots" with single lattice to give good results. I can
only offer anecdotes at this time, but, based on user experiences so
far, this is not uncommon at all.

I don't usually think of microbeam alone as a significant reduction in
background except that one avoids hitting solvent (which is a major
source of scatter!).  While the beam scatters less air due to its small
size, microbeams also can have much increased flux which compensates for
the smaller size ... so I don't think there will be less air scatter
unless you are just using an aperture with no flux gain.

Regarding helium: we regularly use helium at 95K combined with a helium
enclosure. The setup is awkward for manual sample mounting, but quite
convenient for automounting. The main advantage here is that direct beam
scattering with air is nearly eliminated.

How important is this effect? It depends upon how much free path there
is between the end of the optic (slits, etc) and the beamstop.  
It also depends upon how small your sample is and how strongly it
diffracts. In a carefully-controlled experiment using the identical
crystal and orientation range with both nitrogen and helium, I saw a
signal-to-noise improvement by over a factor of 3 in the 2.5 Angstrom
range and lower. At higher resolution the benefit decreases, but still
looks no worse than a factor of 2. This is with a 50 micron crystal
illuminated with an 18 micron beam.

I am currently working on guidelines for when helium and microbeam are
necessary (based on both simulations and explicit measurements).  
At the present time, my feeling is that crystals below 50 micron can
certainly make the extra hassle worthwhile. It really depends upon how
badly you want that extra resolution. In the case above, it pushed the
resolution from above to below the 2.0 Angstrom mark based on I/SIG.

As for radiation damage, I do think it is important that a beam
intensity profile be as flat at the top as possible (not a sharp
hotspot) .. so defocusing a little may be useful. This is not usually a
factor that regular users have any control over. I don't have any hard
data on this.



Richard Gillilan
MacCHESS
<DIV><FONT size="1" color="gray">This e-mail and any attachments may contain 
confidential, copyright and or privileged material, and are for the use of the 
intended addressee only. If you are not the intended addressee or an authorised 
recipient of the addressee please notify us of receipt by returning the e-mail 
and do not use, copy, retain, distribute or disclose the information in or 
attached to the e-mail.
Any opinions expressed within this e-mail are those of the individual and not 
necessarily of Diamond Light Source Ltd. 
Diamond Light Source Ltd. cannot guarantee that this e-mail or any attachments 
are free from viruses and we cannot accept liability for any damage which you 
may sustain as a result of software viruses which may be transmitted in or with 
the message.
Diamond Light Source Limited (company no. 4375679). Registered in England and 
Wales with its registered office at Diamond House, Harwell Science and 
Innovation Campus, Didcot, Oxfordshire, OX11 0DE, United Kingdom
</FONT></DIV> 

Reply via email to